Single Fault Condition for Internally Powered Medical Device
rickbert
I'm trying to figure out which tests need to be done for leakage current for two battery powered deviced. One has a rechargable battery that is powered by an off the shelf DC adapter (60601 certified). The second uses a standard 9V battery. Both have electrodes that are applied parts. The test lab says that the worst case single fault condition is Mains voltage on a SIP/SOP. I see that for a the rechargable battery, but why do I need to apply 110V to an electrode plug for the 9V battery device? I need to put these justifications into the protocol, so I want to get them right and make them sound clean ("because the test house says so" isn't going to cut it).
Thanks for the help.
P
PaulGr
I think the testlab is right. If your device has a SIP/SOP, than it can be connected to other electrical equipment which might be defective leading to mains voltage on the SIP/SOP. I would think the leakage current measurements are then relevant for both adapter and internally powered models?
From your post it looks like your SIP/SOP is some 'electrode plug' could you clarify a little what this is? Perhaps it is not a SIP/SOP part at all.
Paul
rickbert
Looking at it more, I see what you mean. The SIP/SOP works for the rechargeble device. The second device has no Sip/Sop. It has a connection for electrode wires, but that's it. I was considering this connection a SOP, but I think that is incorrect. This device is a simple box with a 9V battery, a couple knobs, a port for electrodes, cables, and disposable electrodes. I guess I could apply mains voltage to the electrodes, but the whole device is floating so I don't see what that would do. Thanks for the help!
P
PaulGr
You are welcome. For further reading, there are some good topics in the cove related to IEC60601-1 and internally powered equipment, like this topic. (read the comments of Peter Selvey on testlabs with limited experience on battery powered devices)
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
G IEC 61010 - Single Fault Condition - Protective Impedance Implementation IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E Single Fault Condition Simulation IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
G What is the exact meaning of Single Fault condition? EU Medical Device Regulations 4
R Single Fault Condition - Motor Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 17
R Is foreseeable misuse considered as single fault condition? IEC 60601 - Medical Electrical Equipment Safety Standards Series 33
G Single Fault Condition for 3-Phase Medical System - IEC 60601 Clause 8.1b IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
M What are Single Fault Conditions and how to test - IEC 80601-2-13 Other Medical Device Related Standards 0
P Question related to Single Fault Conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
W Medical Device Single fault test-protection against fire IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M IEC 60601-1 Single fault conditions of electronic PCB components IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
N Is this a single integral drug device combination product EU MDR CE Marking (Conformité Européene) / CB Scheme 1
R CND nomenclature for single use instruments CE Marking (Conformité Européene) / CB Scheme 1
P Violation of CE mark - Re-use of single-use Products CE Marking (Conformité Européene) / CB Scheme 2
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N Where to find Regulations for Reprocessing and Reuse of Single-Use Devices Other Medical Device Related Standards 2
D Partial FAI - AS9102 - One single drawing has 10 part numbers AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
Sidney Vianna Informational IAF and ILAC Seek Contractor for Establishment of a Single International Organization for Accreditation ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 2
F UDI - Unit of Use and Class I, single-use devices EU Medical Device Regulations 4
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
D How to get EUDAMED Single Registration Number (SRN) EU Medical Device Regulations 17
CycleMike GD&T - Hole pattern - Print (attached) has a single Datum Reference Frame Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
K Does company who manufacture but does not design or carry out clinical trials is responsible for CE marking for single use medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 3
F 2017/745 Article 31 Single Registration Number Medical Device and FDA Regulations and Standards News 5
M Informational EU draft act – Single-use medical devices – safety and performance requirements for reprocessing Medical Device and FDA Regulations and Standards News 0
Q IATF rule for single site - Ingots from scrap metal recycling company IATF 16949 - Automotive Quality Systems Standard 0
R Supplier Controls we can place on Single-Source Suppliers ISO 13485:2016 - Medical Device Quality Management Systems 2
R Critical suppliers (Definition of) and MDSAP (Medical Device Single Audit Program) ISO 13485:2016 - Medical Device Quality Management Systems 15
Ed Panek Can a single supplier fit two or more categories for risk? ISO 13485:2016 - Medical Device Quality Management Systems 2
T No Defined Shelf Life/Expiration Date - Disposable single-use, non-invasive, non-sterile Other Medical Device Regulations World-Wide 2
F Reprocessing or refurbishing? Single Use Medical Device CE Marking (Conformité Européene) / CB Scheme 0
S How to make Single Sign On (SSO) Comply e-sig requirements? ISO 13485:2016 - Medical Device Quality Management Systems 4
E Sample size for design verification of variable in single use device Design and Development of Products and Processes 20
Ed Panek Sensitive Supplier Issue - Single source supplier ISO 13485:2016 - Medical Device Quality Management Systems 6
G Single DFU for multiple medical devices in one box Other Medical Device Regulations World-Wide 0
M Medical Device News Health Canada - Medical Device Single Audit Program (MDSAP) Transition Plan Canada Medical Device Regulations 2
A ISO 2859 Single Sampling - Clarification regarding the sampling table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
bio_subbu Indian government issues guidance on Grouping Medical Devices in a Single Submission Other Medical Device Regulations World-Wide 1
F Marketing a single medical device with multiple indications Other US Medical Device Regulations 4
T Definition Sole source VS. Single Source - Definitions Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 2
Edward Reesor MDSAP (Medical Device Single Audit Program) Costs Canada Medical Device Regulations 7
Mikey324 GR&R - Little to no part to part variation in single part number Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 14
R How many Corrective Action Request can be Issued by DCMA for a single issue? Nonconformance and Corrective Action 2
I Label Expiration Date - Single Use Combination Medical Device EU Medical Device Regulations 2
M Merge Technical File, DMR, and Device File into a single document? Other Medical Device and Orthopedic Related Topics 3
P Single Use Disposable Dental Syringe - CE Marking Advice EU Medical Device Regulations 4