Hello everyone
This is my first post on elsmar. I am a test engineer(recently joined and new in medical standards).
As per clause 5.8 of IEC 60601-1, it is recommended to do testing in the sequence given in annex B and annex B is saying to do Humidity preconditioning treatment(Cl. 5.7) after risk management related clauses. When I am seeing old reports to take some reference, people are doing touch current tests(Cl. 8.7) both before and after Humidity test.
Is this the right way of doing touch current testing?
This is my first post on elsmar. I am a test engineer(recently joined and new in medical standards).
As per clause 5.8 of IEC 60601-1, it is recommended to do testing in the sequence given in annex B and annex B is saying to do Humidity preconditioning treatment(Cl. 5.7) after risk management related clauses. When I am seeing old reports to take some reference, people are doing touch current tests(Cl. 8.7) both before and after Humidity test.