Hello all,

We are developing a class C medical device and have evaluated a list of SOUP anomalies from the manufacturer. Just want to know what is a typical way of documenting the list of anomalies that were evaluated.
  • Part of a risk analysis session?
  • Include the anomaly list in the architecture document?
  • Link the online SOUP list from manufacturer directly (this would be prone to change of links/reference)
Looking to see how you would document it best
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Linking to an online SOUP resource is not a good idea and you have no control over.
What I've seen done a couple of times is SOUP report that includes, among others, a description of known defects.

Shimon
 

Tidge

Trusted Information Resource
#3
At a minimum, you should capture the known state of the SOUP at the time you implement it, including any published anomolies. As @shimonv hints, the online resource could evaporate leaving you with no records. It used to be the case that suppliers of SOUP-y things like chipset-specific compilers and libraries would remove all links and documentation for almost any reason: The chipset is no longer produced, the mfg has a new suite of tools, etc.

There are many possibilities as to "where" you want to keep a record of the review of SOUP anomalies. I hesitate to be prescriptive, but a simple answer is "however you handle the review of anomalies before launch... why not review the known SOUP anomalies the same way, but as a single unit?" (1) Different companies will take a different approach to maintaining the anomaly list (post-launch); most companies are uniform in their approach that closed/resolved anomalies are not kept on a post-launch anomaly list, but are instead archived as part of the development process. (2)

I cannot guarantee the behavior/expectations of any external reviewer when it comes to software; I've seen some that keep strictly to deliverables, I've seen others obsessed with seeing raw source code. My personal experience is that there is more variability among external reviewers of software (both in the development of software in devices, as well as software control of processes) than in any other compliance area.

(1) Anomaly tracking should include risk analysis for both open/closed anomalies, but not every software anomaly's risk analysis necessarily requires a line of analysis in a risk file. For example: during initial development, a piece of code that simply doesn't work and requires rework to meet a requirement wouldn't necessarily have or need a dedicated risk analysis line, as the risk is simply that requirements aren't met. If the code is used without rework and still doesn't meet requirements, that's a bigger issue as it is an unresolved anomaly which you'll be carrying forward.

(2) If there is a SOUP anomaly than contributes to risk (pre- or post-controls) I would expect it to be analyzed in a risk document such as a Software HA, especially as you have no control over the SOUP and don't have the option of 'fixing' it. If there is a SOUP anomaly cannot be resolved, then I would expect such a specific anomaly to carried forward on the post-launch anomaly list.
 
Thread starter Similar threads Forum Replies Date
J Custom Tools as SOUP? (Software of Unknown Provenance) IEC 62304 - Medical Device Software Life Cycle Processes 9
A OTS and SOUP Software Documentation Requirements Other US Medical Device Regulations 9
K IEC 62304 - Functional and performance requirements for SOUP items IEC 62304 - Medical Device Software Life Cycle Processes 2
F Firmware as SOUP - Sensor with third party produced firmware IEC 62304 - Medical Device Software Life Cycle Processes 2
K SDK: SOUP or not? IEC 62304 - Medical Device Software Life Cycle Processes 8
P SOUP anomaly evaluation for MMA (Mobile Medical Application) IEC 62304 clause 7.1.3 IEC 62304 - Medical Device Software Life Cycle Processes 4
J Converting SOUP to "SOKP" ... IEC 62304 - Medical Device Software Life Cycle Processes 1
U The Importance of Correct Sample Size Calculation - Greek Alphabet Soup Book, Video, Blog and Web Site Reviews and Recommendations 7
S DHF/DMR/MDF for a software-only, cloud-based, single-instance device Medical Information Technology, Medical Software and Health Informatics 2
H Software Validation for FFS Packaging Machine Qualification and Validation (including 21 CFR Part 11) 1
E Any sample of a full software life cycle IEC 62304 report ( any class )? IEC 62304 - Medical Device Software Life Cycle Processes 1
Q ISO 13485 7.5.6 Validation - Off the shelf Software ISO 13485:2016 - Medical Device Quality Management Systems 3
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
J Do Software Subcontractors need to be ISO13485 compliant in the EU? EU Medical Device Regulations 3
D Safety data sheets software REACH and RoHS Conversations 2
N What are the software audit and control steps Reliability Analysis - Predictions, Testing and Standards 2
N Validating Software before getting approved as Class 2 device US Food and Drug Administration (FDA) 5
M Clinical Decision Support Software Question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P Missing 1m visual alarm signal in case of software/display failure, mitigation? ISO 14971 - Medical Device Risk Management 3
B Software service provider as critical supplier ISO 13485:2016 - Medical Device Quality Management Systems 5
S Asterisk in DOE minitab software Using Minitab Software 23
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 1
M Advice needed for SEH Compliance Software and ISNETWord Compatabiliy Occupational Health & Safety Management Standards 2
bruceian Software Quality Metrics Software Quality Assurance 11
optomist1 How Secure Are Our Software Systems Software Quality Assurance 7
M 'Active' device? Software/laptop with attached camera 'looking' at passive metal probe EU Medical Device Regulations 3
D Software validation team Misc. Quality Assurance and Business Systems Related Topics 3
O Any info on release date of FDA “Computer Software Assurance for Manufacturing and Quality System Software” document? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 0
L Radiology software Class I exemption Medical Device and FDA Regulations and Standards News 3
O Software for comparing text of PDF files Contract Review Process 2
J Implementing an ISO 13485 QMS Software ISO 13485:2016 - Medical Device Quality Management Systems 6
K Software Updates in the Field and ISO scope ISO 13485:2016 - Medical Device Quality Management Systems 2
M Recurrent event analysis software (python) General Auditing Discussions 2
Y UL 1998 Standard: software classes Software Quality Assurance 0
P Need a programmer for QVI's VMS software for optical inspection machine Inspection, Prints (Drawings), Testing, Sampling and Related Topics 0
S IEC 62304 software costs and time Medical Device and FDA Regulations and Standards News 3
S IEC 62304 - Software verification cost IEC 62304 - Medical Device Software Life Cycle Processes 3
Sravan Manchikanti Software Risk Management & probability of occurrence as per IEC 62304 IEC 62304 - Medical Device Software Life Cycle Processes 8
I Form templates for software (iso9001) Document Control Systems, Procedures, Forms and Templates 0
H Software Interface Translation IVD Regulation EU Medical Device Regulations 0
C 8.5.1.1 Control of Equipment, Tools, and Software Programs - Questions about the extent of control of NC programs AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
M IEC 62304 Software changes - Minor labeling changes on the GUI IEC 62304 - Medical Device Software Life Cycle Processes 3
silentmonkey Rationalising the level of effort and depth of software validation based on risk ISO 13485:2016 - Medical Device Quality Management Systems 10
T Do I need a qualified compiler for class B software? IEC 62304 - Medical Device Software Life Cycle Processes 3
S Manufacturing Execution Systems Software Costs Manufacturing and Related Processes 0
E 13485:2016, Sections 4.1.6, 7.5.6 and 7.6 - Validation of Software - Need some Advice please ISO 13485:2016 - Medical Device Quality Management Systems 3
R Medical Device Software Certification IEC 62304 - Medical Device Software Life Cycle Processes 1
S HIPAA-compliant monitoring software (advice needed) Hospitals, Clinics & other Health Care Providers 1
A Software bug fixes after shipping a product EU Medical Device Regulations 3
J Medical software Patient outcome Medical Information Technology, Medical Software and Health Informatics 2

Similar threads

Top Bottom