Selecting weights for daily Calibration Check - Using just one weight?

BradM

Staff member
Admin
#11
Hello there! :bigwave:

First, are these balances used in an FDA setting, where they may require GLP compliance? If so, then your answer may take another approach where you need to demonstrate balance precision.

Having said that, the balance needs to be verified (for routine use) at the range it is used at. so if all you ever use the balance for is measuring product between 100-500 grams, then that would be your target range.

If you use your balance for one weight, and only one weight, then a single check is sufficient. However, rare is the case that a balance is only used for one measure. People from other labs, other areas could come in and weight product also.

Otherwise, you need to verify the balance across the range of use.

Now... tolerance is another matter. Like Howste correctly identified, Handbook 44 is a good document. However, most people with precision analytical balances tend to find that document.... not very helpful. :tg:

It may be best to establish your own tolerances based on process requirements. Also, be sure and document the performance checks and any noticed drift. That way you can establish how stable the balance is.

Finally, this should not take the place of having the balance calibrated by a qualified vendor. They can come in, remove error, perform corner loads, etc.
 
Elsmar Forum Sponsor
Thread starter Similar threads Forum Replies Date
J Selecting factors for screening DoE design Statistical Analysis Tools, Techniques and SPC 8
M Informational How To Avoid Compliance & Timeline Risks When Selecting A Medical Device Supplier Medical Device and FDA Regulations and Standards News 0
O ISO 13485 for software company - Selecting suppliers ISO 13485:2016 - Medical Device Quality Management Systems 3
C Selecting potential internal auditors Internal Auditing 3
V Selecting System Precision Criteria - System Suitability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Q Advice on selecting an ISO 9001:2015 Lead Auditor Certification Course ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 22
B Selecting Parts for an MSA (Measurement System Analysis) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
S Selecting materials for implants to comply with ISO 10993 biocompatibility Other Medical Device Related Standards 4
B Process Control - Selecting Parts - NDC & Study Variation Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Selecting components to meet UL 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B Risk Management while selecting the Supplier Document Control Systems, Procedures, Forms and Templates 3
T Selecting which MDD 93/42/EEC Annex to follow - PACS software system CE Marking (Conformité Européene) / CB Scheme 2
AnaMariaVR2 5 keys to selecting the right CM (Contract Manufacturer) ISO 13485:2016 - Medical Device Quality Management Systems 1
M Pre-Selecting Materials for ISO 10993-1 Biocompatibility Compliance Other ISO and International Standards and European Regulations 5
P Selecting a Registrar and completing our ISO 9001 Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 15
S Selecting Wires for Measuring P.D. of Thread Gages General Measurement Device and Calibration Topics 9
M Suggestions for Selecting RF Wireless Components Other US Medical Device Regulations 3
W Selecting a Registration Agent in China China Medical Device Regulations 17
C Selecting Consultants for getting drug approval in USFDA Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
G Gage R&R Procedures: Selecting parts (sample) to measure Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
W Selecting Non-Conformities to address Nonconformance and Corrective Action 16
C Selecting the Right Problem (book chapter 1) The Reading Room 8
G Gage R&R: Selecting Number of Trials and Appraisers and Parts Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
B Selecting a Sample and Determining the Reference Value for a Bias Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
Le Chiffre Selecting a predicate device for FDA 510(k) approval Other US Medical Device Regulations 14
W Question about sample size code letter and selecting sampling plan Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
D Selecting a Tolerance when multiple specifications exist - Different Beers Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
B Optimally Selecting 30 Pieces of Product out of 34 - How do I? Statistical Analysis Tools, Techniques and SPC 9
J Selecting the best Gage - Buy according to production criteria IATF 16949 - Automotive Quality Systems Standard 0
A Is there any kind of Ranking available for selecting the Registrars? Registrars and Notified Bodies 31
L I need a reference for selecting a notified body. Can anyone make a recommendation? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 21
L Selecting ISO 9001:2000 Internal Auditors Internal Auditing 12
Q Criteria for Selecting a Consultant or Consulting Firm Consultants and Consulting 2
B Selecting Six Sigma Black Belts Six Sigma 3
S QOS for Tooling and Equipment - Having difficulty selecting a measurable QS-9000 - American Automotive Manufacturers Standard 2
P Qualification of weights used for curing epoxy Manufacturing and Related Processes 3
D Lesser expensive to calibrate various test weights General Measurement Device and Calibration Topics 5
P Uncertainty of Measurements for Adhesive Weights for Peel Testing Capability, Accuracy and Stability - Processes, Machines, etc. 2
M Establishing Ranges for Weights for Calibration Verification of a Micro-Balance Calibration Frequency (Interval) 1
S Count (Bolts and Nuts) Scale - Class of Calibration Weights Calibration and Metrology Software and Hardware 3
D How to determine the Calibration/Test Weights with their classes General Measurement Device and Calibration Topics 5
G (Free) Mass Calibration Software - Calibration of Weights to UKAS Standards? Calibration and Metrology Software and Hardware 3
A How to calculate weights for customer satisfaction score? Misc. Quality Assurance and Business Systems Related Topics 4
C What is the least acceptable TAR value? Mettler UMX2 micro balance - Class E2 weights General Measurement Device and Calibration Topics 4
B Are Brass weights obsolete? Should they be replaced by Steel weights? General Measurement Device and Calibration Topics 7
qualprod Ideas to stop bad practices in daily work ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
optomist1 Rates Daily or Hourly Process Improvement Training Consultants and Consulting 2
C Quality, efficiency and effectiveness of daily work routines Misc. Quality Assurance and Business Systems Related Topics 4
I Can subgroup size change daily? (X-bar & R) Capability, Accuracy and Stability - Processes, Machines, etc. 5
M Daily Temperature and Humidity Graph template Document Control Systems, Procedures, Forms and Templates 2

Similar threads

Top Bottom