Should I perform Gage R&R only at the beginning of a new project?

Graciel

Involved In Discussions
#1
Dear all,

should I perform R&R only at the beginning of a new project. Will I never need to reproduce again? (If it´s based on IATF, does it say exactly when?)

Thank you.
 
Elsmar Forum Sponsor

Jim Wynne

Staff member
Admin
#2
MSA should not have to be repeated unless there have been significant changes in the system. The object is to demonstrate that a particular measurement system, including devices and operators, is appropriate for measurement of a particular dimension or feature.
 

Ninja

Looking for Reality
Staff member
Super Moderator
#3
Jim is correct, but consider what he says in detail...

If you've done a GRR for measurements by this method for 1-4 inches before...but the new project is using the same method for measuring 12 inches...
Consider that things like this may be "significant changes in the system"...even if the technique and tools remain the same.
 

Miner

Forum Moderator
Staff member
Admin
#4
should I perform R&R only at the beginning of a new project.
Please clarify what you mean by new project. Do you intend a new product introduction, or are you referring to a continuous improvement project such as Six Sigma?
 

Graciel

Involved In Discussions
#5
Please clarify what you mean by new project. Do you intend a new product introduction, or are you referring to a continuous improvement project such as Six Sigma?
hI, I´m refering if I need to perform MSA to each new product, product change or process change.
 

Ninja

Looking for Reality
Staff member
Super Moderator
#6
I used a drop gage to measure the thickness of product ranging from 12um through 1.1mm...same gage, same technique.

I used product from 12um through 0.010" (~250um) for the GRR, and included a "known" sheet of Mylar at 75um as one of the 10 parts.

When it came time to do 1.1mm (1100um) I considered how the gage worked (linear displacement up to 1.000inch) and chose to do a 2x2x5 R&R on parts from 0.010" through 1200um. Saw no issue, and moved on.
In the mean-time, we launched at least 60 new products in the 12um - 250um range...never even considered doing 60 more GRR's.
Apply logic...the goal is to have accurate enough measurement to know good from bad...

After establishing a "known good measuring system within these parameters"...product changes and process (for making the product) changes have nothing to do with the measurement system.
 

Miner

Forum Moderator
Staff member
Admin
#7
I agree with the previous comments. The only reason to redo an MSA is if :
  • there is a fundamental change in the measurement system,
  • you are using the measurement system to measure a fundamentally different feature (width vs. diameter),
  • Using the device in a different way (outside diameter to inside diameter to step height for calipers)
  • measuring in an entirely different range of the device. Note: this would normally be covered by Linearity, but could have an impact on R&R for certain devices. For example, the size of the feature might affect how an operator holds the product and gage.
 
Thread starter Similar threads Forum Replies Date
B AS9102 FAI & Lower Level Drawings - How should we perform the FAI? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
V Product Audits - Who should perform Product Audits? IATF 16949 - Automotive Quality Systems Standard 17
B Should we perform Equipment final calibration-Operations closing ISO 13485:2016 - Medical Device Quality Management Systems 8
M Customer is also my supplier - Should I perform an audit to his process? IATF 16949 - Automotive Quality Systems Standard 6
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 4
Ed Panek External Standards List - Should this document include previously revised standards? ISO 13485:2016 - Medical Device Quality Management Systems 4
T How should I approach REACH, CM, etc. as a job shop? RoHS, REACH, ELV, IMDS and Restricted Substances 18
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Should I take an online course for a career in Occupational Health and Safety? Career and Occupation Discussions 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
K Should APQP/PPAP has its own section in a QM? Quality Management System (QMS) Manuals 1
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
S Which Sampling Plan(s) Should I Use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
A Document release vs its related training. Which should come first? ISO 13485:2016 - Medical Device Quality Management Systems 18
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Master Samples - What should we be keeping? IATF 16949 - Automotive Quality Systems Standard 9
G Supplier delivered recent PPAP, should he deliver yearly layout inspection? IATF 16949 - Automotive Quality Systems Standard 4
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
J Should Loading and Unloading be Included in Cycle Times? Lean in Manufacturing and Service Industries 14
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
S Should safety checks be included in the Control Plan? IATF 16949 - Automotive Quality Systems Standard 5
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
D Is there a specific location for PPE such as safety glass holders and glove dispensers should be mounted Occupational Health & Safety Management Standards 10
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M Who should receive the bills from suppliers and vendors, account payable or procurement? Consultants and Consulting 4
V IATF 16949 8.4.1 Control of externally provided processes, products and services - Should the CB be on our Approved Supplier List? IATF 16949 - Automotive Quality Systems Standard 10
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Companies that maintain your machine should be in ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Use of "Shall" versus "Should" in Procedures ISO 13485:2016 - Medical Device Quality Management Systems 26
D Class II medical device - When should a complaint be closed? Customer Complaints 6
Sidney Vianna IATF 16949 News Presentations from the latest IATF Stakeholder Event - Expectation that IATF 16949 certification should equate with product quality. Misguided? IATF 16949 - Automotive Quality Systems Standard 7

Similar threads

Top Bottom