Should Corrective Action be segregated from Preventive Action?

sagai

Quite Involved in Discussions
#1
Do we have to have a segregated record for corrective action and preventive action?
Is there any problem recording CAPA in general without saying This is CA, That is PA?
(I know the difference, I am wondering if we should distinguish them when recording.)
I tend to believe there is no such requirement in the standard, but curious about others opinion, Many thanks.
 
Elsmar Forum Sponsor
G

Geoff Withnell

#2
Do we have to have a segregated record for corrective action and preventive action?
Is there any problem recording CAPA in general without saying This is CA, That is PA?
(I know the difference, I am wondering if we should distinguish them when recording.)
I tend to believe there is no such requirement in the standard, but curious about others opinion, Many thanks.
I can't think of any requirement to do so. Some flag on the PAs, to make it easier to pull them out as examples during audits, may be worthwhile, but required, no.
 

RA Guy

Involved In Discussions
#3
13485 has separate clauses for CA and PA, so I tend to think there is a requirement to identify actions as CA or PA. Otherwise how do you demonstrate complying with both clauses in an audit situation? At a minimum, the CAPA determination step would need to identify if a CA or PA.
 

sagai

Quite Involved in Discussions
#4
What if the PAs are those CAPAs having no registered NonConformities against to, while CAs are the ones having registered NonConformities to?
Many thanks.
 
G

Geoff Withnell

#6
The OP said he knew the difference. The question is "Do we have to have a segregated record for corrective action and preventive action?" No, the records do not need to be segregated. In an audit situation, if the auditor says "show me some corrective actions", show him some. The MR should be familar with the actions being taken, and can likewise ID PAs. If you want to flag them for your convenience, it is perfectly fine to do so, and I would even recommend it. But being the easiest, most convenient way to do it, in my opinion, your opinion, the OP's opinion or ewven the auditor's opinion does NOT make it "required".
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#7
Do we have to have a segregated record for corrective action and preventive action?
Is there any problem recording CAPA in general without saying This is CA, That is PA?
(I know the difference, I am wondering if we should distinguish them when recording.)
I tend to believe there is no such requirement in the standard, but curious about others opinion, Many thanks.
When ISO 9001:2000 came about, 11[sup]1[/sup]/[sub]2[/sub] years ago, a well known CB set a policy that registrants certified by them were mandated to have SEPARATE procedures for corrective and preventive actions. They got a lot of backlash and have rescinded the policy. Actually one of the clarifications in the ISO 9001:2008 standard had to do with that aspect.

As for your question concerning segregating corrective action records from preventive action records, you should be able to distinguish which is which, even though there is a lot of non-sense, when it comes to preventive actions. We have some of the longest threads on that subject, here at The Cove. Also, don't forget the wisdom contained in the ISO TC 176 APG paper titled Auditing Preventive Action.
 
Last edited:

RA Guy

Involved In Discussions
#8
Hi Geoff:

Perhaps we are not reading "segregated" in the same way. I understand it as indicating (or distinguishing as SV states above) that a CA is a CA and a PA is a PA. As far as I am concerned one can use the same process, templates, logs, etc. for both CAs and PAs.

If there is no flag or other mechanism to "segregate" the CAs from the PAs, how is the "if its not documented, its not done" rule of thumb satisfied? It can't be by the memory or instiutional knowledge of the MR or anyone else. What if they leave the company?

In an audit, anything that comes out verbally, must be backed up by the record or other objective evidence. If the auditee is the only source verbally telling the auditor what's what, I'd suggest that the auditor is injecting his/her opinion by accepting that with no documented evidence.
 

sagai

Quite Involved in Discussions
#9
I know not I was the one was called, but sorry Sir, "objective evidence" is not necessary a document, a record or anything tangible. As far as I concern in the ISO world. Cheers!

---
3.8.1
objective evidence
data supporting the existence or verity
of something

NOTE Objective evidence may be ob-
tained through observation
, measure-
ment, test (3.8.3), or other means.
 

RA Guy

Involved In Discussions
#10
Sagai, I agree and that's why I said "by the record or other objective evidence."
I am not sure in your CAPA example what other type of evidence could be presented besides a document or record.
My original post stated 13485 has 2 separate requirements. Both requirements need to be demonstrated by objective evidence. The record is the best way in MHO.
I am open to other suggesteions on how this could be done without documentary evidence "segregating" the 2 types of action.
 
Thread starter Similar threads Forum Replies Date
R When should a Corrective action be created? (Procedures not being followed) Nonconformance and Corrective Action 23
A Whom should I send the FSCA (Field Safety Corrective Action) to? EU Medical Device Regulations 2
C Should Preventative Action follow a Corrective Action Nonconformance and Corrective Action 27
J What should generate a CAPA (corrective action - preventive action) US Food and Drug Administration (FDA) 16
Q How many records should I keep when implementing Corrective Actions Records and Data - Quality, Legal and Other Evidence 3
Q CAPA - What criteria should be used to initiate Corrective and Preventive Actions ISO 13485:2016 - Medical Device Quality Management Systems 8
T Who should take the lead for 8D Corrective Actions Preventive Action and Continuous Improvement 13
S Who should do the CA's (Corrective Actions) from Internal Audits Internal Auditing 9
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 4
Ed Panek External Standards List - Should this document include previously revised standards? ISO 13485:2016 - Medical Device Quality Management Systems 4
T How should I approach REACH, CM, etc. as a job shop? RoHS, REACH, ELV, IMDS and Restricted Substances 18
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Should I take an online course for a career in Occupational Health and Safety? Career and Occupation Discussions 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
K Should APQP/PPAP has its own section in a QM? Quality Management System (QMS) Manuals 1
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
S Which Sampling Plan(s) Should I Use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
A Document release vs its related training. Which should come first? ISO 13485:2016 - Medical Device Quality Management Systems 18
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Master Samples - What should we be keeping? IATF 16949 - Automotive Quality Systems Standard 9
G Supplier delivered recent PPAP, should he deliver yearly layout inspection? IATF 16949 - Automotive Quality Systems Standard 4
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
J Should Loading and Unloading be Included in Cycle Times? Lean in Manufacturing and Service Industries 14
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
S Should safety checks be included in the Control Plan? IATF 16949 - Automotive Quality Systems Standard 5
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
D Is there a specific location for PPE such as safety glass holders and glove dispensers should be mounted Occupational Health & Safety Management Standards 10
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M Who should receive the bills from suppliers and vendors, account payable or procurement? Consultants and Consulting 4
V IATF 16949 8.4.1 Control of externally provided processes, products and services - Should the CB be on our Approved Supplier List? IATF 16949 - Automotive Quality Systems Standard 10
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6

Similar threads

Top Bottom