Sampling plan for orthopedic implant - Process Validation
enorht
I am working as a process validation engineer in Europe. Our notified body recommended us to perform process validation when the output cannot be verified at 100%.
The recent changes in standards (for example ISO 19227 for the validation of cleaning process of orthopedic implants, ...) ask to use "statistical techniques with rationale for sample sizes". And i have difficulties to define a statistical rationale because:
- we have very small lot (the batch size is between 10 and 100 pcs around)
- we have no process history
- all my processes are outsourced and my subcontractor are really bad in quality field, they have never validated a process and I am new in process validation field too
I have attributes data (GO/NO GO) and variable data (length for example).
1) Attribute
For the attributes, i need to choose an AQL and a LTPD. For priority, we are focusing on validating all the critical output (TOC, THC, cytotoxicity, ...). For that, we have a defined an AQL of 0.25% (no process history, no rationale) and for the LTPD, I do not know how to choose it...
2) Variable
We plan to perform a capability study and plan to prove that the process is capable (Pp and Ppk > 1.33). However, after that, i do not know how to choose my sampling... Do i have to the ppk to a LTPD and an AQL?
Do you know how to choose LTPD and AQL acceptable for critical default (sterility, cleanliness, critical dimensions, ....) when the lot size is really small (orthopedic implants, ...)?
Thanks,
planB
enorth,
IMHO, you cannot succeed with these prerequisites.
Without diving into specifics:
  • in case you a running short of actual product for your validations, can you define and build worst-case (dummy) products, in sterilization often referred to as "PCD (Process challenge device)" that represent you actual products in the relevant process characteristics?
  • In case you are planning to use standard sterilization methods (such as irradiation or ethylene oxide), the related standards are pretty specific about required sample sizes.
HTH,
Gerhard
enorht
Hi Gerhard,
I have no issue for the processes like the gamma sterilization because the standard is very clear on the methodology and the sampling.
However, it is not the case for the machining, laser marking, packaging processes where no sampling plan is described.
It is asked to use "statistical techniques with rationale for sample sizes ". I do not know which AQL and LTPD to use for my critical characteristics. And which sampling to use for small batch production...
enorht
Hello, a little bump since I began some testings...
Context: It is very difficult to explain the necessity of validation to our supply so we do not have a solid OQ (meaning no identification of critical parameters such as speed, ...). I just have a machine capability performed on a complex and aluminum product using all the axes (8 or 9) of one equipment. As all the equipment are of the same brand, we decided to accept this "OQ".
I am currently performing a performance qualification of CNC equipments. They are used for the manufacturing of orthopaedic implants (screws, rods, nuts, ...). We have decided to manufacture for each kind of product (based on their physical dimensions) at least three batches representative of the routine. We already identified for each of them critical dimensions and critical aspect. Most of them are measured but few of them are controlled by a GO/NO test or visual aspect.
we agreed with the supplier for the three first batches of each product reference to control and record:
- 100 measures for each dimension per batch --> my aim was to have enough data to perform a capability study and calculate the lower bound of capability index with a confidence level of 95%. I saw that 100 measures are a good sampling to have relevant confidence in the capability indexes (and i will have 300 measures if there are no variation inter lot). But i did not find relevant sources.
- 100% of control GO/NO GO and 100% for the visual aspect. honestly, the production was launched without ending the protocol writting so we decided to have the maximum of values. but I do not know if I have to write a rationale with an approach of LTPD/AQL for critical defauts or do i need to also perform a capability study (with minitab). How can i justify this sampling in my protocol?
The aim is to validate the manufacturing process by showing that:
- for the dimensions measured, the process is stable and have a pp/ppk lower bound at 95% of confidence > 1.33. After that, should I calculate the Z score, calculate the %ppm and link to the AQL/LTPD too?
- for the characteristics controlled by visual inspection or GO/NO GO controls, i do not know if i i have to use capability studies too or claim a NQA/LTPD %...
Thanks for your help,
Wolf.K
Did you get some insights into which AQL ald LTPD to set, and how to calculate sample sizes? Currently I have the same problem - I "inherited" a manufacturing process and try to understand, what they have done in the past... And the standards are not very clear and not very helpful. :-(
L Sampling Plan Risks AQL - Acceptable Quality Level 6
R Reduced sampling plan for sterial products APQP and PPAP 0
B AS9100D Sampling Plan AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
Nicole Desouza Sampling plan for a customer who wants AQL 1.0 (per ANSI Z1.4) AQL - Acceptable Quality Level 5
M Sampling Plan for Alumin High Pressure Die Castings Manufacturing and Related Processes 0
S How to make a Sampling Plan - Conveyor belts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
S Help develop sampling plan for prototype samples (medical devices) Statistical Analysis Tools, Techniques and SPC 7
M Zero Defect Sampling Plan Standards Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
L ISO 2859-2 LQL sampling plan linked to %confidence/%reliability Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
K Sampling Plan for transparent tubular plastic parts from 32 different cavities Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
A AQL Sampling Plan SOP / WI - Medical Device Industry Inspection, Prints (Drawings), Testing, Sampling and Related Topics 14
J Sampling Plan Questions from the CQE Practice Exam Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Sampling Plan for for Chemical Analysis Testing of Medical Devices EU Medical Device Regulations 3
V Sampling Plan for Destructive Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
J Can ANSI Z1.4 sampling plan be used for sampling alone without the Acc/Rej criteria? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
C Sampling plan for destructive tests on small lots ISO 13485:2016 - Medical Device Quality Management Systems 20
B Designing a Sampling Plan - AQL AQL - Acceptable Quality Level 19
N Use of arrows in ISO 2859-1 Table 1 and 2-A sampling plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S AS9100 Sampling Plan Requirements - 8.2.4 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S Military Sampling Plan 105E vs. ISO 2859-1 AQL - Acceptable Quality Level 2
kedarg6500 Acceptance Sampling Plan for Variables - Product Performance Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
M Manufacturing Process Validation Destructive Testing Sampling Plan Advice Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
S Where could I get reference to an OQA Sampling Plan of 1 per 1000 ? Quality Manager and Management Related Issues 4
R What to Report when Using C=0 Sampling Plan Supplier Quality Assurance and other Supplier Issues 15
R Results: Monte Carlo Simulation of Acceptance Sampling Plan AQL - Acceptable Quality Level 3
G Sampling Plan according to SAE ARP9013 C=0 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
G Sampling Plan for Inspection of Metal Parts after CNC Machining Inspection, Prints (Drawings), Testing, Sampling and Related Topics 34
S Class III Sampling Plan - Explain AQL vs. 95%C & 90%R Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
G Sampling Plan for a Performance Qualification of an Inflation System Qualification and Validation (including 21 CFR Part 11) 2
S How to establish a Sampling Plan from the Rework percentage? AQL - Acceptable Quality Level 6
Q Sampling Plan for Paper and Tape Industry Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
C Medical Device Tracking Sampling Plan using MIL-STD-105E Other US Medical Device Regulations 4
P Lot Acceptance Plan that specifies BS 6001 Double Sampling Plans, general inspection Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
A Destructive Testing Failure Mode Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 5
B Best Attribute Data Sampling Plan - Early Production Containment Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
B In-Process Sampling Plan for a Secondary Package Adjustment Process Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
Proud Liberal Making a simple Sampling Plan by Variables Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
E How to determine the Lot Size for the OQ Sampling Plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
L Electrical Cable Harness Quality Control - Sampling Plan for Pull Test Manufacturing and Related Processes 4
G Sampling Plan - Sample Size Justification for a Bottling Cleaning Process Validation Qualification and Validation (including 21 CFR Part 11) 2
I Alpha and Beta in the Sampling Plan for ISO 2859 & ISO 3951 Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
V Sampling Plan for Final Inspection and Process Verification Inspection, Prints (Drawings), Testing, Sampling and Related Topics 21
M Sampling Inspection at QA - Inputs on the following sampling plan Statistical Analysis Tools, Techniques and SPC 3