I have some questions concerning the leakage tests that require applying mains voltage to SIP/SOP ports. I have read in other posts that if you specify in the manual that SIP/SOP can only be connected to other medical devices is no longer necessary to apply mains voltage to SIP/SOP.
My question concerns the Ethernet port, if it is intended to be connected to a network is it necessary to apply mains voltage?.
Another question I have is about MOPP in SIP/SOP, my understanding is that if the SIP/SOP are to be connected to other medical devices is not necessary to provide 2 MOPP to the SIP/SOP, that would be only necessary when SIP/SOP are designed to be connected to non-medical devices.
In case that the Ethernet port would need 2 MOPP would it be possible to specify that a external network isolator is needed when the Ethernet is to be connected to a network (instead of providing it internally).
My last question is how, (when it is necessary) mains voltage is applied to the SIP/SOP. (do all the SIP/SOP lines need to be shorted and connected to mains voltage?, and if there are more than one SIP/SOP, do all the SIP/SOP need to be connected to mains voltage at the same time?)
Thanks in advance
My question concerns the Ethernet port, if it is intended to be connected to a network is it necessary to apply mains voltage?.
Another question I have is about MOPP in SIP/SOP, my understanding is that if the SIP/SOP are to be connected to other medical devices is not necessary to provide 2 MOPP to the SIP/SOP, that would be only necessary when SIP/SOP are designed to be connected to non-medical devices.
In case that the Ethernet port would need 2 MOPP would it be possible to specify that a external network isolator is needed when the Ethernet is to be connected to a network (instead of providing it internally).
My last question is how, (when it is necessary) mains voltage is applied to the SIP/SOP. (do all the SIP/SOP lines need to be shorted and connected to mains voltage?, and if there are more than one SIP/SOP, do all the SIP/SOP need to be connected to mains voltage at the same time?)
Thanks in advance