Selecting Parts for an MSA (Measurement System Analysis)

B

Black arrow

#1
Hi,

We will soon check if our measurement system is adequate for spc and cap.studies so % Study Var and ndc are important here. We have already selected parts and from my point of view it is important to select part from the normal process so those were picked during the whole day. So far so good i guess.

BUT My friend works at a company that also includes customer complaints parts in their MSA, complaints they have had during the year.
Is that correct approach?
 
Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
Re: Select parts-MSA

If they are using %Tolerance, this will not affect their results. They should not do this if they are using %SV or ndc because it will bias the results and make them look better than reality.
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#3
The intent is to have parts that are representative of the full range of variation that the process typically produces. Customer complaint parts are certainly part of this range but if you include too many of them you will skew your study (biased toward the worst parts)


I'm also not sure that a single day is representative of the full range of variation that your process will experience.

I do know that if you use the AIAG default of 10 parts you will have a poor estimate of the process variation regardless of the above concerns. 10 parts is just too small to reliably estimate the process variation. I typically use 30 parts (measured only twice).
 
B

Black arrow

#4
Hi,

I selected 50 parts and measured them and calculated standard deviation. Then i selected 10 parts from those 50 so they will have same standard deviation (spread) as those 50 i selected first.

Hmm, Bev you said " process typically produces". How many customer complaints parts should he really select here? If he for example produce 500 parts and 2 of them are customer complaints, should he involved them? I could find it tricky here to define what is typically normal process.
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#5
IHmm, Bev you said " process typically produces". How many customer complaints parts should he really select here? If he for example produce 500 parts and 2 of them are customer complaints, should he involved them? I could find it tricky here to define what is typically normal process.
it really depends. if the customer complaint parts have values that are outside the range of values captured by your internal parts then you shoul dadd some customer compaint parts. if they within the range of the internal part values there is no additional value that I can see at this point to adding customer complaint part.
 
B

Black arrow

#6
it really depends. if the customer complaint parts have values that are outside the range of values captured by your internal parts then you shoul dadd some customer compaint parts.
Even though it is 1 customer complaint part of 1000 produced parts ?
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#7
OK it's not about whether or not the parts are customer complaint parts. nor it about achieving some %TV or %PV goal. what is critical is to understand the variation of your measurement system.

so the question is: does the Customer complaint part(s) have a value that is larger or smaller than your in house sample set? if so, include it. if not don't include it.

these things are never cookbook unless you just want to 'check the box' and day you did it.

if you are truly trying to understand, you must think about your measuremetn system and how the process works to create variation in the characteristic of interest. you then design the study to answer the question(s) your are intending to understand.
 
Last edited:
Thread starter Similar threads Forum Replies Date
B Process Control - Selecting Parts - NDC & Study Variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G Gage R&R Procedures: Selecting parts (sample) to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G Gage R&R: Selecting Number of Trials and Appraisers and Parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
J Selecting factors for screening DoE design Statistical Analysis Tools, Techniques and SPC 8
M Informational How To Avoid Compliance & Timeline Risks When Selecting A Medical Device Supplier Medical Device and FDA Regulations and Standards News 0
O ISO 13485 for software company - Selecting suppliers ISO 13485:2016 - Medical Device Quality Management Systems 3
C Selecting potential internal auditors Internal Auditing 3
V Selecting System Precision Criteria - System Suitability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Q Advice on selecting an ISO 9001:2015 Lead Auditor Certification Course ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
B Selecting components to meet UL 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Risk Management while selecting the Supplier Document Control Systems, Procedures, Forms and Templates 3
T Selecting which MDD 93/42/EEC Annex to follow - PACS software system CE Marking (Conformité Européene) / CB Scheme 2
AnaMariaVR2 5 keys to selecting the right CM (Contract Manufacturer) ISO 13485:2016 - Medical Device Quality Management Systems 1
M Pre-Selecting Materials for ISO 10993-1 Biocompatibility Compliance Other ISO and International Standards and European Regulations 5
P Selecting a Registrar and completing our ISO 9001 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
S Selecting Wires for Measuring P.D. of Thread Gages General Measurement Device and Calibration Topics 9
M Suggestions for Selecting RF Wireless Components Other US Medical Device Regulations 3
S Selecting weights for daily Calibration Check - Using just one weight? General Measurement Device and Calibration Topics 10
W Selecting a Registration Agent in China China Medical Device Regulations 17
C Selecting Consultants for getting drug approval in USFDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
W Selecting Non-Conformities to address Nonconformance and Corrective Action 16
C Selecting the Right Problem (book chapter 1) The Reading Room 8
B Selecting a Sample and Determining the Reference Value for a Bias Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Le Chiffre Selecting a predicate device for FDA 510(k) approval Other US Medical Device Regulations 14
W Question about sample size code letter and selecting sampling plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
D Selecting a Tolerance when multiple specifications exist - Different Beers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Optimally Selecting 30 Pieces of Product out of 34 - How do I? Statistical Analysis Tools, Techniques and SPC 9
J Selecting the best Gage - Buy according to production criteria IATF 16949 - Automotive Quality Systems Standard 0
A Is there any kind of Ranking available for selecting the Registrars? Registrars and Notified Bodies 31
L I need a reference for selecting a notified body. Can anyone make a recommendation? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 21
L Selecting ISO 9001:2000 Internal Auditors Internal Auditing 12
Q Criteria for Selecting a Consultant or Consulting Firm Consultants and Consulting 2
B Selecting Six Sigma Black Belts Six Sigma 3
S QOS for Tooling and Equipment - Having difficulty selecting a measurable QS-9000 - American Automotive Manufacturers Standard 2
I When is necessary to have RoHS declaration on non-electrical parts? REACH and RoHS Conversations 1
Z Rapid audit template for plastic parts manufacturing process Manufacturing and Related Processes 12
C Scrapping on production floor - Setup parts 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
G Bad Parts cause Customer line stop IATF 16949 - Automotive Quality Systems Standard 12
G Calibration of "Master Parts" Used as Gauges Calibration Frequency (Interval) 5
P FDA Approved Product Contact Parts ISO 13485:2016 - Medical Device Quality Management Systems 6
M Identifying Applied Parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Nonconforming parts accepted by operator with intention Human Factors and Ergonomics in Engineering 31
J Mechanical inspection techniques of close tolerance parts Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
J Sister-company providing parts is only ISO 9001 registered IATF 16949 - Automotive Quality Systems Standard 7
A AS9100D 8.1.4 Prevention of Counterfeit Parts AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
C Difficulties getting RoHS/REACH for cheap parts RoHS, REACH, ELV, IMDS and Restricted Substances 1
G Same parts but new customer - What will the auditor ask me? IATF 16949 - Automotive Quality Systems Standard 2
G Need to change KPI we called NC parts (maximum 3%.) to FTQ (first time quality) IATF 16949 - Automotive Quality Systems Standard 4
Q Do you classify applied parts? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1

Similar threads

Top Bottom