I am having a disagreement with the person who is training me to teach FMEA classes. I wrote a blog post in which I used an example FMEA that had a severity of 10. The actions taken involved a design change that eliminated the potential failure effect so I made a new estimate for the severity.

My trainer told me that I was wrong to do this. He said that in the revised metrics section of the FMEA I should have kept the severity at 10 and re-estimated the occurrence and detection based upon the improvements. This particular failure can?t occur due to the listed failure mode so I believe a new severity is appropriate.

We are in agreement that the new design should also be evaluated as a new point in the FMEA; however, I am convinced that the severity can be reduced with a proper design change. The AIAG manual concurs; however, my trainer believes that this new severity is used in a new line and not as part of the revised metrics.

I disagree with my trainer, but have agreed to teach things his way until such time as I can provide concrete evidence that I am correct.
Can anybody please provide that evidence? Or, explain to me why I am wrong.

Thanks!
 
Elsmar Forum Sponsor

somashekar

Staff member
Super Moderator
#2
Re: Severity reduction in a design FMEA

Is the design change something like making the 10000 V shock to say 1000 v shock in the unlikely chance of a shock delivery ...
Change the severity rating.
Is the design change something like making the probability of the 10000 V shock delivery much more unlikely ...
No change in the severity rating as still a 10000 V shock can be delivered.
 
M

Matt_B

#3
Re: Severity reduction in a design FMEA

My change is like making the 10000v shock only 1000v. In the example, I had systems connected in such a way that damage to an external area was transferred directly to a critical system that led to a catastrophic loss of the entire unit. The design change was to remove the interlinks so surface damage is not directly transmitted inside of the unit.
Unfortunately, I need evidence that I can present to my trainer.
 

somashekar

Staff member
Super Moderator
#4
Re: Severity reduction in a design FMEA

You are doing stuff only on the occurrence. However if in any case, it is still directly transmitted inside to the critical system., the catastrophic loss of the entire unit still exists. The severity exists .... Right ?
 
M

Matt_B

#5
Re: Severity reduction in a design FMEA

In the example, the catastrophic loss was no longer possible; hence, the severity reduction.
Regardless of my example, I say a severity reduction is possible with a design change and the new severity can be listed with the new RPZ on the right side of the FMEA form. I just can?t find any evidence to prove this.
 

somashekar

Staff member
Super Moderator
#7
Re: Severity reduction in a design FMEA

In the example, the catastrophic loss was no longer possible; hence, the severity reduction.
Regardless of my example, I say a severity reduction is possible with a design change and the new severity can be listed with the new RPZ on the right side of the FMEA form. I just can?t find any evidence to prove this.
In the example, if you can say the catastrophic loss is now a minor loss, and such a catastrophic loss will no longer manifest, then your design had addressed the severity and you are good to reduce it...
 
M

Matt_B

#8
Re: Severity reduction in a design FMEA

In the example, if you can say the catastrophic loss is now a minor loss, and such a catastrophic loss will no longer manifest, then your design had addressed the severity and you are good to reduce it...
I know; but my trainer and I agreed that I will do things his way until I can provide evidence or an example from a standard to prove that I was right.
He insists that the severity remain the same regardless of actions and the severity is only lowered by a new entry in the FMEA. He said the severity in the revaluation of the RPZ after corrective actions must remain the same.
 
A

adamsjm

#9
Re: Severity reduction in a design FMEA

What Design Requirement change was made to eliminate the Failure Mode (anti-function)?
Reduction of Occurrence to a 1 (Failure Eliminated) and/or a Detection of 1 (Error [Cause] prevention as a result of part design.) will NOT reduce severity, only a Design Requirement elimination will.
 

somashekar

Staff member
Super Moderator
#10
Re: Severity reduction in a design FMEA

Your design solution can be mitigating the severity or occurrence. So depending on the solution you are good to change whichever without a new entry in the dFMEA.
 
Thread starter Similar threads Forum Replies Date
B Severity Reduction by Product or Process Design Change? Seeking Good Example Design and Development of Products and Processes 2
R DFMEA/PFMEA mitigation of high severity (9-10) in low volume products IATF 16949 - Automotive Quality Systems Standard 1
silentmonkey How to measure severity if my product is designed for emergency use and failure would result in death? ISO 14971 - Medical Device Risk Management 9
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
S The Severity of a Medical Device Hazard - Risk Analysis Clarification ISO 14971 - Medical Device Risk Management 6
MrTetris Informational Risk Register - Same hazardous situation, different severity of harms ISO 14971 - Medical Device Risk Management 7
A PFMEA severity score different applications FMEA and Control Plans 11
X Customer Service FMEA - Specifically Customer Care / Severity Rankings Service Industry Specific Topics 3
D How to fill a Process FMEA - Three columns - Severity, occurrence and Detection FMEA and Control Plans 16
R DFMEA Severity 9-10 with Occurrence 3 and Detection 3 - Actions FMEA and Control Plans 3
S What should the FMEA Severity for SC & CC Characteristics be? FMEA and Control Plans 5
N AIAG 4th ed. FMEA Manual "Fit" Changed to "Appearance" - Fit severity? FMEA and Control Plans 2
G Severity in Process FMEA for Rework Process FMEA and Control Plans 1
O PFMEA - Reducing Severity (S) Ranking - Manufacturing Process Design Changes FMEA and Control Plans 4
S FMEA's - Severity and Classification of Print Characteristics FMEA and Control Plans 1
N Risk Severity Estimation for Medical Devices as per ISO 14971 ISO 14971 - Medical Device Risk Management 12
C Determining PFMEA Severity for Armored Parts FMEA and Control Plans 5
C FMEA DOGMA about Severity - Am I right or Wrong? FMEA and Control Plans 2
K 4th Edition AIAG FMEA Occurrence, Severity and Detection Criteria for Plastics FMEA and Control Plans 1
F Risk = Likelihood * Severity - Can we mitigate the Severity? Occupational Health & Safety Management Standards 44
E Risk Management selection Probability of Occurrence and Severity ISO 14971 - Medical Device Risk Management 24
S Severity of 9 or 10 should be identified as Special characteristic? FMEA and Control Plans 10
G Internal vs. External PFMEA Severity Rankings FMEA and Control Plans 12
R Risks which must be Distinctly Identified - Harm, Hazard, Severity ISO 14971 - Medical Device Risk Management 7
C Severity of Effect on Process (Manufacturing/Assembly Effect) Scrap or Reworking FMEA and Control Plans 3
W Severity Rating CORRECTIVE ACTIONS - Opinions Wanted Please Nonconformance and Corrective Action 15
H DFMEA Severity Ranking of Non-Compliance to EMC Standards FMEA and Control Plans 2
P Chemical Purity Testing - FMEA Severity Scale FMEA and Control Plans 2
T FMEA Severity 9 &10 - Example of "with warning?" FMEA and Control Plans 6
E Harms Severity Scoring - Combination of Hazards ISO 14971 - Medical Device Risk Management 4
B The Distribution that represents the Severity of Nonconformances Nonconformance and Corrective Action 2
A AIAG Design FMEA Severity Rankings - Noncompliance Government Regulation? FMEA and Control Plans 9
D Severity or Occurrence? Performing an FMEA on Motorcycle Operation FMEA and Control Plans 9
L FMEA Occurrence, Severity and Detection Criteria FMEA and Control Plans 3
I Ford FMEA requirements for Severity and Occurrence Rankings Customer and Company Specific Requirements 1
G FMEA Severity Ranking for an Automotive Audio System FMEA and Control Plans 9
Chennaiite FMEA Severity Rating for Potential Manufacturing Effect mentioned in the FMEA manual FMEA and Control Plans 9
kedarg6500 FMEA Severity Scale - Higher for Failure at Customer's or End-Customer's place? FMEA and Control Plans 19
M FMEA Severity Ratings for Health Care Information System FMEA and Control Plans 10
C Does the PFMEA Severity have to match the DFMEA Severity? FMEA and Control Plans 3
P Measuring Severity of Construction Nonconformance Report Nonconformance and Corrective Action 3
L Standard for process control level depending on gravity (severity)? FMEA and Control Plans 7
I Defining a Severity Level of a non-conformance Nonconformance and Corrective Action 13
Y Severity, Occurrence and Detection FMEA Ranking Tables - Guidelines? FMEA and Control Plans 8
L FMEA - Should a severity of 9 or 10 be always followed by recommended actions? FMEA and Control Plans 2
J Severity of Finding for "Past Due Gage" found on the shop floor General Measurement Device and Calibration Topics 40
S FMEA Severity of 9 or 10 - Shower does not operate FMEA and Control Plans 13
J Ranking Severity of Non-Conforming Material Defects Nonconformance and Corrective Action 5
V Emphasizing Severity more than Occurrence and Detection vs. Overall RPN FMEA and Control Plans 20
R Severity, Occurance and Detection ratings seem qualitative FMEA and Control Plans 14

Similar threads

Top Bottom