Setting accuracy limit for a read out

#1
Hi all,

we always calibrate the whole measuring chain (logic, we think)

But for a new process we need very dry air with a dew point of -60 °C
All the suppliers of measurement systems for this kind of dry air only calibrate the sensor and its mA output.
They claim that a readout never drift !

How do we set the limit for the readout that we will be calibration using a mA loop calibrator?
(We normally only use those for trouble shooting)

Do we split the entire tolerance in two parts, half of the tolerance for the sensor and half tolerance for the the readout?
Or
Do we use the output that is written on the certificates and supply that to the readout
Than check the readout display against the reference value that was applied to the sensor?
 
Last edited:
Elsmar Forum Sponsor

Ninja

Looking for Reality
Staff member
Super Moderator
#2
They claim that a readout never drift !
Hello Johanna,
What type of "readout" are you working with?
If it is a PLC with a ROM lookup table...it doesn't drift. If it's an electronic display with a conversion algorithm...it doesn't drift.
It can break, but the ROM data won't drift.

I work(ed) a lot with thermocouples, tied to a readout for either monitoring or control (single display, PID or PLC based).
We attributed ALL errors to the thermocouple, since the readout is simply a computer translating mV to a degree value based on a ROM lookup table.

My guess is that your sensor and display work the same way (but I used the word "guess", so don't stop there...)

HTH
 

dwperron

Trusted Information Resource
#3
All the suppliers of measurement systems for this kind of dry air only calibrate the sensor and its mA output.
They claim that a readout never drift !


The readouts do not usually drift, but they can and should be calibrated.
The probes for low dewpoint sensors definitely do drift! One thing you need to be sure off is that when you get the sensor recalibrated that they give you the as found measurements. There is a really good chance that it will be found to be out of tolerance at one of the extremes of its range.
 
#4
If it is a PLC with a ROM lookup table...it doesn't drift. If it's an electronic display with a conversion algorithm...it doesn't drift.
It can break, but the ROM data won't drift.
The LUT indeed is part of the ROM and wont change.
but the input cards (the electronics that are reading the signal) can drift.


I work(ed) a lot with thermocouples, tied to a readout for either monitoring or control (single display, PID or PLC based).
We attributed ALL errors to the thermocouple, since the readout is simply a computer translating mV to a degree value based on a ROM lookup table.
In this case the mV input card can show drift.
But compared to the drift expected from thermocouples, it is indeed a small factor.

The probes for low dewpoint sensors definitely do drift! There is a really good chance that it will be found to be out of tolerance at one of the extremes of its range.
We received from one sensor manufacturer a white paper claiming a drift lower than 3 °C dp over 12 months.
We will anticipate by drying the air to a level lower than -65 °C dp and start with a calibration interval of 6 months


One thing you need to be sure off is that when you get the sensor recalibrated that they give you the as found measurements.
As found and as left measurements are mandatory in our quality manual.
We consider it the only correct way of calibration.

The readouts do not usually drift, but they can and should be calibrated.
Do you split the tolerance between the sensor and the readout?
Or
Do you use the outputs mentioned on the certificate to simulate the readout and check if the complete loop is in tolerance?

Both approaches should have pros and cons,
but we have no practical experience in breaking the loop in several part for calibration (only for troubleshooting purposes)
 

dwperron

Trusted Information Resource
#5
Do you split the tolerance between the sensor and the readout?
Or
Do you use the outputs mentioned on the certificate to simulate the readout and check if the complete loop is in tolerance?

Both approaches should have pros and cons,
but we have no practical experience in breaking the loop in several part for calibration (only for troubleshooting purposes)


I have handled these two ways.
Normally I just send the sensor off for calibration and use a process calibrator to calibrate the readout. The sensor calibration results will expect that you are using a calibrated instrument as a readout, and you are correct that errors in the readout are typically very small compared to the sensor.

if the sensor and readout are being calibrated as a set I would typically combine their tolerances using the RSS method.
 
Thread starter Similar threads Forum Replies Date
B Setting flexible packaging specification limits ISO 14971 - Medical Device Risk Management 4
I Setting action completion dates Nonconformance and Corrective Action 3
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
S Heat Sealers Upper and lower limit setting during OQ for Heat Sealer Qualification and Validation (including 21 CFR Part 11) 0
Ed Panek GDPR in Urgent Healthcare Setting Other ISO and International Standards and European Regulations 1
M Setting deadlines (ex. 45 days) for Document Registration & Review Cycle Document Control Systems, Procedures, Forms and Templates 3
Ashok sunder What are the criteria for setting a target KPI value against a quality objective? Benchmarking 7
K Setting Quality Objectives (for the first time) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 17
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
M Best Practice for setting tolerances on a Drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
T Setting up a quality department in an Engineering Firm Quality Management System (QMS) Manuals 11
B Software for setting up New Warehouse Manufacturing and Related Processes 0
S Setting up the process - First piece inspections Statistical Analysis Tools, Techniques and SPC 3
A Add-On software for VMC to calculate setting time, Idle Time, OEE etc.. Manufacturing and Related Processes 2
G Setting Up a Basic Quality Lab for a small manufacturing company General Measurement Device and Calibration Topics 2
A Setting up an LLC as a Consultant Paid Consulting, Training and Services 27
A Setting initial Expiry Date of an IVD per BS EN 23640:2015 Other Medical Device Related Standards 12
Q Setting Quality Objectives with a Timeframe ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Proud Liberal Setting up Prolink's QC-Gage to a Keyence IM-6225 Vision System Statistical Analysis Tools, Techniques and SPC 2
P Setting Acceptance Quality Limits (AQL) for New Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
Hami812 Cpk Setting Tolerance - Cart before the Horse? (Wifi Routers) Capability, Accuracy and Stability - Processes, Machines, etc. 3
K Bioburden 0 cfu for Dose Setting using the VDmax25 Other Medical Device Related Standards 9
A New company, no employees, Setting up a QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
F Setting up a new sterilization unit in a new facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
C ISO 14001 Recertification without setting Targets ISO 14001:2015 Specific Discussions 3
T Guidance - Setting up an integrated quality & environmental management system Career and Occupation Discussions 12
M Setting up a M1003 QA Program for the AAR Various Other Specifications, Standards, and related Requirements 2
A Setting up a Medical Device Distributor in Malaysia Other Medical Device Regulations World-Wide 16
R Irradiation Sterilization Dose Setting and Dose Audit Flow Charts ISO 13485:2016 - Medical Device Quality Management Systems 0
J Setting Up New QMS - New US Location ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
A Setting air gauging outside of a part tolerance Capability, Accuracy and Stability - Processes, Machines, etc. 2
S Target Setting - Internal metric(s) to drive Internal Processes Six Sigma 1
S MSA Help or guidelines - Plastics - Setting up an MSA on a multi cavity fixture Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G Setting up a corporate umbrella QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Setting Quality Objectives ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
W Setting up from scratch - Looking for help (ISO9001 / ISO13485) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F Setting up a New QMS for Certification when ISO 9001 is about to Change ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
P New to AS9100 Rev. C - Setting KPIs and FAI Reports AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
B Setting Lower Control Limit on Attributes Charts Statistical Analysis Tools, Techniques and SPC 9
J Lean 5S - Setting Min/Max Inventory Levels Lean in Manufacturing and Service Industries 3
V Setting Specification Limits for UOD (Uniformity) for Live Bacteria or Spores Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
D How do I make the Corporate QMS work in the manufacturing setting? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
J Setting Quality Objectives for Service, Repair and Sales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Q Setting Specification Limits Quality Tools, Improvement and Analysis 11
C Mitutoyo SJ210 Settings - Is there an Industry Standard Surface Measurement Setting? General Measurement Device and Calibration Topics 4
W Setting up a Zmike for Pin Gage Calibration General Measurement Device and Calibration Topics 3
W Zero setting of correlation graph when plotting best fit linear regression Statistical Analysis Tools, Techniques and SPC 7
G Setting up a 3 Factor, 2 Level,2 response DOE in Minitab Using Minitab Software 7
B Quality Goals and Improvements in a Bioanalytical Laboratory Setting Quality Tools, Improvement and Analysis 1
N Setting up MSA system in a machine shop environment. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5

Similar threads

Top Bottom