Thanks DanMann
Yes, the Software Documentation level determination (Basic vs Enhanced) is done assuming 100% probability of all Sw failures modes (but the Sw
FMEA uses actual estimated risks).
But the FDA guidance does not state explicitly that you should assume hardware mitigations should be assumed to be 100% ineffective for this documentation estimation - and it would not make any sense, in my view.
Even if you do estimate prior to bolted-on hardware mitigations for Sw failures, I don't see why or how you can estimate prior to intrinsically safe hardware design.
The device is a diagnostic cardiac stimulator Class IIb - where for above example, Sw failure could try to program higher than intended current, but hardware configuration limits it.