Scope in procedure and content for 7.6 Monitoring and Measuring Equipment
Q
QAMTY
Hi all.
I'm in the phase of implementing an ISO 9001:2008 and have some questions in 7.6, please share your ideas. 7.6 require to comply with
a) be calibrated or verified, or both, at specified intervals, or prior to use, against measurement standards
traceable to international or national measurement standards; where no such standards exist, the basis
used for calibration or verification shall be recorded (see 4.2.4);
b) be adjusted or re-adjusted as necessary;
c) have identification in order to determine its calibration status;
d) be safeguarded from adjustments that would invalidate the measurement result;
e) be protected from damage and deterioration during handling, maintenance and storage.
and some others...
But I have heard that in calibrating the equipment with external suppliers some special tests have to be done,
e.g. R&R graphs, repeatability, reproducibility, etc.
My doubt is:
As I understand, I only have to take care of issues required mentioned on top (ISO requirements), additionally, I need to have the competency to perform the measurements , I only have micrometers and rules.
I have a procedure where I mention everything, but I'm not mentioning, R&R, repeatability, reproducibility,etc.
Do I have to mention that?
I think that these test are to be done only by the supplier of calibration in order to determine the status of a an equipment,
Please shed light on me., what I have to mention in my procedures?
My best regards
drgnrider
? But I have heard that in calibrating the equipment with external suppliers some special tests have to be done,
My doubt is:
My best regards
Special tests ? no these are not required. You may, however, perform these tests, (or have them performed), to prove a reason for using a calibration interval greater than the manufacturer recommended interval, (typically one year). Neither I, nor my predecessor, ran these tests, but he had enough historical data to make a decent estimate of how the gages are faring using the current intervals.
:2cents:Personally, I would not mention these tests, other than possibly as a side note 'these are tests that can be used to...'; depending how you word them in your procedures then that is what you must do. If they are not in your procedure, they are options that you can use if, and when, you deem necessary.
Competency ? this can be something as simple as someone with experience deeming you competent after training you. Most calibration is using the tool to measure against a known standard and recording the result.
- Before I started as a machinist I had never used a micrometer, I became proficient using them on the shop floor and now I am only using them on known standards, not parts, and recording my results (aka: calibrating). I added to my skills by seeing what was required of calibration and adjusting our calibration procedures accordingly (they were missing some steps).
As for the standard and your procedure - as long as your procedure addresses all the points of the standard, that is the minimum that is required, anything beyond that up to you... you just have to be doing that.
drgnrider
Calibration & ISO Coordinators
Hi all.
My doubt is:
My best regards
[/SIZE][/FONT][/SIZE][/FONT]
If you are going to calibrate something, to do it properly you should know the measurement uncertainty. This why an accredited lab reports this factor. Without it - like checking a caliper against a gauge block - is just verification, NOT calibration. You have to determine whether calibration or verification is needed. It sounds to me that you might need to get some education on calibration before doing much else. To suggest that competency comes from training is putting things around the wrong way - defining competencies comes before taking action - including training - to develop those skills.
So, if YOU don't know if you need to perform any of the analyses such as R & R, linearity, bias, stability etc, then clearly you need to be competent in those things. I'd recommend getting help from someone who IS competent in the subject (some labs also run training)
Q
QAMTY
Thanks drgnrider and Andyn
Good comments
Even if I'm not going to be a guru in calibration, at least
I have to know the basics of equipment calibration terms and definitions
Thanks again
Actually, a procedure for calibration isn't required. Gauge R&R isn't required. Measurement uncertainty isn't required.
All are good to do, but is above and beyond ISO 9001.
Any of them could be a customer requirement though.
Johnson
Hi all.
and some others...
My doubt is:
My best regards
Firstly the concept and purpose of calibration and Gage R&R should be clarified. To make it simple (but the statement is not accurate enough):
The calibration is to check the correctness and accuracy of a measuring device compared to it's specification. It uses "master" or "reference standard". Normally done by the third party. Your parts are not used.
But Gage R&R is used to check if one selected measuring device is suitable /capable to measure your parts with defined tolerance. It use your parts and the measuring is done by your inspectors or operators. For example, you should not use a caliper (with resolution of 0.01 mm) to measure the parts with tolerance of 10+/-0.020 mm, you may use micrometers (with resolution of 0.001 mm)
Based on above explaination and look at ISO9001 7.6 again. It could be concluded:
A measuring device need to "be calibrated or verified", but R&R is not mentioned. R&R is required in automotive OEM or ISO/TS16949.
I would suggest you use Gage R&R. It is very simple and usefull since you" only have micrometers and rules.".
Hope it helps !
"above and beyond" may be your understanding, however, just like many requirements in ISO 9001 - even simple document control, WHAT you do to calibrate equipment to ensure it is "suitable" MAY require an organization to perform these steps. Just be cause ISO/TS 16949 called it out specifically (or by ref. to the MSA manuals etc) may only be because it's not clear enough for those who would "prefer" not to do it.
The control of measuring equipment is often a misunderstood requirement where some most people know a little bit and are too quick to dismiss a number of activities as "not required". Just like many other ISO 9001 requirements don't tell you WHAT has to be done...
AndyN;572604 Just like many other ISO 9001 requirements don't tell you WHAT has to be done...[/QUOTE said:
And thus the comment of these activities are above and beyond the requirements of ISO 9001:2008.
Again, as I said before, they still may be desirable, just not a requirement.
In all seriousness, could you explain HOW (without reference to any of these established techniques) an organization is going to demonstrate that the equipment they use to measure something - is "suitable"? I'm all ears...:popcorn:
By training the users on how to properly use the equipment and to select the right measureing tool.
Much of selecting the right tool is common sense. You don't use a tape measure to measure something that needs to be within 5 thousands of an inch. You select a tool that has suffecient resolution.
Put the popcorn away. You don't need it here.
T Vendor (Supplier) Approval Procedure - Scope of the procedure ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
H IEC 62368-1 Scope of Testing CE Marking (Conformité Européene) / CB Scheme 3
A Laboratory scope on the IATF IATF 16949 - Automotive Quality Systems Standard 5
I Restricting scope to one part number for one customer IATF 16949 - Automotive Quality Systems Standard 6
A ISO 9001- any advise please with transitioning from a paper based system to eQMS and multi site scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S F-108 Request for Expansion of Scope of Accreditation ISO 17025 related Discussions 2
S Scope of management review ISO 13485:2016 - Medical Device Quality Management Systems 5
D Do all products fall under scope of AS9100? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M PROBLEM IN SCOPE OF ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
N ISO 13485 Scope Reduction Other ISO and International Standards and European Regulations 2
M AS 9100 Scope for Repair Station with PMA approval AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
M Intended Use vs Actual Use and Scope of Risk Management EU Medical Device Regulations 8
Stefan Mundt Split Scope = Separate Audits ??? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
lisap Design falling outside ISO scope Misc. Quality Assurance and Business Systems Related Topics 11
A Scope of ISO 13485 certification as legal manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 4
R ISO 13485 certification scope ISO 13485:2016 - Medical Device Quality Management Systems 5
Q Kap.4.3 Determining the scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B ERP software validation - risk assessment vs validation scope ISO 13485:2016 - Medical Device Quality Management Systems 11
G Why are 17025 scope angle measurements limited to 5 to 85 degrees? ISO 17025 related Discussions 2
3 Scope question ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
G Internal Laboratory Scope IATF 16949 - Automotive Quality Systems Standard 10
T How to handle out-of-scope activities. ISO 13485:2016 - Medical Device Quality Management Systems 7
Enghabashy Supply chain main policies ,scope, risk assessments & relavant KPI Supply Chain Security Management Systems 2
M Does the scope of ISO 9001:2015 applies to tenders, pricing and sales department of a medical devices distributor? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
S Can outsourced processes be excluded from QMS scope? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
D IATF 16949 M&TE Laboratory scope IATF 16949 - Automotive Quality Systems Standard 7
J Informal vs formal scope creep... managing non-medical devices through system processes ISO 13485:2016 - Medical Device Quality Management Systems 2
R "Medical devices" required in scope ISO 13485:2016 - Medical Device Quality Management Systems 2
R CB/Auditor Requiring a change in scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 9
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 4
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 19
H Implementing ISO22301 on a Limited Scope Business Continuity & Resiliency Planning (BCRP) 2
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
Q The scope of ISO 21534 Other Medical Device Related Standards 0
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
K Medical Device Repairs and ISO 13485 Scope ISO 13485:2016 - Medical Device Quality Management Systems 7
K Software Updates in the Field and ISO scope ISO 13485:2016 - Medical Device Quality Management Systems 2
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
Crimpshrine13 Laboratory Scope - Calibration vs. Test Methods - IATF 16949 IATF 16949 - Automotive Quality Systems Standard 3
silentmonkey Are risks in supply chain and development activities within scope of MDD? EU Medical Device Regulations 4
D Limited Scope for second site Question? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
J Scope of ISO 9001 clause 10.2 in the product life cycle ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
hockeyhead Existing process never included in AS9100 scope of certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 10
G APQP Scope and scale tool APQP and PPAP 2
R NRTL - Scope Question - Off-the-Shelf Plug In IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Scope of Combined ISO 9001 and IATF 16949 QMS - Non-automotive customers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5