I'm trying to tease out the difference between software validation and design V&V, when the medical device is a SaMD. Any insights?

- Once you validate the software, what is left to do for design V&V?
- Is validation of the software part of design verification? Design validation? Neither one?
 
Elsmar Forum Sponsor

chris1price

Trusted Information Resource
#2
I would say software validation is equivalent to Design Verification, ie ensuring the correct code has been written. However you still need Design Validation, so Human Factors studies and making sure the code does what was intended in the first place.
 

yodon

Staff member
Super Moderator
#3
Unfortunately, guidance docs, etc. do a pretty poor job of describing expectations for "software validation" and, frankly, 62304 doesn't help since they introduce even a 3rd term, "software system testing" (and explicitly excludes validation). As @chris1price points out, you'll still confirm requirements are met; i.e., verification and still have to do (system) validation. Some software requirements are often necessary to test at a lower level (debug versions).

FDA has a Guidance doc on software validation which is probably your best resource. If you read it, you'll see they consider software validation more than testing, it's more of a well-defined software lifecycle (processes) that covers the life of the system.
 

Watchcat

Trusted Information Resource
#6
I would say software validation is equivalent to Design Verification
I've always said that, but to medical device people. When you are dealing with a startup that involves only software people, no experience with medical devices...I have found it challenging to explain.
 

Watchcat

Trusted Information Resource
#7
they consider software validation more than testing, it's more of a well-defined software lifecycle (processes) that covers the life of the system.
Yes, more familiar with the guidance. But I don't think it addresses the difference between software validation and design validation for a medical device. Possibly a consequence of the mutually agreed upon fantasy that software stands alone.

On the other hand, I did have a breakthrough on that part, explaining to a software guru that all software is ultimately "embedded," including SaMDs, and he seemed to agree.
 

Ed Panek

QA RA Small Med Dev Company
Staff member
Super Moderator
#8
Do you have any past audit reports for this device from a regulatory authority and how those parts were submitted/responded to?
 

Watchcat

Trusted Information Resource
#9
Nope, it's a startup getting ready for their first submission to FDA. They at least seem to have some professional software developers, but not with medical device experience.

Beyond that, I already have another similar startup, so I'd like to come up with a way to not only explain this to them, and also to figure out how I go about answering my first question...first and foremost, to my own satisfaction.

- Once you validate the software, what is left to do for design V&V?
I think the digital tsnunami may have finally arrived. I was really hoping to retire before that happened, sigh.
 

Tidge

Trusted Information Resource
#10
I would say software validation is equivalent to Design Verification, ie ensuring the correct code has been written. However you still need Design Validation, so Human Factors studies and making sure the code does what was intended in the first place.
I don't disagree with the second sentence, but I think the first treads on dangerous ground. I wouldn't say that part out loud or write it down.
 
Thread starter Similar threads Forum Replies Date
R Validation of Software used in Verification Testing ISO 13485:2016 - Medical Device Quality Management Systems 2
M Initial Importer/Distributor and Software Validation IEC 62304 - Medical Device Software Life Cycle Processes 1
E ISO 13485 software validation ISO 13485:2016 - Medical Device Quality Management Systems 7
Y ISO 13485:2015 Software Validation IQ/OQ/PQ ISO 13485:2016 - Medical Device Quality Management Systems 13
R Debug mode in software/device validation IEC 62304 - Medical Device Software Life Cycle Processes 2
M Software verification and validation AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 4
P Software verification and validation procedure IEC 62304 - Medical Device Software Life Cycle Processes 6
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
S SOP for ISO 13485:2016 Quality related Software validation ISO 13485:2016 - Medical Device Quality Management Systems 13
K ERP System Software Validation - ISO13485 2016 4.1.6 Design and Development of Products and Processes 11
D Software validation in Medical Equipment Other Medical Device and Orthopedic Related Topics 20
C Looking for simple Software Validation IQ templates. Qualification and Validation (including 21 CFR Part 11) 4
C Software validation - Off The Shelf Software - Web hosted ISO 13485:2016 - Medical Device Quality Management Systems 6
R Validation of Medical Device Hardware containing Software - How many to Validate ISO 13485:2016 - Medical Device Quality Management Systems 1
F 21 CFR Part 11 - Implicit requirements - Validation plan for a Software as a Service Other US Medical Device Regulations 1
R ISO 13485 Software validation procedure and Quality Objectives Monitoring wanted Document Control Systems, Procedures, Forms and Templates 1
S Validation of COTS Equipment plus Software Qualification and Validation (including 21 CFR Part 11) 12
D Software Validation - Contract manufacturer of Components (PCBA's) Qualification and Validation (including 21 CFR Part 11) 7
Pmarszal Software Validation Training Course - Recommendations Training - Internal, External, Online and Distance Learning 3
T Software Validation Certificate (ISO 13485:2016) ISO 13485:2016 - Medical Device Quality Management Systems 19
R FDA Requirements - Printing Equipment Software Validation Qualification and Validation (including 21 CFR Part 11) 1
G Windows 10 OS build Software Validation US Food and Drug Administration (FDA) 1
S Where to keep Enterprise Resource Planning software (ERP) Validation Records ISO 13485:2016 - Medical Device Quality Management Systems 1
C Software validation (4.1.6 ISO 13485:2016) ISO 13485:2016 - Medical Device Quality Management Systems 20
M Software Validation Guidance Suggestions Various Other Specifications, Standards, and related Requirements 6
S QMS software validation - Documentation ISO 13485:2016 - Medical Device Quality Management Systems 5
A SOP for software validation of software in medical device IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 5
R CNC Software Validation requirements as per ISO 13485:2016 Other ISO and International Standards and European Regulations 8
R Software validation - off the shelf X-Ray Software Quality Assurance 3
S What is the clause in ISO 13485 for SAP Software Validation? ISO 13485:2016 - Medical Device Quality Management Systems 3
P Software validation of Data Exporter ISO 13485:2016 - Medical Device Quality Management Systems 3
A Process Validation of QMS Software ISO 13485: 2016 Cl. 4.1.6 ISO 13485:2016 - Medical Device Quality Management Systems 26
P Software Validation for Equipment - Question ISO 13485:2016 - Medical Device Quality Management Systems 5
D Medical Device Software Tool Validation - Compilers! IEC 62304 - Medical Device Software Life Cycle Processes 7
N When is Medical Device Software Validation required? ISO 13485:2016 - Medical Device Quality Management Systems 6
O Process Mapping prior to Validation and Software to use to Map Process Maps, Process Mapping and Turtle Diagrams 12
S Software Validation – Clause 4.1.6 of ISO 13485:2016 ISO 13485:2016 - Medical Device Quality Management Systems 12
B Could I combine IQ, OQ and PQ for Minitab Software Validation ? Software Quality Assurance 3
R Document Management Software : Validation and other requirements Medical Information Technology, Medical Software and Health Informatics 8
E Software Validation - Clinical Trials US Food and Drug Administration (FDA) 3
K What are the minimum requirements for Process Validation (Software)? ISO 13485:2016 - Medical Device Quality Management Systems 5
R ISO - Clause 7.5.1.3 - CMM Software Program Validation ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
I Validation Software Experience? (Valgenesis, etc.) Quality Assurance and Compliance Software Tools and Solutions 1
M Injection Molding Machine Software Validation Calibration and Metrology Software and Hardware 2

Similar threads

Top Bottom