Setting flexible packaging specification limits
We are using a heat sealed flexible pouch to vacuum package a product. We're using ASTM testing standards to characterize puncture strength, tear strength, and abrasion resistance.
How does one set / justify a specification limit for a heat sealable metallized film material? for instance, for the puncture resistance, is there a rule of thumb that for certain applications (this one is a medical device) the material should resist X newtons at puncture failure? tear strength should be R newtons for ease of opening? abrasion resistance >N cycles rub testing?
I'm looking for a way to draw a link between real world transport and use and the standards. I am appreciative of any feedback.
indubioush
You need to tie this back to your risk analysis. What would be the sequence of events and hazard associated with a punctured pouch? Is this even a foreseeable event? What do you expect your pouch to undergo in the real world in a worst case scenario?
We have a pretty comprehensive risk analysis we're working from. these will be carried in the field and will also need to test to MIL-STD shipping standards.
I am trying to draw some correlation between real world handling and how that relates to ASTM testing. for instance, the pouches are vacuum sealed and form fitting, so how can I relate that to results that come from a standardized puncture test fixture and similarly for other tests.
indubioush
Exactly. If you can't draw a correlation between what happens in the real world and the testing you will be doing, you need to ask yourself whether the testing is value added.
What ASTM testing standards are you referring to? Also, when you characterize something, that does not necessarily mean you will be developing a specification right away. You need to determine and document your product requirements based on customer requirements and regulations, not some "rule of thumb" that you heard from somewhere.
For example, like you stated, you could have a requirement regarding the ease of tearing open a pouch. Your design input truly can be "pouch must be easily opened by users." So then you do a characterization study where you test the tear strength of various materials and evaluate the ease of tearing each, both by hand and by measuring the force on an instrument. This could lead to the design output that describes the material to be used for your pouch. Hope that helps.
Tidge
In the case of ASTM standards, you can reach out to the members of the responsible committee to discuss with them the applicability of any particular standard. It is possible that a member may direct you to other (more appropriate) standards.
If there is a "standards gap" relating to a material, process or product... standards committees can get very excited!
K ISO 9001 Auditing in a Healthcare setting Process Audits and Layered Process Audits 15
J Setting accuracy limit for a read out General Measurement Device and Calibration Topics 4
I Setting action completion dates Nonconformance and Corrective Action 3
S Qualification question - ISO 13485 - Setting up a small lab Reliability Analysis - Predictions, Testing and Standards 2
S Heat Sealers Upper and lower limit setting during OQ for Heat Sealer Qualification and Validation (including 21 CFR Part 11) 0
Ed Panek GDPR in Urgent Healthcare Setting Other ISO and International Standards and European Regulations 1
M Setting deadlines (ex. 45 days) for Document Registration & Review Cycle Document Control Systems, Procedures, Forms and Templates 3
Ashok sunder What are the criteria for setting a target KPI value against a quality objective? Benchmarking 7
K Setting Quality Objectives (for the first time) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 17
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
M Best Practice for setting tolerances on a Drawing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 14
T Setting up a quality department in an Engineering Firm Quality Management System (QMS) Manuals 11
B Software for setting up New Warehouse Manufacturing and Related Processes 0
S Setting up the process - First piece inspections Statistical Analysis Tools, Techniques and SPC 3
A Add-On software for VMC to calculate setting time, Idle Time, OEE etc.. Manufacturing and Related Processes 2
G Setting Up a Basic Quality Lab for a small manufacturing company General Measurement Device and Calibration Topics 2
A Setting up an LLC as a Consultant Paid Consulting, Training and Services 27
A Setting initial Expiry Date of an IVD per BS EN 23640:2015 Other Medical Device Related Standards 12
Q Setting Quality Objectives with a Timeframe ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Proud Liberal Setting up Prolink's QC-Gage to a Keyence IM-6225 Vision System Statistical Analysis Tools, Techniques and SPC 2
P Setting Acceptance Quality Limits (AQL) for New Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
K Bioburden 0 cfu for Dose Setting using the VDmax25 Other Medical Device Related Standards 9
A New company, no employees, Setting up a QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
F Setting up a new sterilization unit in a new facility 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
C ISO 14001 Recertification without setting Targets ISO 14001:2015 Specific Discussions 3
T Guidance - Setting up an integrated quality & environmental management system Career and Occupation Discussions 12
M Setting up a M1003 QA Program for the AAR Various Other Specifications, Standards, and related Requirements 2
A Setting up a Medical Device Distributor in Malaysia Other Medical Device Regulations World-Wide 16
R Irradiation Sterilization Dose Setting and Dose Audit Flow Charts ISO 13485:2016 - Medical Device Quality Management Systems 0
J Setting Up New QMS - New US Location ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
A Setting air gauging outside of a part tolerance Capability, Accuracy and Stability - Processes, Machines, etc. 2
S Target Setting - Internal metric(s) to drive Internal Processes Six Sigma 1
S MSA Help or guidelines - Plastics - Setting up an MSA on a multi cavity fixture Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
G Setting up a corporate umbrella QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
T Setting Quality Objectives ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
W Setting up from scratch - Looking for help (ISO9001 / ISO13485) ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
F Setting up a New QMS for Certification when ISO 9001 is about to Change ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 23
P New to AS9100 Rev. C - Setting KPIs and FAI Reports AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
B Setting Lower Control Limit on Attributes Charts Statistical Analysis Tools, Techniques and SPC 9
J Lean 5S - Setting Min/Max Inventory Levels Lean in Manufacturing and Service Industries 3
V Setting Specification Limits for UOD (Uniformity) for Live Bacteria or Spores Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 4
D How do I make the Corporate QMS work in the manufacturing setting? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
J Setting Quality Objectives for Service, Repair and Sales ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Q Setting Specification Limits Quality Tools, Improvement and Analysis 11
C Mitutoyo SJ210 Settings - Is there an Industry Standard Surface Measurement Setting? General Measurement Device and Calibration Topics 4
W Setting up a Zmike for Pin Gage Calibration General Measurement Device and Calibration Topics 3
W Zero setting of correlation graph when plotting best fit linear regression Statistical Analysis Tools, Techniques and SPC 7
G Setting up a 3 Factor, 2 Level,2 response DOE in Minitab Using Minitab Software 7
B Quality Goals and Improvements in a Bioanalytical Laboratory Setting Quality Tools, Improvement and Analysis 1