Should the Calibration Laboratory be kept locked?

C

Corsair

#1
Hello,
Recently, I went to visit a supplier for a QMS audit and found their internal Calibration Lab open and unlocked, The AS9100 do not mention that the Lab- needs to be locked at all times and the supplier internal procedure does not mentioned neither.
So is this a GMP or needs to be stated somewhere ?
You can't write them up if is not in a procedure or Standard.
Thanks in advance:thanx:
 
Elsmar Forum Sponsor

BradM

Staff member
Admin
#2
Re: Should the Calibration Lab be kept locked ?

What is the general nature of the lab? Do they need to garb to go in?

What type of calibrations are performed?

Are calibration standards sitting out where they could be adjusted? Are there fragile/ delicate/ expensive equipment that could be damaged?

Do they have a process where people are instructed to bring their instruments to the lab for calibration? Drop off and pick up?

did they have instruments/equipment sitting around with inappropriate labels on them?

I got this from an A2LA document:

5.3 S.3 There shall be provisions for restricting access to the field laboratory when unrestricted access could nvalidate the test or calibration results.
But again, it depends on how the lab is being used.

If you are going to suggest the lab be locked, there should be a compelling reason why. Locked doors/ rooms are a pain to work with. :)
 

Ron Rompen

Trusted Information Resource
#3
Re: Should the Calibration Lab be kept locked ?

I will echo Brad's comments. If there isn't a compelling reason to lock the lab door, then it shouldn't be required. Yes, there may be times when it IS appropriate, but in general I don't see the need.
I have worked in plants where EVERY access was locked, and it was a complete pain in the butt to try and get anything done.
 
C

Corsair

#4
Thanks for the feedback.
It is an aerospace medium size organization with capabilities for gauges, mic's, meters, etc-. I would have the Lab- locked just to safeguard the equipment and records.
I guess will be wise to go with the A2LA Std then.
 

drgnrider

Quite Involved in Discussions
#6
Adding to BradM's list -
.. is there a problem where people take items before they are ready to be returned for use?
 

dwperron

Trusted Information Resource
#7
This is something that is covered in ISO 17025:

5.8.1 The laboratory shall have procedures for the transportation, receipt, handling, protection, storage, retention and/or disposal of test and/or calibration items, including all provisions necessary to protect the integrity of the test or calibration item, and to protect the interests of the laboratory and the customer.

In a 17025 world it would be the responsibility of the lab to demonstrate their procedures as to how they would keep calibration items secure. It's just good practice to maintain security and the integrity of the items in the lab.
 
Thread starter Similar threads Forum Replies Date
Anerol C Where should I start as the person responsible for the calibration laboratory? General Measurement Device and Calibration Topics 7
A Preparing manual for Calibration laboratory! How should it be named? General Measurement Device and Calibration Topics 1
J What Calibration Standards should I have? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Damaged Load Cell After Calibration - Who Should Pay? General Measurement Device and Calibration Topics 3
Ajit Basrur Should calibration range of Oscilloscope cover the test ranges? General Measurement Device and Calibration Topics 8
E Should companies that perform calibration services be ISO 17025 accredited? ISO 17025 related Discussions 13
V Should there be a incident/deviation for breakdown or missed calibration date? Nonconformance and Corrective Action 7
M Should a low cost hand held refractometer be in a calibration program? Calibration Frequency (Interval) 8
B Should we perform Equipment final calibration-Operations closing ISO 13485:2016 - Medical Device Quality Management Systems 8
J CNC Calibration - What should be calibrated on a HAAS machine? General Measurement Device and Calibration Topics 7
K Should QS/TS certified internal test/calibration labs be required to be ISO17025 ISO 17025 related Discussions 16
J Should weigh count scales be in the Calibration System? QS-9000 - American Automotive Manufacturers Standard 4
J Calibration, should we.... Go-No-Go, thread, pin, and ring gauges QS-9000 - American Automotive Manufacturers Standard 2
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 2
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
S Should safety checks be included in the Control Plan? IATF 16949 - Automotive Quality Systems Standard 5
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
D Is there a specific location for PPE such as safety glass holders and glove dispensers should be mounted Occupational Health & Safety Management Standards 10
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M Who should receive the bills from suppliers and vendors, account payable or procurement? Consultants and Consulting 4
V IATF 16949 8.4.1 Control of externally provided processes, products and services - Should the CB be on our Approved Supplier List? IATF 16949 - Automotive Quality Systems Standard 10
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Companies that maintain your machine should be in ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Use of "Shall" versus "Should" in Procedures ISO 13485:2016 - Medical Device Quality Management Systems 21
D Class II medical device - When should a complaint be closed? Customer Complaints 6
Sidney Vianna IATF 16949 News Presentations from the latest IATF Stakeholder Event - Expectation that IATF 16949 certification should equate with product quality. Misguided? IATF 16949 - Automotive Quality Systems Standard 7
L Clause 0.4 of ISO 9001 and EHS - Where should I stop the inclusion of EHS in my QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Ed Panek Part 11 Self Certify Memo - What else should it cover? Qualification and Validation (including 21 CFR Part 11) 5
H Should I mention machine/Equipment password In SOP? Qualification and Validation (including 21 CFR Part 11) 4
D How long should we keep the spare parts available for our medical device, after we have stopped the production? ISO 13485:2016 - Medical Device Quality Management Systems 0
H Statistical Techniques Procedure - What should be included Document Control Systems, Procedures, Forms and Templates 4
Q How should I analyze measurement correlation between me and customer? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Sidney Vianna Interesting Discussion ISO 9001:2024 - What should be changed in the next Edition of ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 82
N Should it even be on the hazard analysis (software)? FMEA and Control Plans 2
V Which batches should or could be considered for design validation and design verification? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
L A Taiwan company want to sell Class I medical device (510(k) exempt) on Amazon, should we register with FDA? US Food and Drug Administration (FDA) 4
M Routine testing of medical electrical systems - What specific electrical safety tests should be performed? IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
G ISO 17025:2017 7.1.2 - Should I produce a document for the customer? ISO 17025 related Discussions 8

Similar threads

Top Bottom