Should the Calibration Laboratory be kept locked?

C

Corsair

#1
Hello,
Recently, I went to visit a supplier for a QMS audit and found their internal Calibration Lab open and unlocked, The AS9100 do not mention that the Lab- needs to be locked at all times and the supplier internal procedure does not mentioned neither.
So is this a GMP or needs to be stated somewhere ?
You can't write them up if is not in a procedure or Standard.
Thanks in advance:thanx:
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#2
Re: Should the Calibration Lab be kept locked ?

What is the general nature of the lab? Do they need to garb to go in?

What type of calibrations are performed?

Are calibration standards sitting out where they could be adjusted? Are there fragile/ delicate/ expensive equipment that could be damaged?

Do they have a process where people are instructed to bring their instruments to the lab for calibration? Drop off and pick up?

did they have instruments/equipment sitting around with inappropriate labels on them?

I got this from an A2LA document:

5.3 S.3 There shall be provisions for restricting access to the field laboratory when unrestricted access could nvalidate the test or calibration results.
But again, it depends on how the lab is being used.

If you are going to suggest the lab be locked, there should be a compelling reason why. Locked doors/ rooms are a pain to work with. :)
 

Ron Rompen

Trusted Information Resource
#3
Re: Should the Calibration Lab be kept locked ?

I will echo Brad's comments. If there isn't a compelling reason to lock the lab door, then it shouldn't be required. Yes, there may be times when it IS appropriate, but in general I don't see the need.
I have worked in plants where EVERY access was locked, and it was a complete pain in the butt to try and get anything done.
 
C

Corsair

#4
Thanks for the feedback.
It is an aerospace medium size organization with capabilities for gauges, mic's, meters, etc-. I would have the Lab- locked just to safeguard the equipment and records.
I guess will be wise to go with the A2LA Std then.
 

drgnrider

Quite Involved in Discussions
#6
Adding to BradM's list -
.. is there a problem where people take items before they are ready to be returned for use?
 

dwperron

Trusted Information Resource
#7
This is something that is covered in ISO 17025:

5.8.1 The laboratory shall have procedures for the transportation, receipt, handling, protection, storage, retention and/or disposal of test and/or calibration items, including all provisions necessary to protect the integrity of the test or calibration item, and to protect the interests of the laboratory and the customer.

In a 17025 world it would be the responsibility of the lab to demonstrate their procedures as to how they would keep calibration items secure. It's just good practice to maintain security and the integrity of the items in the lab.
 
Thread starter Similar threads Forum Replies Date
Anerol C Where should I start as the person responsible for the calibration laboratory? General Measurement Device and Calibration Topics 7
A Preparing manual for Calibration laboratory! How should it be named? General Measurement Device and Calibration Topics 1
C Should resolution be included in uncertainty budget for digital caliper or micrometer calibration? Measurement Uncertainty (MU) 5
J What Calibration Standards should I have? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Damaged Load Cell After Calibration - Who Should Pay? General Measurement Device and Calibration Topics 3
Ajit Basrur Should calibration range of Oscilloscope cover the test ranges? General Measurement Device and Calibration Topics 8
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
V Should there be a incident/deviation for breakdown or missed calibration date? Nonconformance and Corrective Action 7
M Should a low cost hand held refractometer be in a calibration program? Calibration Frequency (Interval) 8
B Should we perform Equipment final calibration-Operations closing ISO 13485:2016 - Medical Device Quality Management Systems 8
J CNC Calibration - What should be calibrated on a HAAS machine? General Measurement Device and Calibration Topics 7
K Should QS/TS certified internal test/calibration labs be required to be ISO17025 ISO 17025 related Discussions 16
J Should weigh count scales be in the Calibration System? QS-9000 - American Automotive Manufacturers Standard 4
J Calibration, should we.... Go-No-Go, thread, pin, and ring gauges QS-9000 - American Automotive Manufacturers Standard 2
H When should the first PSUR be issued? EU Medical Device Regulations 5
I If i do not want to be an initial importer should i register with FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
O Should a Covid vaccine and testing policy be included as part of ISO9001 or AS9100 risk management? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 6
B UKRP to what level should you audit Class I Technical Documentation? UK Medical Device Regulations 0
C When should you quit programming? Job Openings, Consulting and Employment Opportunities 3
Ed Panek External Standards List - Should this document include previously revised standards? ISO 13485:2016 - Medical Device Quality Management Systems 4
T How should I approach REACH, CM, etc. as a job shop? RoHS, REACH, ELV, IMDS and Restricted Substances 18
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
A Should I take an online course for a career in Occupational Health and Safety? Career and Occupation Discussions 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
K Should APQP/PPAP has its own section in a QM? Quality Management System (QMS) Manuals 1
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
S Which Sampling Plan(s) Should I Use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
A Document release vs its related training. Which should come first? ISO 13485:2016 - Medical Device Quality Management Systems 18
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Master Samples - What should we be keeping? IATF 16949 - Automotive Quality Systems Standard 9
G Supplier delivered recent PPAP, should he deliver yearly layout inspection? IATF 16949 - Automotive Quality Systems Standard 4
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
J Should Loading and Unloading be Included in Cycle Times? Lean in Manufacturing and Service Industries 14
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33

Similar threads

Top Bottom