Selecting components to meet UL 60601-1
blackrooster
Hello All:
I am new to UL60601-1 and could use a little direction on component selection.
We are designing a machine that we want to comply with UL60601-1. In my opinion it is a bit of a stretch to consider it a medical device but some of our customers have indicated that they would like FDA approval. The machine does not have any parts that would be applied to a patient so I think I am just concerned with operator safety.
My question is about selecting components to meet UL60601-1. I see that some things like power supplies are easy to find with a UL60601-1 listing. However there are a lot of component types where I don?t see any options. I don?t know whether I am not looking at the right suppliers or if I am looking for something that doesn?t exist.
As one example, the control of this machine would be well suited for a Programmable Logic Controller (PLC). Unfortunately I don?t see anything similar to this that is intended for a medical electrical device. Are there any off the shelf type controllers that are appropriate for UL60601-1? I would rather not have to design a custom PCB for this machine.
I would appreciate any input that you guys can share with me.
Thanks in advance.
Jaydub
Here is my take:
Only components that form a means of protection or the failure of which would create a hazard, would need to be considered. For example, this most likely includes everything on the primary side of an isolation transformer. These components should meet any relevant safety standards and be used within their ratings. For example most power switches have been evaluated by UL and a myriad of other agencies. They don't need to specifically meet IEC 60601-1, but rather the IEC or ISO standard for power switches. A switching power supply probably does need to meet IEC 60601-1 requirements for your product to pass leakage current and dielectric strength tests.
A PLC probably is not a safety critical component and you could use any one you wish.
Peter Selvey
According to UL 60601-1:2003, there is a special clause (3.10DV.1.1) which requires specific components to meet UL standards, and this is limited to wiring, tubing, optocouplers, PWBs, lithium batteries and CRTs.
For other parts, it is the judgment of each engineer whether a part is:
(a) included in the critical component list, and
(b) included in the critical component list with required UL recognition
There are many cases where a part is considered safety critical, but no UL is expected. The part will appear on the component list just so the factory inspector can check if it has changed.
If the engineer insists on UL recognition, usually there is usually some reason. For example, the part itself may be so critical that factory inspection at the part manufacturer is also reasonable. In such a case, choosing a UL recognized part makes things easier (but other routes are also possible).
One problem is the tendency to list UL recognition even if not required. For example, the engineer will often list the maker and material of the rating label (which is reasonable) but then go on to list the UL recognition. However, the recognition is often just for UL 94 (flammability) which has nothing to do with complying with a rating label or UL 60601-1. It then looks as if the UL recognition is required for the rating label, but actually it is not.
blackrooster
Thanks to both of you for the reply.
My understanding is that the primary criteria for 60601 compliance are dielectric, creepage, clearance, and leakage current. I see how a typical switching power supply is unlikely to meet all of these and so a 60601 power supply is needed.
A few more questions if someone doesn?t mind?
1) Are there other kinds of components besides a switching power supply that you generally would want to look for a UL60601 rating?
2) Is it safe to say that I don?t need to worry about anything low voltage (24V)?
3) As an example, it sounds like a contactor for starting a motor would be a MOP because it carries the mains voltage. I don?t think there are UL60601 contactors so I should just look for one that meets the IEC standard for the device and then verify dielectric, creepage and clearance dimensions. Does this sound right or am I missing something?
Sorry if these are really basic questions. I am obviously very new to 60601 and I?m just trying to get a feel for what is expected.
Peter Selvey
Actually UL 60601-1 is not applicable to power supplies. It is kind of an industry quirk which allows power supplies to be certified, strictly against the rules but allowed because it makes the end product evaluation simpler. But it is the exception rather than the rule. Most components have special standards that are written for that component, and the expectation is to use the applicable standard for the component, not UL 60601-1.
If the part is required to be certified, there should be some relevance to the scope of certification. For UL, their certification is based on UL 60601-1 but there is also a disclaimer which limits this to "shock, fire and mechanical hazards" only. So in general the focus is on mains/patient isolation; parts protecting against fire and mechanical hazards. Most of these parts are in the mains circuit.
blackrooster
That makes sense. I got confused because someone cautioned me against using UL508 components when trying to comply with UL60601 and I assumed that I would be able to find UL60601 equivalents for everything. I didn't understand why power supplies were the only devices I could find.
So my general approach will be to select components that comply with their applicable standards and then verify that the creepage, clearance, and dielectric ratings fall within the limits of 60601. Low voltage components are probably easy to meet the requirements and the danger is low. Mains voltage components have a much higher risk of shock or fire hazards so this will be the focus of any 60601 evaluation.
M Suggestions for Selecting RF Wireless Components Other US Medical Device Regulations 3
J Selecting factors for screening DoE design Statistical Analysis Tools, Techniques and SPC 8
M Informational How To Avoid Compliance & Timeline Risks When Selecting A Medical Device Supplier Medical Device and FDA Regulations and Standards News 0
O ISO 13485 for software company - Selecting suppliers ISO 13485:2016 - Medical Device Quality Management Systems 3
C Selecting potential internal auditors Internal Auditing 3
V Selecting System Precision Criteria - System Suitability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Q Advice on selecting an ISO 9001:2015 Lead Auditor Certification Course ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
B Selecting Parts for an MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
B Process Control - Selecting Parts - NDC & Study Variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Risk Management while selecting the Supplier Document Control Systems, Procedures, Forms and Templates 3
T Selecting which MDD 93/42/EEC Annex to follow - PACS software system CE Marking (Conformité Européene) / CB Scheme 2
AnaMariaVR2 5 keys to selecting the right CM (Contract Manufacturer) ISO 13485:2016 - Medical Device Quality Management Systems 1
M Pre-Selecting Materials for ISO 10993-1 Biocompatibility Compliance Other ISO and International Standards and European Regulations 5
P Selecting a Registrar and completing our ISO 9001 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
S Selecting Wires for Measuring P.D. of Thread Gages General Measurement Device and Calibration Topics 9
S Selecting weights for daily Calibration Check - Using just one weight? General Measurement Device and Calibration Topics 10
W Selecting a Registration Agent in China China Medical Device Regulations 17
C Selecting Consultants for getting drug approval in USFDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
G Gage R&R Procedures: Selecting parts (sample) to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
W Selecting Non-Conformities to address Nonconformance and Corrective Action 16
C Selecting the Right Problem (book chapter 1) The Reading Room 8
G Gage R&R: Selecting Number of Trials and Appraisers and Parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
B Selecting a Sample and Determining the Reference Value for a Bias Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Le Chiffre Selecting a predicate device for FDA 510(k) approval Other US Medical Device Regulations 14
W Question about sample size code letter and selecting sampling plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
D Selecting a Tolerance when multiple specifications exist - Different Beers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Optimally Selecting 30 Pieces of Product out of 34 - How do I? Statistical Analysis Tools, Techniques and SPC 9
J Selecting the best Gage - Buy according to production criteria IATF 16949 - Automotive Quality Systems Standard 0
A Is there any kind of Ranking available for selecting the Registrars? Registrars and Notified Bodies 31
L I need a reference for selecting a notified body. Can anyone make a recommendation? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 21
L Selecting ISO 9001:2000 Internal Auditors Internal Auditing 12
Q Criteria for Selecting a Consultant or Consulting Firm Consultants and Consulting 2
B Selecting Six Sigma Black Belts Six Sigma 3
S QOS for Tooling and Equipment - Having difficulty selecting a measurable QS-9000 - American Automotive Manufacturers Standard 2
S Reporting REACH - Electronic components distributor REACH and RoHS Conversations 5
R Components to a finished medical device, MDR requirements Other US Medical Device Regulations 1
U Procedure pack components EU Medical Device Regulations 3
T Sterilized components expiration date ISO 13485:2016 - Medical Device Quality Management Systems 1
T Sterilized components shelf life Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
C Warehousing two separate components to a finished device while meeting regulations Other US Medical Device Regulations 7
G Dealing with non conformity caused by Supplier Components detected in the production line IATF 16949 - Automotive Quality Systems Standard 14
C Shelf life electronic components 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Process FMEA Template with examples - Cold and Hot Forged components FMEA and Control Plans 4
F Biocompatibility evaluation for Hardware/Interface Components? Other Medical Device Related Standards 12
K 510K new manufacturer but same components Other US Medical Device Regulations 6
D Using non-conforming components even though the final assembly is conforming? Manufacturing and Related Processes 5
R DHR question: Traceability of components ISO 13485:2016 - Medical Device Quality Management Systems 2
R What is meant by Use of COMPONENTS WITH HIGH-INTEGRITY CHARACTERISTICS in ME EQUIPMENT (clause no 4.9) IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
J Medical Device Components US to EU commercialization EU Medical Device Regulations 7