Separation requirement for Class II equipment with functional earth

Roland chung

Trusted Information Resource
#1
Hi folks,

I have a question about the Class II equipment with functional earth. It is easy to think that functional earth does not provide MOP at location where large fault current (e.g. short circuit) may occur. The standard therefore requires two MOPs between internal screens, including internal wiring connected to them and ACCESSIBLE PARTS.

But I am a little bit confused on the wording for insulation the standard used which mentioned above. I can at least think of 3 scenes:

1) The internal screen functionally earthed is not separated from MAINS, then two MOPs required from internal screen to accessible part. (In line with standard requirement)

2) The internal screen functionally earthed is separated from MAINS by one MOP, then an additional one MOP from internal screen to accessible part is sufficient.

3) The internal screen functionally earthed is separated from MAINS by two MOPs, then no insulation required between internal screen and accessible part.

Any other opinions?

Thanks,
Roland
 
Elsmar Forum Sponsor
R

raymond02302

#2
From the meaning of wording of standards, opinion 3 should be right. It is only such condition for equipment also with acceessible functional terminial. For other equipment without accessible functional terminal, I can't foresee any hazard that will arise for equipment structured as per opinin 1 and 2.

BTW, this structure is not allowed for medical equipment used in home healthcare environment.
 

Peter Selvey

Staff member
Super Moderator
#4
I haven't checked the standard on this one, but I think the term "functional earth" basically assumes it's part which is accessible, i.e. (e.g.) an externally accessible terminal for the user may optionally use to connect a ground wire to for noise reduction.

I know for Class II equipment, it may get confusing, because the design may use the 3rd pin of the power supply cord as the FE terminal, which means the user may not be able to directly touch it. But the original concept was for a terminal separate to the power cord, like you see on many sensitive instruments.

Does this make sense?
 

Roland chung

Trusted Information Resource
#5
You may be right regarding the original concept. We could find some clues from Figure 2 to Figure 4 in IEC 60601-1:2005. All the functional earth described in the Figures are exposed.

Let's look back to the question, the standard says the insulation between internal screens, including internal wiring connected to them and ACCESSIBLE PARTS, shall provide two MEANS OF PROTECTION. According to the requirement, the earth conductor of power supply cord shall be used only as the functional earth connection to a FUNCTIONAL EARTH TERMINAL for these screens. That means the internal screens are functionally earthed.

There is obviousely no insulation between internal screens and accessible parts (functional earth), since they are connected directly. I am totally confused.
 
D

Dimitri2

#6
I see it as follows: your functional earth and all other ASSESSIBLE parts must be 2 MOOPs away from the MAINS as well as from any other power entry point.

Would you please explain what your "internal screen" really ist?
 
D

Dimitri2

#8
Figure 2 does not show any insulation of FE from the MAINS. It just explains the terms.
At Fig. 3 you can see that between MAINS and FE there are at least 2 MOPPs whereas a screen between windings is connected to PROTECTIVE EARTH CONDUCTOR.
 
Thread starter Similar threads Forum Replies Date
M Do I need separation in my circuit with a medical charger? IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
D Separation of F-type applied part and remaining parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 7
B F-type applied part - Separation from ALL(?) other parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
Ahang Separation between secondary circuit and accessible parts IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
O Difference in 60601-1-2 3rd and 4th ed separation RF portable equipment CE Marking (Conformité Européene) / CB Scheme 1
Q Rule 5.2h - Site Separation - Audit Manday Calculation (IATF 16949) IATF 16949 - Automotive Quality Systems Standard 17
S Separation Devices as required in 16.5 of IEC 60601-1:2005 IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
R Separation of Network Port from other Circuits including Patient Circuit IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Employee Transfer, Separation and Clearance Procedure and Forms ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
B "Gated Conditions" - Separation of automotive and non-automotive product IATF 16949 - Automotive Quality Systems Standard 13
C Separation of Facilities - Part 145 Repair Station AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 37
B Separation of Duties and Responsibilities ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 14
D What is the acceptance criteria/requirement for Stability Study? General Measurement Device and Calibration Topics 3
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
L Water requirement for Non-sterile topical OTCs Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
C Requirement to link Quality Manual to ISO 9001 clause numbers? ISO 13485:2016 - Medical Device Quality Management Systems 13
B FDA requirement for CAPA Signoff ISO 13485:2016 - Medical Device Quality Management Systems 6
W Pfmea function requirement and failure mode FMEA and Control Plans 6
Y 510k requirement for bringing in gloves Medical Device and FDA Regulations and Standards News 0
S PSUR - MDR First Submission Requirement? EU Medical Device Regulations 2
J When PPAPs are not a contractual requirement APQP and PPAP 9
I Overwhelmed with attribute MSA requirement for visual inspection IATF 16949 - Automotive Quality Systems Standard 8
Brizilla Employee Data Privacy Policy - ISO 9001:2015 requirement(s)? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
P Clinical Trial Requirement - No equivalent devices in the market CE Marking (Conformité Européene) / CB Scheme 3
J Requirement for Retention of Records of Withdrawn Documents of External Origin Document Control Systems, Procedures, Forms and Templates 3
T ISO 17025: Lockout/Tagout Requirement ISO 17025 related Discussions 1
A The requirement of the PMS plan--suitable indicators and threshold values EU Medical Device Regulations 0
S Battery powered device - electrical protection requirement IEC 60601 - Medical Electrical Equipment Safety Standards Series 18
K Regulatory requirement of SaMD with machine learning component IEC 62304 - Medical Device Software Life Cycle Processes 3
A ASL requirement when the supplier is certified for ISO 13485 ISO 13485:2016 - Medical Device Quality Management Systems 6
P Regulatory or registration requirement for disinfectant in ASEAN Registrars and Notified Bodies 1
P IATF 16949 requirement - error-proofing in control plan IATF 16949 - Automotive Quality Systems Standard 3
C ISO/ IEC 17021 Resource requirement (need help) Document Control Systems, Procedures, Forms and Templates 5
D ISO 14971:2019 vs MDR Annex 1, Requirement #4 - "Manufacturers shall inform users of any residual risks" ISO 14971 - Medical Device Risk Management 5
B IATF16949 audit requirement - Auditor request UCL and LCL must be show Xbar-R, IATF 16949 - Automotive Quality Systems Standard 7
J 21 CFR 821 Medical Device Tracking Requirement 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
J Is Device Tracking Card an actual requirement under FDA regulation? Other US Medical Device Regulations 0
W Misinterpretation of requirement acceptable as root cause? Problem Solving, Root Cause Fault and Failure Analysis 19
S ECG Cable Banana to Snap or Tab electrode Adapters -- ISO 10993 Requirement Other Medical Device Related Standards 0
D IATF 16949 Requirement for CMMI in a Global Company Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 0
W Regarding Requirement for Hi pot & Leakage Current test Station Set-up IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
L Stage 2 audit - Requirement for 3 months of records ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
M Requirement to manufacture under a quality management system EU Medical Device Regulations 4
A Risk Number for each software requirement IEC 62304 - Medical Device Software Life Cycle Processes 7
T ISO 17025:2017 requirement 5.7.b. about maintenance the integrity of the management system ISO 17025 related Discussions 1
I Laboratory Fridge / Freezer Inventory - Requirement(s) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
A Several sections of traditional 510(K) have the same or similar requirement of content. US Food and Drug Administration (FDA) 4
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
C Compliance with ISO 17025 requirement 8.4.2 - Controls - Records recovery ISO 17025 related Discussions 4
Q User Requirement Specification for HR (Human Resource Management System) Manufacturing and Related Processes 1

Similar threads

Top Bottom