OK, you have a point.

But IF not... and IF we only need to be accurate within, ooh, let's say a couple of inches (she says, bowing to the imperial measurement system still in use by you guys), would you still argue that tape measures need calibration?? I sure wouldn't.
Yep, IF the measurement is specifically documented and required
From whence comes this requirement?
 
Elsmar Forum Sponsor

Randy

Super Moderator
#22
From whence comes this requirement?
It's a challenge because one really has to understand a system approach, but look at 7.6a to begin with...then think 7.2.1, 8.2.3, 8.2.4 and how they can apply to having measure something....There's a difference between taking a measurement for self assurance and taking a measurement that has been specifically defined by a customer, a regulation or the organization itself.

"Customer X says......no longer than 63 or shorter than 58" then measuring is necessary and the instrument used falls under 7.6

If a measurement, any measurement MUST be taken then the instrument used has to be either individually calibrated or verified against something else that is calibrated and traceable to either a international or national measurement standard (like NIST) or in the absence of some standard then the calibration process itself must be recorded.

It's all very plain in balck and white...No interpretation needed

It is touchy ground at times though because sometimes we see ambiguous requirements such as "Grey" in color....Colors can be measured, but how many different "colors" of Grey are there? Which Grey is the correct Grey?
 
C

cereo

#23
Again, this is not a customer requirement, nor a regulatory requirement. The shipping company (a supplier, not a customer) charges extra if we exceed 20 feet (no acceptance limits provided) so our procedure states to verify that the load is under 20 feet. We are careful to stay well under that. Nothing falls out of the sky, nobody gets hurt, product isn't affected and the customer never knows, we just lose some profit margin if we get careless and don't check before shipping.
We DO calibrate our scales as these are used to meet DOT requirements.
 

Randy

Super Moderator
#25
Again, this is not a customer requirement, nor a regulatory requirement. The shipping company (a supplier, not a customer) charges extra if we exceed 20 feet (no acceptance limits provided) so our procedure states to verify that the load is under 20 feet. We are careful to stay well under that. Nothing falls out of the sky, nobody gets hurt, product isn't affected and the customer never knows, we just lose some profit margin if we get careless and don't check before shipping.
We DO calibrate our scales as these are used to meet DOT requirements.
YOUR PROCEDURE SAYS? Guess what? Self inflicted wound and you've got to control the measuring tapes.

Your procedure is a "Self Created" requirement that must be fulfilled and verifiable as to its accomplishment and the shipping companies requirement must also be accomplished and verifiable as well.

Why?

Because if you don't meet it, the customer doesn't get their stuff! UH-OH, you didn't achieve CUSTOMER SATISFACTION because the shipping load was tooooooo long by %age of an inch.
 
J

JaneB

#26
YOUR PROCEDURE SAYS? Guess what? Self inflicted wound and you've got to control the measuring tapes.

Your procedure is a "Self Created" requirement that must be fulfilled and verifiable as to its accomplishment and the shipping companies requirement must also be accomplished and verifiable as well.
Randy is completely, utterly, 100% on the money here (as so often).

A requirement is a requirement. It doesn't matter in a sense who created the requirement, as currently it is a requirement in your system. And thus you must control the tapes.

Sensible solution: change that requirement. (Assuming you can, of course).
 
J

JaneB

#27
If a measurement, any measurement MUST be taken then the instrument used has to be either individually calibrated or verified against something else that is calibrated and traceable to either a international or national measurement standard (like NIST) or in the absence of some standard then the calibration process itself must be recorded
I agree that in such a case the equipment comes under 7.6.

But what about the:

Where necessary to ensure valid results
bit in the 3rd para of 7.6?

If there's fairly wide latitude / tolerance (eg, measurement can be + or - a couple of inches or so) and one is using regular, everyday tape measures, say, then are you still holding that ALL equipment must be calibrated as you describe?
 
C

ChrissieO

#28
Haven't been on the cove for a few weeks and just catching up. I have read the first two pages of responses to this and as usual agree mostly with Jane, Randy, Marc, Andy and the rest of the usual suspects.

Just to put my "twopenneth" (its a brit thing) in. The first thing that hit me from the OP is that the auditor is a typical "tick box" auditor with no signs of thinking outside the box or evaluating the processes.

Yes he has made some valid points but he has also made some points that are abosultely meaningless and of no added value.

As Randy pointed out, some of this is self inflicted injury. Are your procedures written for a perfect "ISO" world and then you are expected to follow them? If they are not realistic, revist them and update them to what you actually do. Were they written by some quality geek that trained in the dark ages or were they written by the people that actually use them.

Paper stamped documentation - do still people still use paper:notme:

But as it is the companies own standard there is not much the OP can do apart from ensure that the dinasaur of an auditor is put out to grass very soon.

Sorry about my rant but I fed up of seeing thread after thread about auditors in the world of 9001 and TS who are still nit picking at minor issues and using check box mentality auditing..................thought this went in to "Room 101" (is that another brit thing?) with QS9000!!!!!!!

Get over it move on and worry about things that actually effect the business.....thay are called PROCESSES!!!!!!!!!

ok :mad: rant over....honest;)

Cx
 
Last edited by a moderator:

Jim Wynne

Staff member
Admin
#29
It's a challenge because one really has to understand a system approach, but look at 7.6a to begin with...then think 7.2.1, 8.2.3, 8.2.4 and how they can apply to having measure something....There's a difference between taking a measurement for self assurance and taking a measurement that has been specifically defined by a customer, a regulation or the organization itself.

"Customer X says......no longer than 63 or shorter than 58" then measuring is necessary and the instrument used falls under 7.6

If a measurement, any measurement MUST be taken then the instrument used has to be either individually calibrated or verified against something else that is calibrated and traceable to either a international or national measurement standard (like NIST) or in the absence of some standard then the calibration process itself must be recorded.

It's all very plain in balck and white...No interpretation needed

It is touchy ground at times though because sometimes we see ambiguous requirements such as "Grey" in color....Colors can be measured, but how many different "colors" of Grey are there? Which Grey is the correct Grey?
I agree that it's very plain. We won't get into the interpretation thing again, :D but the plain language of 7.6 says that calibration is required when necessary to ensure valid results and nowhere in the standard does it say that calibration is required if a measurement is required.
 

Randy

Super Moderator
#30
Let me put it real, real simple. The need for calibration or validation of devices in instances exactly like described has been appealed to an accreditation body by an organization that was audited and the appeal was lost. The NC stood as written;)
 
Thread starter Similar threads Forum Replies Date
T Sanity Check: 6 Sigma with a Complex Business Process Six Sigma 5
D Sanity Check - Audit of Distribution Centers required by New Registrar ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 19
P Sanity Check on AQLs (Acceptable Quality Level) AQL - Acceptable Quality Level 3
Q ISO 9001 - 3 BIG questions to help a Newbie's Sanity ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
P U-chart Sanity/File Review - Is the chart constructed properly? Statistical Analysis Tools, Techniques and SPC 2
P Audit check for IT company (ISO 9001) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
M Management review check-list Management Review Meetings and related Processes 3
M Documents to be included in 1.3 Material check in PPA report APQP and PPAP 4
D FAA inspection at my repair station - Rating Check Federal Aviation Administration (FAA) Standards and Requirements 0
G MSA check list to audit IATF 16949 - Automotive Quality Systems Standard 10
M How Supplier Capacity check is done by competitors Manufacturing and Related Processes 0
M IATF 16949 8.5.1.3 Verification of job set-ups - Do we need secondary check? IATF 16949 - Automotive Quality Systems Standard 7
J Standard used to determine if check fixture gauge is acceptable Reliability Analysis - Predictions, Testing and Standards 2
M MDR requires the NB to perform "tests in order to check that the quality management system" EU Medical Device Regulations 3
G Is it common to do a repeatability check during calibration of an instrument? General Measurement Device and Calibration Topics 5
V CAPA effectiveness check Nonconformance and Corrective Action 3
A Statistical Analysis - Check if these organisms at different concentrations affect the growth of wheat seedlings Using Minitab Software 4
G Oversized Threads - What sized stab pin would be best to quickly check for oversized thread holes General Measurement Device and Calibration Topics 5
G Stab Pin Size For Check Gage Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
B AS9110 Audit Check Sheet wanted Excel or Word Document AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
J AQL for Effectiveness Check - What AQL tiers to use on this type of sampling? Nonconformance and Corrective Action 9
G From an ISO 17025 auditor perspective must micrometer calibration check anvil flatness? General Measurement Device and Calibration Topics 4
F Position tolerance check gage manufacturing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
S The procedure for replicate check in calibration lab as per ISO 17025 General Measurement Device and Calibration Topics 2
F Design check list for Cable and Wiring Harnesses Manufacturing and Related Processes 6
R Mass production handover check sheet (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 1
bobdoering Handy training tip for SPC - the Check Engine Light Analogy Statistical Analysis Tools, Techniques and SPC 0
R Please help me with attached copy of AS 9100D internal audit check sheet AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
G What to Check - Set Up vs. In Process Inspections Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
B FDA Philippines has no check or balance outside the health department Supplier Quality Assurance and other Supplier Issues 3
Crusader Product Completion, Birth Cert, Product Check sheet - packaged with product Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
C Serialization at Final Quality Check ISO 13485:2016 - Medical Device Quality Management Systems 3
T Example System & Process Audit Check Sheet for IATF 16949 IATF 16949 - Automotive Quality Systems Standard 1
L Signal-to-Noise (S/N) Ratio formula check Design and Development of Products and Processes 2
M AS9100 7.6 - Check Method - PO Information AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
S GR&R for a Stopwatch used to check a built-in Machine Timer Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
R ISO 13485:2016 Clause 6.2 - Effectiveness Check ISO 13485:2016 - Medical Device Quality Management Systems 1
P Looking for a company local in Wyandotte, Mi, to come in check noise levels Manufacturing and Related Processes 2
M Difference between a "check" and a "test" Manufacturing and Related Processes 7
M Proper Gauging - GD&T - Attribute gage to check true positioning on the shop floor Inspection, Prints (Drawings), Testing, Sampling and Related Topics 6
S How to check runout on Id to ID diameters Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
Y Is it possible to check the authenticity of a CFDA registration certificate? China Medical Device Regulations 7
M More than canned audit check-sheets? Auditing the Engineering Department Process Audits and Layered Process Audits 5
P Searching for a complete audit check list based on ISO14001:2004 ISO 14001:2015 Specific Discussions 1
V Internal Audit using Plan, Do, Check, Act - Way to go? Internal Auditing 11
S Elisa kits for poultry disease check up companies US Food and Drug Administration (FDA) 3
F Check List and Questions for Auditing Monitoring and Measuring Devices General Measurement Device and Calibration Topics 1
E How to perform Discrimination check for Dial Indicator? General Measurement Device and Calibration Topics 2
B How to check the parallelism error of v axis w.r.t to lateral surfaces General Measurement Device and Calibration Topics 1
L Most appropriate Auditor to check the FMEA & PCP of a Manufacturing Process IATF 16949 - Automotive Quality Systems Standard 5

Similar threads

Top Bottom