We have a single use non-sterile syringe used in the oral cavity which we would like to place on the US and European market. The syringe is made of polypropylene (same material as regular syringes)
We are being advised by the lab that we should conduct the below tests to validate packaging:
AGING TEST
ASTM F1980
PACKAGING VALIDATION TEST
ASTM F1140 BURST AND CREEP TEST
ASTM F1929 DYE PENETRATION
ASTM F88 SEAL STRENGTH TEST
ASTM F1866 VISUAL INSPECTION (at 0-YEAR/3-YEAR ACCELERATED AGING)
The quote they have provided is, in my opinion, not cheap. I would like advice as to whether we need to have all these tests done given the product in non-sterile (in particular ASTM F1980 for 510(k) exempt and also CE class 1(non-sterile and non measuring) under the new MDR requirements.
Having looked at the ASTM F1980, I do not think it applies to our product as it is not sterile. However, the lab argue that it needs to be done to claim shelf life. They also claim that the tests need to be done to meet the below MDR requirements
Clause 7
Devices shall be designed, manufactured and packaged in such a way that their characteristics and performance during their intended use are not adversely affected during transport and storage, for example, through fluctuations of temperature and humidity, taking account of the instructions and information provided by the manufacturer.
Clause 10.2
Devices shall be designed, manufactured and packaged in such a way as to minimise the risk posed by contaminants and residues to patients, taking account of the intended purpose of the device, and to the persons involved in the transport, storage and use of the devices. Particular attention shall be paid to tissues exposed to those contaminants and residues and to the duration and frequency of exposure.
I would appreciate it if you could advice me as I am unclear of the requirements and feel the lab is pushing to get business rather meet the necessary regulatory requirements.
Do I really need the above listed tests for our product for FDA and CE?
We are being advised by the lab that we should conduct the below tests to validate packaging:
AGING TEST
ASTM F1980
PACKAGING VALIDATION TEST
ASTM F1140 BURST AND CREEP TEST
ASTM F1929 DYE PENETRATION
ASTM F88 SEAL STRENGTH TEST
ASTM F1866 VISUAL INSPECTION (at 0-YEAR/3-YEAR ACCELERATED AGING)
The quote they have provided is, in my opinion, not cheap. I would like advice as to whether we need to have all these tests done given the product in non-sterile (in particular ASTM F1980 for 510(k) exempt and also CE class 1(non-sterile and non measuring) under the new MDR requirements.
Having looked at the ASTM F1980, I do not think it applies to our product as it is not sterile. However, the lab argue that it needs to be done to claim shelf life. They also claim that the tests need to be done to meet the below MDR requirements
Clause 7
Devices shall be designed, manufactured and packaged in such a way that their characteristics and performance during their intended use are not adversely affected during transport and storage, for example, through fluctuations of temperature and humidity, taking account of the instructions and information provided by the manufacturer.
Clause 10.2
Devices shall be designed, manufactured and packaged in such a way as to minimise the risk posed by contaminants and residues to patients, taking account of the intended purpose of the device, and to the persons involved in the transport, storage and use of the devices. Particular attention shall be paid to tissues exposed to those contaminants and residues and to the duration and frequency of exposure.
I would appreciate it if you could advice me as I am unclear of the requirements and feel the lab is pushing to get business rather meet the necessary regulatory requirements.
Do I really need the above listed tests for our product for FDA and CE?