Hello everyone,

we are a very small company (4 people), and started as a distributor for medical devices.

Meanwhile, we expanded our business and estabilished a contract with an asian manufacturer, providing 3 medical devices under our name. Also I need to say, that we do not design or develop.

Now, we intent to implement the ISO 13485 and want the scope to be similiar to this: "production and distribution of...". We need the word production included in our scope.

According to the ISO 13485, is there any problem with this? Technically our company is the manufacturer.

Best regards,

Paulo
 
Elsmar Forum Sponsor

Sidney Vianna

Post Responsibly
Staff member
Admin
#2
Re: Doubt scope

We need the word production included in our scope.
Can you please elaborate on why you NEED to have the word production in the scope of certification? Is that because of a regulatory requirements such as the Canadian CMDCAS program? Does this post ring a bell concerning your situation?

Technically, based on what you describe, you have outsourced production. How do you control that the production processes comply with the applicable requirements of ISO 13485?

From Guidance Document GD207: Guidance on the Content of ISO 13485 Quality Management System Certificates Issued by Health Canada Recognized Registrars
The MDR do not require importers or distributors of medical devices to have a certified QMS. However, any importer or distributor of a Class II, III or IV devices that choses to become the Manufacturer by labelling and selling the device with their own name, trade-mark, design, trade name or other name or mark owned or controlled by them (see definition of a Manufacturer) shall comply with the QMS sections of the MDR and other applicable sections.
 
Last edited:

Ronen E

Problem Solver
Staff member
Moderator
#3
<snip> Now, we intent to implement the ISO 13485 and want the scope to be similar to this: "production and distribution of...". We need the word production included in our scope.

According to the ISO 13485, is there any problem with this? Technically our company is the manufacturer. <snip>
Hello and welcome to the Cove :bigwave:

IMO, if you so desire, you can broaden the scope to include production. In regulatory terms, you ARE these devices' manufacturer (assuming you appear as such on the lebels). Besides, if you don't make finished medical devices, ISO 13485 does not make sense for you. It doesn't matter that you don't physically make them; the contract manufacturer is simply your extension, from a regulatory point of view. Note though that if you include production, all relevant requirements become applicable. This means that you won't get any exemptions for having actual production outsourced. On the contrary -- you should expect heightened scrutiny of behalf of the certifying org. They will look for effective controls over production, and since this production is thousands of miles away (and in a different culture), controls will have to be super-effective. Look at it as if your company will now have a remote manufacturing site.

Cheers,
Ronen.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#4
On the contrary -- you should expect heightened scrutiny of behalf of the certifying org. They will look for effective controls over production, and since this production is thousands of miles away (and in a different culture), controls will have to be super-effective. Look at it as if your company will now have a remote manufacturing site.
Not sure about this. If you are referring to a certifying organization as a certification body, most of them would consider an ISO 13485 certificate from an accredited CB for the contract manufacturer as "enough and adequate control", despite the fact that we do know for a fact that many certificates can not be trusted.

I still believe that there is a regulatory angle on the need for the word production to be in the scope of certification. While CB's might not make the distinction between (in-house) production and outsourced production, I do believe that regulatory bodies might differ in their approach.
 

Ronen E

Problem Solver
Staff member
Moderator
#5
Not sure about this. If you are referring to a certifying organization as a certification body, most of them would consider an ISO 13485 certificate from an accredited CB for the contract manufacturer as "enough and adequate control", despite the fact that we do know for a fact that many certificates can not be trusted.
This is interesting. I always thought that ISO 13485 compliance on the supplier's part is only one ingredient, and the customer (the legal manufacturer in this case) is nevertheless obliged to exercise adequate controls to make sure that its supplier's outputs are satisfactory. In my terminology "Effective controls" doesn't necessarily mean elaborated; if the supplier ISO 13485 outputs are satisfactory and integrate well with the customer's QMS, then a few simple clauses may be all that is required. Further, when I wrote "heightened scrutiny" I didn't mean going into the supplier's QMS in detail; rather, I meant that the interface between the supplier's and customer's QMSs is likely to be closely looked at. This is what I would have done. It seems that this is a very common weak link when collaborating over great distances and cultural gaps.

However, this is just my own interpretation.

Cheers,
Ronen.
 

somashekar

Staff member
Super Moderator
#6
<snip> Now, we intent to implement the ISO 13485 and want the scope to be similar to this: "production and distribution of...". We need the word production included in our scope.

According to the ISO 13485, is there any problem with this? Technically our company is the manufacturer. <snip>
You can have your scope of "Manufacture and sales of ... " in your scope when you have manufacturing in-house or outsourced.
Ask your CB how.
Keep your QMS ready with the controls that you extend and be able to demonstrate these being effective.
Quality management agreements that you establish between you and the manufacturing site can come handy as a framework for effective controls.
 
L

Laura Halper

#7
I agree with Somashekar's comments. I think the term "manufacture" has a broader connotation than "production". For example, the manufacturer may out-source production.
 

Ronen E

Problem Solver
Staff member
Moderator
#8
I agree with Somashekar's comments. I think the term "manufacture" has a broader connotation than "production". For example, the manufacturer may out-source production.
It is my impression that "Production" is less used and less well-defined than "Manufacturing" in the international medical devices regulatory context. Hence the confusion.
 
L

Laura Halper

#9
Yes, I think you are right. The regulations use the term "manufacturing", whereas ISO 13485 uses the term "production".

To further confuse things, the regulations use the term "manufacturer" in two ways -- the legal Manufacturer and also the person who physically makes the product. They may or may not be the same entity.
 

Ronen E

Problem Solver
Staff member
Moderator
#10
Yes, I think you are right. The regulations use the term "manufacturing", whereas ISO 13485 uses the term "production".

To further confuse things, the regulations use the term "manufacturer" in two ways -- the legal Manufacturer and also the person who physically makes the product. They may or may not be the same entity.
ISO 13485 uses both, sometimes interchangeably(?).

For instance:

NOTE Design transfer activities during the design and development process ensure that design and development outputs are verified as suitable for manufacturing before becoming final production specifications.
So it doesn't help clarifying the situation.

Seems that in the ISO 13485 context "Production" is more limited to the actual fabrication of devices, while "Manufacturing" sometimes applies in the broader context of the entire process that leads to placing the device on the market (and so, as you've pointed out, includes "legal manufacturing" as well).

I wonder if the next revision will resolve this confusion.

Cheers,
Ronen.
 
Thread starter Similar threads Forum Replies Date
D ISO 13485 scope (implantable) - Polymers for dental application EU Medical Device Regulations 8
M ISO 13485:2016 Certification Scope ISO 13485:2016 - Medical Device Quality Management Systems 2
M Scope for ISO 13485 Certification of a Translation Service Provider ISO 13485:2016 - Medical Device Quality Management Systems 17
M How Specific in an ISO 13485:2016 Scope for a Contract Manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 9
A Scope of ISO 13485 certificate ISO 13485:2016 - Medical Device Quality Management Systems 1
K Restricted Scope of ISO 13485 Certification ISO 13485:2016 - Medical Device Quality Management Systems 7
R Scope of ISO 13485 certification ISO 13485:2016 - Medical Device Quality Management Systems 6
P Dropping ISO 9001 limits the scope of the ISO 13485 audit? ISO 13485:2016 - Medical Device Quality Management Systems 6
S ISO 13485 scope of certification - We are a virtual manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 1
K MDSAP scope vs ISO 13485 Scope Canada Medical Device Regulations 4
R Scope of Certification for a Design Organization with Outsourced Mfg. - ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 10
M OEM-OBL ISO 13485 Certification - Devices mentioned in the product scope ISO 13485:2016 - Medical Device Quality Management Systems 4
chris1price What are the Scope Numbers on ISO 13485 Certificates ISO 13485:2016 - Medical Device Quality Management Systems 2
R ISO 13485 Certificate Scope modified by Notified Body EU Medical Device Regulations 9
S What is the scope of EN ISO 13485 and ISO 13485 ? ISO 13485:2016 - Medical Device Quality Management Systems 3
A Scope of ISO 13485 Registration and Component Makers ISO 13485:2016 - Medical Device Quality Management Systems 7
V ISO 13485 certification scope in Spanish ISO 13485:2016 - Medical Device Quality Management Systems 5
N Scope of ISO 13485 Certificate for Multiple Devices EU Medical Device Regulations 3
B Class IIb Medical Device using Annex II route - Need to Update ISO 13485 Scope? ISO 13485:2016 - Medical Device Quality Management Systems 2
L Non-medical device is in the ISO 13485 audit scope? ISO 13485:2016 - Medical Device Quality Management Systems 11
Ajit Basrur Scope for ISO 13485 vs. ISO 15378 - Molding Parts for Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
Q The scope of ISO 21534 Other Medical Device Related Standards 0
K Medical Device Repairs and ISO Scope ISO 13485:2016 - Medical Device Quality Management Systems 3
K Software Updates in the Field and ISO scope ISO 13485:2016 - Medical Device Quality Management Systems 2
J Scope of ISO 9001 clause 10.2 in the product life cycle ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
D ISO 9001:2015 4.3 Determining the Scope of the QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
M Scope of Combined ISO 9001 and IATF 16949 QMS - Non-automotive customers ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M ISO 27001 ISMS scope for companies with subsidiaries IEC 27001 - Information Security Management Systems (ISMS) 0
L Dying a slow SCOPE death - NEW ISO APG Paper on Scope and Applicability May 2020. AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 3
P ISO 80369-7 standard - Interpreting which Parts should be in scope Other Medical Device Related Standards 7
I Sales Documents in scope for ISO-9001:2015? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
qualprod ISO 9001:2015 8.5.1 f - Scope with customer special process requirements ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
S ISO 9001:2015 - Scope of Registration ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 35
N ISO 9001 certification with limited scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
Q ISO 9001, section 4.3 Determining the scope of our QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
Q ISO 9001:2015 - Extent of Policy and Scope with Context and Interested Parties ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
K ISO 9001:2015 clause 9.2.2 a. - Define the audit criteria and scope Internal Auditing 2
B ISO 14001:2015 4.3 - How to make the scope of the EMS available to interested parties ISO 14001:2015 Specific Discussions 4
F ISO 9001:2015 4.3 - Determining the scope of the quality management system ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
J ISO 17025, Laboratory Scope and Certifications Requests ISO 17025 related Discussions 4
G How to define the scope of QMS as per ISO 9001:2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 21
Anerol C IATF or ISO TS 16949 rules about Scope of QMS IATF 16949 - Automotive Quality Systems Standard 7
M Taking on a job outside of our ISO Scope ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
I ISO 9001 Scope - Exclusion of Design in an engineered solutions company ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 28
T Does your Quality Dept Control Procedures Outside the Scope of ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
S ISO 9001:2008 Certification Scope does not mention "manufacturing" ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
F ISO/TS 16949 internal audit scope and annual plan Internal Auditing 2
F What is the scope for "Total Numbers of Employees on site" per ISO/TS 16949 ? IATF 16949 - Automotive Quality Systems Standard 4
R Changing ISO 9001 Certification Scope Quality Manager and Management Related Issues 4

Similar threads