Roland chung
Trusted Information Resource
Hi all,
When an IEC 60320 appliance outlet is used to construct an ME system, does clause 16.9.2.1 (multiple socket-outlet) apply?
- If applies, the appliance outlet which is IEC 60320-2-2 approved already would be required to comply with IEC 60884-1 also. It seems to me unreasonable.
- If not, there would be no requirement for classification of multiple socket-outlet. According to 16.9.2.1 c), second dash), only Class I construction is allowed. I would think Class II appliance outlet should not be used.
Furthermore, I am very confused about the 16.9.2.1 c), second dash. Does this mean that multiple socket-outlet mounted on Class II ME equipment without functional earth (Class II ME equipment is an item of the ME system) shall also be Class I construction? Or does the standard reject the Class II ME equipment without functional earth to construct ME system by means of multiple socket-outlet?
Please help to clarify. Thank you very much!
Roland
Roland