Soak Time to Acclimate prior to Calibration

S

svalego

I'm looking for a requirement which would state how long a piece of equipment (mechanical, electronic, temperature, etc.) needs to acclimate once it arrives from shipment to a calibration lab prior to the unit being serviced. I know this would vary depending on the equipment but did not know if anyone knew of a Standard which would state the time needed to soak prior. Any help would be appreciated.
 
N

ncwalker

There is no standard that I know of. You are correct in that it will vary from piece of equipment to piece of equipment. What has to happen is the components that comprise the pickups need to reach standard temperature. (Or close to it).

For example: On a CMM the thing that is doing the measuring is usually a glass piece with graduations etched on it. This piece of glass would need to come to standard temperature to take all the variation due to temperature (expansion/contraction) out of the equation. So you could check this by using a pyrometer and shooting the piece. Which, of course, is NOT readily available for direct viewing always. (One would think they would wire this in to the CMM, I mean, thermocouples are relatively cheap). This could happen quickly or take quite a long time depending on the object. It's a heat transfer question.

You could test it by measuring something AT STP in the lab and looking at the error. Example: If I have a 1" gage block in the lab and I know it's at STP because it has been in the lab for weeks, and I measure this and get 1.001", it could be the thing I am measuring with hasn't soaked long enough. It could also be error in the device. :)

In truth, because the heat transfer has an asymptotic boundary, it will take an infinite amount of time for the object to reach the "same" temperature as the lab. So the reality is you only have to wait until the error is "good enough." That depends on what you are measuring. Are you making 2x4s or bearings?
 
Thread starter Similar threads Forum Replies Date
M AMS2770 Aluminum Heat Treat - Determining Soak Time Various Other Specifications, Standards, and related Requirements 1
S AMS2771 - Soak Time Interpretation (Thermal Conditioning) - T6 or T61? Manufacturing and Related Processes 8
J Determination of software safety class (62304) prior to software risk analysis ISO 14971 - Medical Device Risk Management 3
CaliperJim Acclimation time prior to calibration General Measurement Device and Calibration Topics 8
D How to address the content deviation of 'cannot apply criteria of risk acceptability prior to...' ISO 14971 - Medical Device Risk Management 1
C Determining an as received OOT condition during calibration does not affect measurements taken prior to calibration. General Measurement Device and Calibration Topics 16
A Single-Use accessory disinfected prior to use Other Medical Device Related Standards 6
R Disruptions that happen prior to surveillance audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
B Is 14971 Annex C checklist now in ISO/TR 24971 required to complete prior to 510k filing? ISO 14971 - Medical Device Risk Management 3
M How to apply IEC 62304 Problem Resolution Process for bugs discovered prior to release i.e. during initial development? IEC 62304 - Medical Device Software Life Cycle Processes 2
blackholequasar ISO 13485 certification prior to Medical Device Manufacturing... worth it? ISO 13485:2016 - Medical Device Quality Management Systems 4
S Inventory prior to AS9120 certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
G ISO 13485 Certification - Can we get the ISO 13485 certification prior to shipment of the device? ISO 13485:2016 - Medical Device Quality Management Systems 6
J Quality Objective for QMS prior to Certification AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 10
P Writing a presubmission to the FDA prior to the De novo submission US Food and Drug Administration (FDA) 4
A Is it mandatory to obtain ISO13485 certificate prior to CE marking? EU Medical Device Regulations 8
V Supplier wants to perform pre-qualifications prior to every run instead of locking in parameters. Supplier Quality Assurance and other Supplier Issues 7
L Unacceptable risk for basic safety prior to or after mitigation? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
JoshuaFroud CE Marking and Brexit - Product shipped to a distributor prior to the March deadline EU Medical Device Regulations 2
S Adding Value prior to signing EASA Form 1 EASA and JAA Aviation Standards and Requirements 5
S IMS Manual Review and Revision Prior to Audit ISO 14001:2015 Specific Discussions 4
A ISO 9001:2015 registrar Auditor requesting copies of procedure prior to audit ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 24
Q Top Management Change Prior to Certification ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 12
H Do you have to perform an full internal prior to IATF 16949 audit? IATF 16949 - Automotive Quality Systems Standard 21
K Markets with No Prior Approvals? Without Country of Origin or other Prior Approval Other Medical Device Regulations World-Wide 5
R AS9100 FAI - First Article Inspection to a prior Customers PO AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
A Internal Audit - Findings - Recent Internet Audit (Prior to Certification) IEC 27001 - Information Security Management Systems (ISMS) 7
O Process Mapping prior to Validation and Software to use to Map Process Maps, Process Mapping and Turtle Diagrams 12
JoCam Choose Medical Device Distributors prior to Device Registration? Other Medical Device Regulations World-Wide 6
E Is Design Transfer to Production be complete prior to Release of first production lot 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
Q ISO 9001:2008 requirement for Design Organization Approval prior to "Repair" ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 10
Q AS9100 Repairs - Approval from a non-aerospace customer prior to conducting a repair AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 14
L Who can perform ISO 9001:2008 Internal Audit prior to our Registration Audit? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
C How to Clean Gage Blocks and Gage Pins prior to Calibration General Measurement Device and Calibration Topics 4
L How to add a New Factory under a Prior 510K Establishment US Food and Drug Administration (FDA) 2
D Question... Record retention Prior to FDA NDA application and approval Records and Data - Quality, Legal and Other Evidence 3
L Regarding CDRH Cover Sheet and prior related Submissions Other Medical Device and Orthopedic Related Topics 3
P Should I send survey questionnaire prior to Supplier EMS Audit? Miscellaneous Environmental Standards and EMS Related Discussions 14
C Re-Verification of Nonconforming Product prior to closing our DMR AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 9
C New decree in France: Authorization required prior advertising Medical Devices EU Medical Device Regulations 1
J SFDA Labeling Requirements prior to import to any Chinese territory China Medical Device Regulations 2
S Charging Consulting Fees prior to 510(k) Approval Other US Medical Device Regulations 3
Q Company acquisition prior to QMS and ISO Audits ISO 13485:2016 - Medical Device Quality Management Systems 1
Roberticus Responsibility for Outsourced Process, ALL in-house Mfg/QC done (prior FG PN) ISO 13485:2016 - Medical Device Quality Management Systems 5
M Internal Audit of our ISO 9001 Quality Systems prior to the Registration Audit Internal Auditing 9
J Product Designed prior to Implementation of AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
L Sigma & the Distributor - Rejected pieces from prior shipments to current measures Six Sigma 7
L Method for Poka-Yoke (Mistake Proofing) Validation prior to SOP APQP and PPAP 6
C How do you Validate Solderability of Components Prior to Soldering Operations Manufacturing and Related Processes 5

Similar threads

Top Bottom