Shelf-Life - Accelerated Age Test

medwise

Involved In Discussions
#1
Dear All,

Our device is a class II (a) sterile and is used for skin closure.

This device used to be gamma irradiated but from last two years we use EtO.

The shelf life of the device is 5 years.

Packaging material and packaging machine is exactly the same. Manufacturing process is exactly the same. One change is the sterilization process. From gamma to EtO.

We do sterility test of the product which has undergone EtO and found that it is sterile.

My question is do I have to conduct age testing because we change from gamma to EtO?

We are doing EtO from 2008 so the only option is Accelerated age test., which we are happy to do. But I was wondering wouldn't risk assessment be sufficient (including the sterility test) because the packaging material and machine has been the same.

Any guidance will be much appreciated.

Thank you

Kindest regards,
Romit :)
 
Elsmar Forum Sponsor
O

ontheopenroad

#2
You would need to revalidate your shelf life (using accelerated aging as you suggested, followed-up by real time study). You should revalidate because:

a) change in sterilization method may have a negative (or even positive) impact on the package integrity.
b) change in sterilization method may have a negative (or even positive) impact on device integrity.

:2cents:
 

Statistical Steven

Statistician
Staff member
Super Moderator
#4
Romit -

Unfortunately at the time of the sterilization change you did not initiate a real time stability program. Based on the risk assessment, you could claim that a real time stability plan would be appropriate. You have to show that the change in method does not have a long term (stability) impact in the package and/or sterility of the product. Unless you have a good Arrenhius model, accelerated stability might not be effective. My :2cents:

Dear All,

Our device is a class II (a) sterile and is used for skin closure.

This device used to be gamma irradiated but from last two years we use EtO.

The shelf life of the device is 5 years.

Packaging material and packaging machine is exactly the same. Manufacturing process is exactly the same. One change is the sterilization process. From gamma to EtO.

We do sterility test of the product which has undergone EtO and found that it is sterile.

My question is do I have to conduct age testing because we change from gamma to EtO?

We are doing EtO from 2008 so the only option is Accelerated age test., which we are happy to do. But I was wondering wouldn't risk assessment be sufficient (including the sterility test) because the packaging material and machine has been the same.

Any guidance will be much appreciated.

Thank you

Kindest regards,
Romit :)
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#5
ask your reviewing body. The change from irraditiaon to eto is significant as this is a 'special process' and they are fundamentally different sterilization models.

accelerated testing is usually not acceptable. As Steven said most accelerated testing is not well correlated and often the model is not correct in real life.

In the context of biologicals, accelerated testing is often good for detecting differences but not in predicting life...
 
M

MIREGMGR

#6
accelerated testing is usually not acceptable.
In my experience, environmentally accelerated testing frequently is used to establish both sterile-barrier-packaging shelf life and product functional shelf life, so that the product can be brought to market while the real time testing proceeds.

I don't disagree that this approach is uncertain because of the imperfect validity of the acceleration concept, but it's the best alternative to waiting for the real-time period to introduce new medical developments...which itself would have a social cost in terms of improved medical care foregone.

The FDA explicitly expects accelerated testing with approval submissions, while real-time testing proceeds.

My FDA/EC/HC experience is mostly with regulatorily simple products, though. For what particular device classes is the "not acceptable" statement definitely the EC rule?
 

Bev D

Heretical Statistician
Staff member
Super Moderator
#7
I agree that for some devices accelerated testing is allowed and expected while real time studies proceed. for complex biologicals, this is not always true. it can vary depending on your reviewer...certainly pharmaceuticals and vaccines fall in this category although they are not medical devices. A diagnostic test is a medical device and they are complex biologicals that don't fit well into accelerated testing models... I have been restricted to real time testing studies when implementing a sterility change for irriadiation and eto for diagnostic devices... A well crafted plan for sterility properly presented will generally be accepted. that is where a collaborative relationship with your reviewing body helps...
 

medwise

Involved In Discussions
#8
Dear All,

Thank you for your quick response and help. I really appreciate it.:thanks::)

I just forgot to mention that we also want to change the storage temperature of the device from 25 degrees C to 45 degrees C. So can I just do accelerated aging test with 45 degrees C and then do the performance test of the device.

Also, since our device is a wound closure strip and is packaged in such a way that there will be no moisture inside the packaging. For the accelerated test do I still have to consider relative humidity, and if yes then what is the limit of RH%?

:thanx:

Kindest regards
Romit Singh:)
 
M

MIREGMGR

#9
That casts an entirely different light on the situation, in my opinion.

A change of that magnitude needs to be re-evaluated from a design perspective as to effect on device functionality, re-evaluated from a risk perspective, re-evaluated from a packaging perspective, and generally re-validated.

The unassured correspondence between real time behavior and accelerated environmental-conditions aging is a key factor in your question. I don't think accelerated aging would be in any way adequate to support such a storage temperature change.

Your moisture question needs to be addressed at the packaging-design-performance level. Accelerated aging by itself is not an appropriate tool for doing what you want to do.
 
Thread starter Similar threads Forum Replies Date
shrutisancheti Predicate device accelerated aging - 1 year of shelf life is already over Other Medical Device Related Standards 1
W Shelf life and accelerated aging ISO 13485:2016 - Medical Device Quality Management Systems 2
U Accelerated Aging for Shelf Life - Beginning Point Difference CE Marking (Conformité Européene) / CB Scheme 2
6 Accelerated Shelf Life Testing Protocol for Juice Containing Beverages Food Safety - ISO 22000, HACCP (21 CFR 120) 1
M Extension to shelf life of new product under MDR EU Medical Device Regulations 1
S Shelf Life Test for IFU and Implant Card Other Medical Device Related Standards 1
T Sterilized components shelf life Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
L Shelf life of medical devices Other Medical Device Related Standards 4
U Does *anyone* know a lab that will test to EN 455-4 Medical Gloves shelf life determination? EU Medical Device Regulations 1
C Shelf life electronic components 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
F Plastics cans shelf life determination under static load Manufacturing and Related Processes 1
M Medical Devices Shelf Life ---INDIA Other Medical Device Regulations World-Wide 3
S Requirement to Conduct New Shelf-life Testing? (re-do testing for design change) EU Medical Device Regulations 3
M PPE shelf life Medical Device and FDA Regulations and Standards News 1
K EU Shelf life claims for individual materials EU Medical Device Regulations 2
P Disposables Shelf Life Expiration US Food and Drug Administration (FDA) 2
C Medical Device Shelf life studies 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
D Standard guide for shelf life determination of non sterile component with degradation materials Reliability Analysis - Predictions, Testing and Standards 1
Ed Panek Is Shelf Life Change a significant change? CE Marking CE Marking (Conformité Européene) / CB Scheme 7
T No Defined Shelf Life/Expiration Date - Disposable single-use, non-invasive, non-sterile Other Medical Device Regulations World-Wide 2
H Frequency of shelf life/sterilization/bench studies conducted on a medical device Other Medical Device Related Standards 1
B Shelf life of molded and extruded pebax Manufacturing and Related Processes 2
J 510k - Product shelf life validation 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K Food Shelf Life Stability Study Statistical Analysis Tools, Techniques and SPC 1
I Device Shelf Life Extension to 48 months without involving the FDA 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Shelf Life of Combination Sets for Class II Medical Devices ISO 13485:2016 - Medical Device Quality Management Systems 4
Y Shelf Life of Non-Sterile Disposable Drug Inhaler 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
shimonv Over-Labeling Sterile Packaging - Shelf Life 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
N Sample Size Justification for Medical Device Shelf-Life Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
N Do ISO 9001 and ISO 13485 have requirements for Shelf-Life? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
J EU Medical Device Class I - Stability (Shelf Life) Test EU Medical Device Regulations 2
P Need ideas on managing Shelf Life Studies within an organization ISO 13485:2016 - Medical Device Quality Management Systems 1
D Class II Sterile Product Shelf Life Labeling - Should we do a recall? Nonconformance and Corrective Action 4
R FDA Shelf Life Requirements for EO Surgical gown 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
J Non-Sterile Class 2 Medical Device Shelf Life Other Medical Device Related Standards 3
P Does resin have expiry date (shelf life)? Manufacturing and Related Processes 11
C Best Method of Labeling Shelf Life Containers of Urethanes and Plastics Manufacturing and Related Processes 5
A Medical Device Shelf Life vs. Warranty Period Canada Medical Device Regulations 6
N Does Class I (exempt) device have requirements of Shelf Life and Expiration Dating ? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Shelf Life, Expiration Date & Labeling AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
M Shelf Life Control and Identification of items that do not have Shelf Life ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B Does Nitric Acid Expire? (Shelf Life) Manufacturing and Related Processes 11
U Shelf Life Requirement for Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 26
E Is Shelf Life extension a substantial change? EU Medical Device Regulations 6
S Shelf Life Determination - Sterility test (Stasis test) ISO 13485:2016 - Medical Device Quality Management Systems 12
somashekar Authorization to use shelf life sensitive materials beyond its valid shelf life... Manufacturing and Related Processes 7