I was wondering in general how rigorously standard testing must be followed for CE marking. Can we make a reasonable argument for changing the standard tests in some cases?
For example our CPAP utilizes a medical hose to source oxygen to the oxygen mixer. We are performing testing in accordance with ISO 5359. I was wondering if we have to follow the standard exactly even if it is far out of the device specs, such as clause 4.4.14 pressure of the ISO 5369. Our device doesn't have a flowrate of 200L/m. At max it might be 30 L/m for our device. Do we still need to perform a pressure drop test at 200L/m, when the device won't get anywhere near this high?
Marcelo
Re: Softening Standard Requirements
Is the 200L/m in normal or fault condition? If it?s in fault condition, it does not care your device normally would be at 30.
Anyway, a general answer to your question is: from a risk management standpoint, requirements in a safety product standard are possible risk control measures to reduce risk (and testing is simply the verification of implementation and effectiveness of the risk control measure). Possible, not the best or even appropriate to any given device. So yes, you might change the requirements to better "fit" your device situation.
However, generally people are still too focused on blindly following standards, so depending on your assessor, you might have some problems having him accept that (but if you have a good assessor he will verify your rationale and if if it?s sound he will accept the deviation).
F
flyingfrisbee
Thank you for your help. :thanx:
The standard doesn't mention single fault conditions in this particular test. Also, I don't think that our device could reach that high of a flow rate due to the valves and regulators.
I am guessing that I will have to do testing to confirm that it won't reach that pressure even in a single fault, so I can justify my claim of not needing to test for the pressure drop at a 200 L/m flow rate.
Marcelo
Well, I don?t have the last version of ISO 5369 here, but the old versions do mention that the 200L/m pressure drop requirements is related to some lung ventilators that may require peak flows of 200L/m for up to 3 s (please verify in the rationale of the last version if it has this explanation). So it seems that maybe it?s not related to a fault condition.
If your CPAP does not work like this, that would be another good reason not the use the test as defined (you may than justify that your peak flow requirement is xxx, so you changed the testing to yyy).
F
flyingfrisbee
The standard I have mentions lung ventilators as well including the 200L/m for 3s. We are not dealing with ventilators or high velocity flows.
I will do what you suggest for the testing. Thanks for your help.
K Heat Treatment (Softening) on Cast Iron before it can be drilled to form Thread? Manufacturing and Related Processes 4
J Softening Silver Plated Copper Manufacturing and Related Processes 6
SABRINA DJILALI SAIAH Implementation of the ISAS-MEDIA 9001 standard in a TV channel ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 7
Sidney Vianna Interesting Discussion Should ISO 9004 be changed from a guidance standard to a requirements standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 5
B European Harmonized vs IEC standard IEC 62366 - Medical Device Usability Engineering 2
D Which ISO Standard to purchase? ISO 13485:2016 - Medical Device Quality Management Systems 7
C Standard -Machinist Handbook Customer Complaints 4
N EN 813, EN 12277, EN 1497 - Testing some harness prototypes to an EN standard Various Other Specifications, Standards, and related Requirements 0
R Remote Audits for ISO 9001 (or any other standard) General Auditing Discussions 31
S Shore A hardness Tester Acceptance Criteria Standard Elsmar Cove Forum Suggestions, Complaints, Problems and Bug Reports 2
Sidney Vianna What ISO Standard (under the TC 176) supports the UN Sustainable Development Goal #10? ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 11
John Broomfield Vote - Should ISO9004 Become a Requirements Standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 11
Q EN ISO 13485:2016/AC:2018 - AC:2018 being stated in the applicable harmonized standard listing Other ISO and International Standards and European Regulations 1
J SAE J1739 standard IATF 16949 - Automotive Quality Systems Standard 2
Marcel DS LV 123 - BMW, AUDI, Mercedes Standard VDA Standards - Germany's Automotive Standards 1
F Standard Calibration Procedures: Recommended Practice ISO 17025 related Discussions 0
M FDA News FDA Releases Draft Guidance Clarifying Application of ISO 10993-1 Biocompatibility Standard Medical Device and FDA Regulations and Standards News 0
Y UL 1998 Standard: software classes Software Quality Assurance 0
G What to do with correction factors overlooked by an accredited standard (6.4.11)? ISO 17025 related Discussions 5
G Copy of withdrawn ISO 9001:1994 Quality Management Standard ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
Anonymous16-2 SOP (Standard Operating Procedure) Numbering Document Control Systems, Procedures, Forms and Templates 7
Ed Panek Compliance with Standards? When a standard is updated/revised CE Marking (Conformité Européene) / CB Scheme 3
Skooma Deburring and Tolerances - Is there a Standard for deburring or removing material? Manufacturing and Related Processes 3
B Purchasing AS9100D Standard in Japanese AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
O In addition to the standard, what other ISO 13485 sources do people recommend? ISO 13485:2016 - Medical Device Quality Management Systems 5
B ASTM E18-2020 - Rockwell testing standard changes? General Measurement Device and Calibration Topics 2
M Informational RoHS Standard - EN 63000:2018 to replace EN 50581:2012 Medical Device and FDA Regulations and Standards News 11
D Recent changes to ISO 14971 - SOP required for managing standard revisions ISO 13485:2016 - Medical Device Quality Management Systems 1
P Cenelec updated standard information CE Marking (Conformité Européene) / CB Scheme 1
S Is there a standard for sample size during R&D phase Other ISO and International Standards and European Regulations 18
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
K Question on whether IEC 60601-2-62 standard is applied IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
M Viable count Measurement ISO Standard Other Medical Device Related Standards 1
I When exactly was "competence" added to the ISO-9001 standard? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 9
S Managing specification changes on standard parts purchased through trading companies Document Control Systems, Procedures, Forms and Templates 2
MDD_QNA QR Code Standard ISO/IEC 15417:2007 - Does anyone use it? Other Medical Device Related Standards 3
S Average and standard deviation of Cumulative Data Statistical Analysis Tools, Techniques and SPC 5
R Standard Filter for Calibration Colorimeter ISO 17025 related Discussions 1
B Standard for Dielectric Strength Test IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
M Specific Absorption Rate (SAR) Testing - What Standard to Use? Other ISO and International Standards and European Regulations 1