Single use non-sterile syringe used in the oral cavity - Laboratory test advice

#1
We have a single use non-sterile syringe used in the oral cavity which we would like to place on the US and European market. The syringe is made of polypropylene (same material as regular syringes)

We are being advised by the lab that we should conduct the below tests to validate packaging:

AGING TEST
ASTM F1980


PACKAGING VALIDATION TEST
ASTM F1140 BURST AND CREEP TEST
ASTM F1929 DYE PENETRATION
ASTM F88 SEAL STRENGTH TEST
ASTM F1866 VISUAL INSPECTION (at 0-YEAR/3-YEAR ACCELERATED AGING)

The quote they have provided is, in my opinion, not cheap. I would like advice as to whether we need to have all these tests done given the product in non-sterile (in particular ASTM F1980 for 510(k) exempt and also CE class 1(non-sterile and non measuring) under the new MDR requirements.

Having looked at the ASTM F1980, I do not think it applies to our product as it is not sterile. However, the lab argue that it needs to be done to claim shelf life. They also claim that the tests need to be done to meet the below MDR requirements

Clause 7
Devices shall be designed, manufactured and packaged in such a way that their characteristics and performance during their intended use are not adversely affected during transport and storage, for example, through fluctuations of temperature and humidity, taking account of the instructions and information provided by the manufacturer.

Clause 10.2
Devices shall be designed, manufactured and packaged in such a way as to minimise the risk posed by contaminants and residues to patients, taking account of the intended purpose of the device, and to the persons involved in the transport, storage and use of the devices. Particular attention shall be paid to tissues exposed to those contaminants and residues and to the duration and frequency of exposure.

I would appreciate it if you could advice me as I am unclear of the requirements and feel the lab is pushing to get business rather meet the necessary regulatory requirements.

Do I really need the above listed tests for our product for FDA and CE?
 
Elsmar Forum Sponsor

planB

Super Moderator
#2
Ahmed,

the listed ASTM standards are typically used to demonstrate packaging integrity of _sterile_ barrier systems according to ISO 11607-1. You can use these tests to not only demonstrate that your packaging is capable of maintaining sterility of your packaged device over the labelled shelf life and during distribution, but more generally any defined microbial state that you have specified.

I assume that you have actually defined a microbial state of your device? If yes, this explains why your lab is thinking for tests into the direction of sterile barrier systems. The proposed tests aim at the following: demonstrating both seal integrity and whole-package integrity, the two aspects you typically address during your package stability & performance validation.

Bottom line: if you have a defined microbial state then the proposed tests may be reasonable. If you do not have to maintain a defined microbial state then these tests are beyond scope.

HTH,

Gerhard
 

Ronen E

Problem Solver
Moderator
#3
We have a single use non-sterile syringe used in the oral cavity which we would like to place on the US and European market. The syringe is made of polypropylene (same material as regular syringes)

We are being advised by the lab that we should conduct the below tests to validate packaging:

AGING TEST
ASTM F1980


PACKAGING VALIDATION TEST
ASTM F1140 BURST AND CREEP TEST
ASTM F1929 DYE PENETRATION
ASTM F88 SEAL STRENGTH TEST
ASTM F1866 VISUAL INSPECTION (at 0-YEAR/3-YEAR ACCELERATED AGING)

The quote they have provided is, in my opinion, not cheap. I would like advice as to whether we need to have all these tests done given the product in non-sterile (in particular ASTM F1980 for 510(k) exempt and also CE class 1(non-sterile and non measuring) under the new MDR requirements.

Having looked at the ASTM F1980, I do not think it applies to our product as it is not sterile. However, the lab argue that it needs to be done to claim shelf life. They also claim that the tests need to be done to meet the below MDR requirements

Clause 7
Devices shall be designed, manufactured and packaged in such a way that their characteristics and performance during their intended use are not adversely affected during transport and storage, for example, through fluctuations of temperature and humidity, taking account of the instructions and information provided by the manufacturer.

Clause 10.2
Devices shall be designed, manufactured and packaged in such a way as to minimise the risk posed by contaminants and residues to patients, taking account of the intended purpose of the device, and to the persons involved in the transport, storage and use of the devices. Particular attention shall be paid to tissues exposed to those contaminants and residues and to the duration and frequency of exposure.

I would appreciate it if you could advice me as I am unclear of the requirements and feel the lab is pushing to get business rather meet the necessary regulatory requirements.

Do I really need the above listed tests for our product for FDA and CE?
It seems that the lab is either trying to get business or just running auto-pilot (being lazy?). Some of the tests *might* be relevant but it's not obvious and it'd be their onus to explain why (if they care enough about your business).

You do need to address both shelf life and packaging suitability/integrity, but in a more general way and with a little more analysis as to what is actually required.

For general packaging validation and testing maybe have a look here.
 
#4
Ahmed,

the listed ASTM standards are typically used to demonstrate packaging integrity of _sterile_ barrier systems according to ISO 11607-1. You can use these tests to not only demonstrate that your packaging is capable of maintaining sterility of your packaged device over the labelled shelf life and during distribution, but more generally any defined microbial state that you have specified.

I assume that you have actually defined a microbial state of your device? If yes, this explains why your lab is thinking for tests into the direction of sterile barrier systems. The proposed tests aim at the following: demonstrating both seal integrity and whole-package integrity, the two aspects you typically address during your package stability & performance validation.

Bottom line: if you have a defined microbial state then the proposed tests may be reasonable. If you do not have to maintain a defined microbial state then these tests are beyond scope.

HTH,

Gerhard
Thank you Gerhard and Ronen.

We have not defined any microbial state. The lab insists that we need to do these tests to meet the new MDR. I would have said fine if the costs were not as extortionate.

Do we need to specify a shelf life for a medical device under MDR and also are there any general standards that we can use to demonstrate we meet the packing validation?
 

planB

Super Moderator
#5
Ahmed,

if you do not have a defined microbial state then you would only need packaging that protects your device from excessive soiling during distribution. A transport simulation of your device in its distribution packaging configuration, e.g. according to ISTA 2A (follow Ronen's link), could be sufficient.

You only need to specify a shelf life if your device (including its barrier system) is prone to aging - if not, nothing to validate here.

You mentioned that your device is used in the oral cavity: does your device require disinfection prior to use?

HTH,
 
#6
Thanks for your further advice Gerhard.

We do not reccomend disinfection as we manufacture and package the syringes in cleanrooms. We were thinking of introducing a process to randomly select samples for bioburden testing to ensure the packed product meets a certain standards. Not sure if we need this so not implemented.

Do you think syringes made of polypropylene require an aging test?
 

planB

Super Moderator
#7
Ahmed,

I presume you manufacture in a clean-room, because your device does indeed have cleanliness requirements regarding viable and non-viable contamination, right? And you want to maintain this state until the point of use? If yes to both questions, then your lab's proposal does not appear that far-fetched anymore.

If your packaging has to maintain this state, then your packaging is the driver for performing aging studies, typically not your device: (Pure) Polypropylene is typically stable when not exposed to environmental stress, such as (UV) radiation and extreme temperatures and humidity over extended periods of time.

HTH,
 

Ronen E

Problem Solver
Moderator
#8
Ahmed,

I presume you manufacture in a clean-room, because your device does indeed have cleanliness requirements regarding viable and non-viable contamination, right? And you want to maintain this state until the point of use? If yes to both questions, then your lab's proposal does not appear that far-fetched anymore.

If your packaging has to maintain this state, then your packaging is the driver for performing aging studies, typically not your device: (Pure) Polypropylene is typically stable when not exposed to environmental stress, such as (UV) radiation and extreme temperatures and humidity over extended periods of time.

HTH,
1. As much as I currently understand the application and the design, it's a syringe loaded with a prefilled cartridge, so the syringe itself might not have high cleanliness requirements. I don't think any of us here can give responsible, specific advice (yet), because we currently know very little about the device.

2. Where several components are assembled together, substantiating a shelf life is likely to involve more than individual materials stability under a given set of environmental conditions. Factors like creep, stress cracking, compression set etc. may become relevant, depending on the various materials used and the geometrical design. Phenomena like spontaneous (unintended) adhesion, difusion, leaching, absorption, modification of boundary layers etc. should also be considered when components are in close contact or pressed together for prolonged periods. So again, I wouldn't say anything too definite without getting to know the design really well first.
 
Thread starter Similar threads Forum Replies Date
T No Defined Shelf Life/Expiration Date - Disposable single-use, non-invasive, non-sterile Other Medical Device Regulations World-Wide 2
A Cleaning instructions for our non sterile, single use products EU Medical Device Regulations 6
P ETO Sterilisation of Single Use Part supplied in Non-Sterile State EU Medical Device Regulations 12
A Single Multifunction device CE approval - Medical and non-medical EU Medical Device Regulations 4
dgrainger Informational Consultation: Medical devices – single identifier for similar highly individualised devices EU Medical Device Regulations 0
E Process capability in a single batch Capability, Accuracy and Stability - Processes, Machines, etc. 3
S Cleaning validation of single-use tips providing treatment of women’s intimate areas Manufacturing and Related Processes 3
B Single fault analysis and negligible probability of IC internal fault IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
I MOPP required for SIP when considering Single Fault Condition (SFC) in 60601-1 IEC 60601 - Medical Electrical Equipment Safety Standards Series 16
J Critical Supplier Initial evaluation for single purchases Oil and Gas Industry Standards and Regulations 16
B Labeling question about different manufacturers on a single label Manufacturing and Related Processes 4
J Single Fault Testing for Secondary Circuits (60601-1 section 13.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
K ISO 2859-5 (single run production) control plan issues FMEA and Control Plans 2
J Can you assign two EMDN codes to a single device for EUDAMED? EU Medical Device Regulations 3
A Single-Use accessory disinfected prior to use Other Medical Device Related Standards 6
L Single-cell lithium ion battery regulations IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
R Labeling requirements for single use devices Manufacturing and Related Processes 4
renenatasha Single Registration Number (SRN) EU Medical Device Regulations 2
E Sources of every single paragraph of the ISO 9001 standards? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
K Lithium Ion - Single Fault Safe - Deep Discharge IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
N Is this a single integral drug device combination product EU MDR CE Marking (Conformité Européene) / CB Scheme 1
R CND nomenclature for single use instruments CE Marking (Conformité Européene) / CB Scheme 1
P Violation of CE mark - Re-use of single-use Products CE Marking (Conformité Européene) / CB Scheme 2
M What are the basics of Medical Device Single Audit Program (MDSAP)? ISO 13485:2016 - Medical Device Quality Management Systems 7
D SINGLE FAULT CONDITION, short circuit and open circuit of any component (IEC 60601-1 3.1) IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
N Where to find Regulations for Reprocessing and Reuse of Single-Use Devices Other Medical Device Related Standards 2
T Single Fault Condition IEC 60601 Clause 8.7.1 shorting Cr/Cl in Patient Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
D Partial FAI - AS9102 - One single drawing has 10 part numbers AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
Sidney Vianna Informational IAF and ILAC Seek Contractor for Establishment of a Single International Organization for Accreditation ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 2
F UDI - Unit of Use and Class I, single-use devices EU Medical Device Regulations 5
D How to get EUDAMED Single Registration Number (SRN) EU Medical Device Regulations 19
CycleMike GD&T - Hole pattern - Print (attached) has a single Datum Reference Frame Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
K Does company who manufacture but does not design or carry out clinical trials is responsible for CE marking for single use medical devices? ISO 13485:2016 - Medical Device Quality Management Systems 4
F 2017/745 Article 31 Single Registration Number Medical Device and FDA Regulations and Standards News 5
M Informational EU draft act – Single-use medical devices – safety and performance requirements for reprocessing Medical Device and FDA Regulations and Standards News 0
Q IATF rule for single site - Ingots from scrap metal recycling company IATF 16949 - Automotive Quality Systems Standard 0
R Supplier Controls we can place on Single-Source Suppliers ISO 13485:2016 - Medical Device Quality Management Systems 2
R Critical suppliers (Definition of) and MDSAP (Medical Device Single Audit Program) ISO 13485:2016 - Medical Device Quality Management Systems 22
A Touch current in single fault conditions test and earth leakage current in normal conditions test, are they really different tests? IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
Ed Panek Can a single supplier fit two or more categories for risk? ISO 13485:2016 - Medical Device Quality Management Systems 2
F Reprocessing or refurbishing? Single Use Medical Device CE Marking (Conformité Européene) / CB Scheme 0
S How to make Single Sign On (SSO) Comply e-sig requirements? ISO 13485:2016 - Medical Device Quality Management Systems 4
E Sample size for design verification of variable in single use device Design and Development of Products and Processes 20
Ed Panek Sensitive Supplier Issue - Single source supplier ISO 13485:2016 - Medical Device Quality Management Systems 6
G Single DFU for multiple medical devices in one box Other Medical Device Regulations World-Wide 0
M Medical Device News Health Canada - Medical Device Single Audit Program (MDSAP) Transition Plan Canada Medical Device Regulations 2
A ISO 2859 Single Sampling - Clarification regarding the sampling table Inspection, Prints (Drawings), Testing, Sampling and Related Topics 4
bio_subbu Indian government issues guidance on Grouping Medical Devices in a Single Submission Other Medical Device Regulations World-Wide 1

Similar threads

Top Bottom