Should the TC 176 Re-word the Requirements for Preventive Action?

Should the TC 176 have re-worded 8.5.3 to clarify the requirements for prev. action?

  • Yes. Most definitely 8.5.3 needs clarification.

    Votes: 13 72.2%
  • No. Preventive action requirements are very clear. No clarification is needed.

    Votes: 5 27.8%

  • Total voters
    18
J

JaneB

#21
I think too much energy is spent trying to decide what label to put on an action. Let's apply that energy to identifying areas where bad things could happen, and take action to prevent them. That is the greater part of the equation.
Yup, agree. I think that was pretty much what Jim suggested several posts back :)

I voted yes on the poll because yes, I think it could be clearer.
Yes, they are two different terms, and yes, I understand that they are different things.

But in practice, I do disagree that they are always 'crystal clear' and I sure as hell spend one heck of a lot of time attempting to educate clients a/about the difference and b/about meeting the different requirements.

I'd be very interested to see all those in favour of no change list their 5 crystal clear examples of preventive actions.
 
Elsmar Forum Sponsor

Helmut Jilling

Auditor / Consultant
#22
Yup, agree. I think that was pretty much what Jim suggested several posts back :)

I voted yes on the poll because yes, I think it could be clearer.
Yes, they are two different terms, and yes, I understand that they are different things.

But in practice, I do disagree that they are always 'crystal clear' and I sure as hell spend one heck of a lot of time attempting to educate clients a/about the difference and b/about meeting the different requirements.

I'd be very interested to see all those in favour of no change list their 5 crystal clear examples of preventive actions.

I do plan to spend time developing 5 examples today, per Sidney's challenge. However, let's recalibrate a couple of phrases here.

First, I agreed I would like them to reclarify their INTENT, in the next revision. On this, most of us agree.

My previous comment on "crystal clear" was actually

"Of course, as a spokesperson for the camp that says it means just what it says, why can't we just accept it as written? A proactive cousin to corrective action, after launch, before failure, following the same steps as corrective action? Alas, perhaps it is just my simple mind...:cool: ...but to me it is crystal clear.

I did not suggest that preventive action, as a concept, was "crystal clear" to everyone. Quite the contrary. I, too, spend much time explaining to clients what Preventive Actions are.

What I suggested was clear to me, was, that they wrote the details of cl 8.5.3 to exactly mirror cl 8.5.2. Thus it is "crystal clear" to me they did not intend to sweep FMEAs and Preventive Maintenance and all the other examples people have given into cl 8.5.3. The reason it is clear to me is those activities are indeed preventive activities, but they are covered in other specific areas. They are not even mentioned in cl 8.5.3. Thus, it is clear to me they did not INTEND those activies to be the answer to cl 8.5.3.

What I suggested was clear, was since they wrote cl 8.5.3 to exactly mirror cl 8.5.2, and since the official ISO definitions are IDENTICAL except for the use of the words "occurance" vs. "RE-occurance," thus it is "crystal clear" to me they intended preventive actions to be just like corrective actions, on a CA form, but proactive, not reactive. BEFORE the failure occurs. Other than that, they are the same thing. Two twin problem solvng tools, one reactive, one proactive. Nothing more, nothing less. THIS is what I suggested was "crystal clear" to me.

I asked, "why can't we just accept it as they wrote it?" It is a good tool. These other things are good activities as well, but they fit better in the other clauses I have listed in earlier posts.
 
Last edited:

Jim Wynne

Staff member
Admin
#23
I do plan to spend time developing 5 examples today, per Sidney's challenge. However, let's recalibrate a couple of phrases here.
You might find this more challenging than you think. For example, if I can't use the results of a nonconforming condition as impetus for PA, then aren't you excluding everything I might have learned in the past as a result of NC conditions? Therein lies the absurdity of the dichotomy. What is the time limit? If, five years ago, something blew up in my face and resulted in some egregious nonconforming condition, and today I'm faced with a similar situation, may I claim my actions are wholly preventive? The challenge, as I see it, is not to come up with examples of PA, it's to come up with examples of PA that were not the result of experience with bad things happening in the past.
 

Helmut Jilling

Auditor / Consultant
#24
You might find this more challenging than you think. For example, if I can't use the results of a nonconforming condition as impetus for PA, then aren't you excluding everything I might have learned in the past as a result of NC conditions? Therein lies the absurdity of the dichotomy. What is the time limit? If, five years ago, something blew up in my face and resulted in some egregious nonconforming condition, and today I'm faced with a similar situation, may I claim my actions are wholly preventive? The challenge, as I see it, is not to come up with examples of PA, it's to come up with examples of PA that were not the result of experience with bad things happening in the past.

I understand your point, but I would suggest you are overthinking it. That is why I coined the following explanation:

If the CA/PA form you commence to fill out is in REACTION to a failure event, then simply class it a Corrective Action and move on.

If the CA/PA form you commence to fill out is PROACTIVE to a POTENTIAL failure event, (no one compelled you to initiate it), then simply class it a Preventive Action and move on.

I literally do not spend more than a moment or two evaluating which label I place on it. We both agree, the label does not mean much ONCE WE HAVE COMMENCED the action. That is the main point I have been trying to make. I don't make any effort whatsoever to try to correlate it to some event in the past or a book I read or anything else.

Simple, natural, spontaneous. That is why I like this approach. My clients easily understand the concept.
 

Jim Wynne

Staff member
Admin
#25
I understand your point, but I would suggest you are overthinking it. That is why I coined the following explanation:

If the CA/PA form you commence to fill out is in REACTION to a failure event, then simply class it a Corrective Action and move on.

If the CA/PA form you commence to fill out is PROACTIVE to a POTENTIAL failure event, (no one compelled you to initiate it), then simply class it a Preventive Action and move on.
I am not overthinking anything. Practically everything we do wrt PA is reaction to something that happened in the past. How long do we have to wait before we can call the reaction PA? Unless you can answer that question and provide examples that are not reactive to past experience, you have no case.
 

Helmut Jilling

Auditor / Consultant
#26
I am not overthinking anything. Practically everything we do wrt PA is reaction to something that happened in the past. How long do we have to wait before we can call the reaction PA? Unless you can answer that question and provide examples that are not reactive to past experience, you have no case.
Jim, I don't know what you want. If you want a debate, you're not going to get it from me. I finished with the debate-for-sport aspect of this topic months ago.

Your previous post asked what sounded like a simple question, and I explained my simple approach. It works very nice, clean and simply. If that answer does not satisfy you, then nothing else I say on the topic will either. I refuse to allow this topic to be complicated, because my read of the definitions make the intent pretty darn clear to me.

I have already explained all the reasons for that viewpoint several times in previous posts in this and other threads.

If you want a "time," I would recommend you wait four months, three days, and 16.25 hours, then it can be preventive. I think that is a silly question and posture to take. A Proactive vs. Reactive position makes a lot more sense to me and explains simply why there would be any distinction between the two in the first place. I am interested in helping clients to get the results, not debate titles.

I have chosen to follow the concept that the ISO definitions present - that Preventives intend to be proactive, and Correctives focus on reactives, and both will prevent root causes.

I will not read any more into it than that. That works perfectly for me, and all the rest is idle debate without any end in sight.

You may choose to follow this path, or whichever other suits your fancy. But there is no benefit in parsing and dissecting the words. They say what they mean.:bigwave:
 
C

CarlDaniel

#27
Thank you guys for the very informative discussions…some maybe too deep though. Can I seek your advice on something? On our CA/PA form, we defined the corrective action as "To prevent recurrence of the problem due to the most probable cause." and preventive action as "To prevent occurrence of the problem due to other potential causes.". Is this an accurate definition? What if during the root cause analysis exercise we can't find any other potential cause that could lead to the same failure mode. Will it be acceptable to an auditor to write "implement the above corrective action and monitor effectiveness for several weeks"?
 

Helmut Jilling

Auditor / Consultant
#28
Thank you guys for the very informative discussions…some maybe too deep though. Can I seek your advice on something? On our CA/PA form, we defined the corrective action as "To prevent recurrence of the problem due to the most probable cause." and preventive action as "To prevent occurrence of the problem due to other potential causes.". Is this an accurate definition? What if during the root cause analysis exercise we can't find any other potential cause that could lead to the same failure mode. Will it be acceptable to an auditor to write "implement the above corrective action and monitor effectiveness for several weeks"?
From the point of view that I have been making, this would not work. The standard addresses preventive and corrective as two different activities. You use either one or the other. If you are observing a potential failure, then you execute a Preventive Action. If you observe an actual failure, you execute a Corrective Action. From that point, all the causes and actions stay on whichever form you began - preventive or corrective. There is little to be gained to filling out two forms.
 
J

JaneB

#30
Thank you guys for the very informative discussions…some maybe too deep though. Can I seek your advice on something? On our CA/PA form, we defined the corrective action as "To prevent recurrence of the problem due to the most probable cause." and preventive action as "To prevent occurrence of the problem due to other potential causes.". Is this an accurate definition? What if during the root cause analysis exercise we can't find any other potential cause that could lead to the same failure mode. Will it be acceptable to an auditor to write "implement the above corrective action and monitor effectiveness for several weeks"?
Carl,

I think you'd be better advised to stay with the ISO definitions of CA and PA. Making up your own could be unwise. I'd use the ISO 9000 (2006) definitions which I think have already been posted.

If you can't find any other potential cause - fine, stop. The Standard pushes you to consider the need and to act if you find cause. Don't go hunting about for ever. As for 'will it be acceptable to an auditor if we do x'... I just can't answer that. It's one of those 'pieces of string' questions.

Here's a sequence:
1. If a problem, correct it (if you can) - eg, rework. That's just correction (not CA)
2. If the problem could recur, then identify the cause/s and do CA to fix the causes & prevent recurrence. That is CA.
3. (Trickier) If you identify potential causes of potential NCFs, do PA to prevent them.

Sometimes CA does lead on to PA. Sometimes it doesn't. Sometimes PA comes about without CA, all by itself. Depends. Sorry - I KNOW that's not a 100% definite rule. It cannot be.

Re. 'implement the above corrective action and monitor effectiveness for several weeks" - the 'several weeks' monitoring depends on the thing at hand. How long do you need to monitor to ensure your CA has been effective? That's the real question. Might be a week, might be a month, might be several weeks. Don't just blindly pick 'several weeks'.
 
Thread starter Similar threads Forum Replies Date
A Should we assign the PRRC before the date of application of MDR (26 May 2021)? EU Medical Device Regulations 0
J UDI-DI how should we interpret Device version or model to determine if a new UDI-DI is needed? EU Medical Device Regulations 0
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance document to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
A Should I take an online course for a career in Occupational Health and Safety? Career and Occupation Discussions 2
J Should a Class 1 medical device with an option to measure body weight be considered Class 1m? EU Medical Device Regulations 0
K Should APQP/PPAP has its own section in a QM? Quality Management System (QMS) Manuals 1
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
P Should eIFU link per ISO 15223-1:2016 be added to labels out of scope of Reg 207/2012? EU Medical Device Regulations 1
S Which Sampling Plan(s) Should I Use? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 7
A Document release vs its related training. Which should come first? ISO 13485:2016 - Medical Device Quality Management Systems 18
S Which department should prepare the control plan? could you show me a standard regarding to this matter. FMEA and Control Plans 17
J Help settle a disagreement: Should external providers of preventive maintenance be on your ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
N Master Samples - What should we be keeping? IATF 16949 - Automotive Quality Systems Standard 9
G Supplier delivered recent PPAP, should he deliver yearly layout inspection? IATF 16949 - Automotive Quality Systems Standard 4
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
J Should Loading and Unloading be Included in Cycle Times? Lean in Manufacturing and Service Industries 14
E Manufacturers should develop a testing device for covid19 Service Industry Specific Topics 0
T 510(k) submission - Which name should I use in the submission? Other US Medical Device Regulations 3
N ISO 19011:2018 - 5.4.2 "...audit program should engage in appropriate continual development..." Training - Internal, External, Online and Distance Learning 4
G Should I perform Gage R&R only at the beginning of a new project? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
DuncanGibbons Should the requirements FAA/EASA Part 21 be addressed within the QMS and AS9100D quality manual? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M Should 510(k) Predicates be Actively Listed Devices? Other US Medical Device Regulations 12
B Why the Greek god Hephaestus should have done a design FMEA (DFMEA) on his giant robot APQP and PPAP 1
J On PFMEA for danger labels - Label always should be assigned severity 10 ? FMEA and Control Plans 3
H Who should be listed as the manufacturer/distributor on the box? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 15
M MDR, RED and LVD - Should our device comply with them? EU Medical Device Regulations 3
BeaBea How Many Processes should be created for each Department? Process Maps, Process Mapping and Turtle Diagrams 5
M Should volume of sales be factored into risk probability assessments? ISO 14971 - Medical Device Risk Management 33
MrTetris Should potential bugs be considered in software risk analysis? ISO 14971 - Medical Device Risk Management 5
S Should safety checks be included in the Control Plan? IATF 16949 - Automotive Quality Systems Standard 5
M Which incubation condition should be selected to recover both bacteria and fungus effectively Miscellaneous Environmental Standards and EMS Related Discussions 3
D Is there a specific location for PPE such as safety glass holders and glove dispensers should be mounted Occupational Health & Safety Management Standards 10
Robert Stanley Which Registrar Should I Choose for ISO 9001:2015 registration? Registrars and Notified Bodies 10
M Who should receive the bills from suppliers and vendors, account payable or procurement? Consultants and Consulting 4
V IATF 16949 8.4.1 Control of externally provided processes, products and services - Should the CB be on our Approved Supplier List? IATF 16949 - Automotive Quality Systems Standard 10
A We are ISO 13485:2016 should we be audited to ISO 14971 ISO 13485:2016 - Medical Device Quality Management Systems 16
E Received a Major finding during IATF Surveillance audit for loss of BIQS Level 3 (more than 6 SPPS in 6 months)...how should we address SYSTEMIC CA? IATF 16949 - Automotive Quality Systems Standard 11
J Organization merger. Should we keep two separate ISO 13485 certificates? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Companies that maintain your machine should be in ASL? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S Use of "Shall" versus "Should" in Procedures ISO 13485:2016 - Medical Device Quality Management Systems 26
D Class II medical device - When should a complaint be closed? Customer Complaints 6
Sidney Vianna IATF 16949 News Presentations from the latest IATF Stakeholder Event - Expectation that IATF 16949 certification should equate with product quality. Misguided? IATF 16949 - Automotive Quality Systems Standard 7
L Clause 0.4 of ISO 9001 and EHS - Where should I stop the inclusion of EHS in my QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Ed Panek Part 11 Self Certify Memo - What else should it cover? Qualification and Validation (including 21 CFR Part 11) 5
H Should I mention machine/Equipment password In SOP? Qualification and Validation (including 21 CFR Part 11) 4
D How long should we keep the spare parts available for our medical device, after we have stopped the production? ISO 13485:2016 - Medical Device Quality Management Systems 0
H Statistical Techniques Procedure - What should be included Document Control Systems, Procedures, Forms and Templates 4
Q How should I analyze measurement correlation between me and customer? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
Sidney Vianna Interesting Discussion ISO 9001:2024 - What should be changed in the next Edition of ISO 9001? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 114

Similar threads

Top Bottom