Sample Size for Stability Testing

C

clatoz

#1
Hello,

We make primarily Class I, with a few Class II devices. For our stability program, we are required to test 60 samples at each time point for a Pass/Fail test, which gives an upper probability of failure of <5% at the 95% confidence interval. For stability tests which generate numerical results, where we will use regression analysis, we are required to test 17 replicates at each time point. This seems extreme, as from what I've seen in the Pharma industry, typically only 3-5 samples are tested at each time point. Just curious what sample sizes others in the device industry use for their stability testing? Also, if anyone could provide some references for sample size selection for stability testing. I couldn't find anything in the ICH guidelines or ASTM F1980. Thanks in advance.
 
Elsmar Forum Sponsor

Bev D

Heretical Statistician
Staff member
Super Moderator
#2
sample size is always dependent on the standard deviation of the data (among other things). So 17 doesn't actually seem very extreme for stability testing where regression will be applied. you are not merely trying to estimate the MEAN but the results of the individual values. your customer uses individual devices. It does you no good for the average to be above the end of life specification but 49% of your devices are below...:)

I know some companies use smaller sample sizes but this doesn't mean they are correct
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Hello and welcome to the Cove :bigwave:

60 for attribute and 17 for variable doesn't sound excessive to me.

I once heard that the FDA expects ~30 if you're trying to fit one of the standard distributions to predict extremes. They might take less (depending on context) but are normally comfortable with 30.
 
Thread starter Similar threads Forum Replies Date
P Unrealistic Packaging Validation Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 13
J Control chart for huge sample size Statistical Analysis Tools, Techniques and SPC 9
S Is there a standard for sample size during R&D phase Other ISO and International Standards and European Regulations 18
J Sample size definition in an Automotive SMT pilot lot run Misc. Quality Assurance and Business Systems Related Topics 1
E Correct way to certify hydrostatic testing when it is not 100% (and Sample Size) Various Other Specifications, Standards, and related Requirements 6
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
D Determining of sample size for 'Operational Qualification' AQL - Acceptable Quality Level 3
F Bayes Success run Theorem for sample size during OQ&PQ Qualification and Validation (including 21 CFR Part 11) 4
I Device modifications - Clinical sample size rationale EU Medical Device Regulations 5
J Sample size for creating a data base as a reference to a tested variable Other Medical Device and Orthopedic Related Topics 6
T Sample Size for Design Validation Design and Development of Products and Processes 4
C AS9138 Sample Size Determination AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
V Acceptable maximum RSD (relative standard deviation) for an sample size Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
E Sample size for design verification of variable in single use device Design and Development of Products and Processes 19
R Sample size for clinical validation/investigation EU Medical Device Regulations 4
R Process Validation sample size selection Statistical Analysis Tools, Techniques and SPC 0
J Interesting Discussion Sample Size Determination for Medical Device Other Medical Device and Orthopedic Related Topics 9
P Is it possible to make an educated decision using a very very small sample size? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
F Sample Size for Biocompatibility Tests Other Medical Device Related Standards 4
V Firmware Verification Testing & Sample Size Software Quality Assurance 1
R Determining Sample Size for Medical Device Component Validation Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
S Surveillance Sampling Test - Determining Sample Size Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
H Sample Size Calculation (Confidence Interval and Reliability) - Medical Devices Reliability Analysis - Predictions, Testing and Standards 6
P Cleaning and Disinfection Validation Sample Size of 3? Other Medical Device Related Standards 2
V Sample Size Calculation for Image Analysis - Microscopy Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
G Sample Size, Significant Figures, Scale General Measurement Device and Calibration Topics 3
L Heated Sealed Packages - Sample Size for OQ (Operational Qualification) and PQ Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
P Frequency and Sample Size Requirements for an MSA Studies Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
V When and How to apply the Bayes Success Run approach for Sample Size Determination Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
N Sample Size Justification for Medical Device Shelf-Life Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
C Process Qualification Sample size Document Control Systems, Procedures, Forms and Templates 6
G Sample size for IEC 60601-1-11 environmental tests IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
G IEC 60601-1-2 Sample Size issue Other Medical Device Related Standards 1
A How to calculate Mean and standard deviation without sample size in Minitab Using Minitab Software 5
D What is the minimum Sample Size for Weibull Analysis Reliability Analysis - Predictions, Testing and Standards 12
M Calculating Cpk when sample size equals to 1 Capability, Accuracy and Stability - Processes, Machines, etc. 12
F Power and sample size for factorial design Using Minitab Software 4
S Sample Subgroup Size for NP/P chart - nPbar>=5 Statistical Analysis Tools, Techniques and SPC 4
U Sample size to prove parts are good with 99.73% confidence Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
F Sample Size to get Statistically Valid Data Measurement Uncertainty (MU) 6
S Low Sample Size for Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
S SPC Card for One Sample Size - Customer requires for all critical dimensions Statistical Analysis Tools, Techniques and SPC 8
S Determining sample size for inspection to achieve x% confidence re defects Misc. Quality Assurance and Business Systems Related Topics 10
S Determining Sample Size - AQL & LQ AQL - Acceptable Quality Level 10
B Bayes Success-Run on Attribute Data - Determining Sample Size Statistical Analysis Tools, Techniques and SPC 2
J Flammability: Testing Burn Rate with Improper Sample Size fmvss 302 Various Other Specifications, Standards, and related Requirements 1
H Factory Dictated Sample Size Justification Statistical Analysis Tools, Techniques and SPC 9
M Design Verification: "Difference" in Sample Size Calculations? Design and Development of Products and Processes 1
M Calculating Adequate Receiving Inspection Sample Size Statistical Analysis Tools, Techniques and SPC 2
S Salmonella Test Sample Size Food Safety - ISO 22000, HACCP (21 CFR 120) 3

Similar threads

Top Bottom