Before analysing the samples on the instrument, it is recommended to evaluate the repeatability/sensitivity specific to the method-parameters.
this requirement could be more relevant&specific for pharmaceutical industry as it handles little more complex samples. (this is in addition to the routine calibration - performance verification steps )
( quick ref... WHO...EP...USP...article...summary)
Question :-
in order to ensure repeatability following methods are currently in use.
1. calculate RSD of consecutive-multiple-measurements (typically 3) of a known standard. (typically <2%) (Yes, I agree that the no. of measurements are proportional to the sensitivity)
2. ratio of two consecutive measurements. (typically the ratio of two measurements to be 0.98 - 1.02 )
which one of above is superior or preferred.
Here's my quick assessment.,
* for RSD (<2%) , the max range possible is +/- 2%
* however for ratio (+/- 0.02), the max range possible is +/- 1%)
is there a way to assess these scenarios of controls.
Question :-