I'm currently having my first read at IEC 60601-1 and I stumbled upon clause 8.5.2.1 which mentions that patient connections of any F-type applied part needs to be separated from all other parts. However, given my current device's architecture (which I will detail below), this doesn't make sense to me. I'd like to hear your thoughts on this!
Basically, my device is a wearable device which is body-worn for several days and has 2 snaps/electrodes (the applied part) which are used to acquire ECG (therefore a BF applied part). The device is powered by a LiPo battery. The device can only be charged via USB (either through a medical grade charger or by being connected to the computer). Given that the medical grade charger has all the protection required by the IEC 60601, I'm only considering hazard situations to arise from the PC connection (e.g., a peak current from the mains, which flows through the computer, goes into my device via the USB port and reaches the patient through the snaps). To protect against this, I have implemented an USB isolator (the ADuM family isolators), which has the recommended creepage distances and air clearances for 2 MOPP.
Hence, given that I'm protecting the only possible source of hazard, it doesn't make sense to me that I need to separate the ECG circuitry from the remaining circuitry. What are your thoughts on this? To be in compliance with this clause, can I consider the entire wearable device as a Type BF applied part?
Thanks so much!
Diogo
I'm currently having my first read at IEC 60601-1 and I stumbled upon clause 8.5.2.1 which mentions that patient connections of any F-type applied part needs to be separated from all other parts. However, given my current device's architecture (which I will detail below), this doesn't make sense to me. I'd like to hear your thoughts on this!
Basically, my device is a wearable device which is body-worn for several days and has 2 snaps/electrodes (the applied part) which are used to acquire ECG (therefore a BF applied part). The device is powered by a LiPo battery. The device can only be charged via USB (either through a medical grade charger or by being connected to the computer). Given that the medical grade charger has all the protection required by the IEC 60601, I'm only considering hazard situations to arise from the PC connection (e.g., a peak current from the mains, which flows through the computer, goes into my device via the USB port and reaches the patient through the snaps). To protect against this, I have implemented an USB isolator (the ADuM family isolators), which has the recommended creepage distances and air clearances for 2 MOPP.
Hence, given that I'm protecting the only possible source of hazard, it doesn't make sense to me that I need to separate the ECG circuitry from the remaining circuitry. What are your thoughts on this? To be in compliance with this clause, can I consider the entire wearable device as a Type BF applied part?
Thanks so much!
Diogo