Specifying Measurement Standards on an External Calibration PO?

N

nova hotcakes

#1
Hi,
I am assisting my company in preparing for certification to ISO 9001 and have a question about clause 7.6. We have identified equipment to be sent for external calibration. I understand that we must specify the calibration requirements to the external calibration requirements - we plan to do this on the PO. My question is: is it acceptable to the auditors to place a blanket statement on calibration PO's that says: "to be calibrated to manufacturer's specifications and be traceable to NIST standard"? Or do we have to actually specify the NIST standard?

Any advice would be appreciated!
 
Elsmar Forum Sponsor

qusys

Trusted Information Resource
#2
Re: Specification of measurement standards on a external calibration PO?

Hi,
I am assisting my company in preparing for certification to ISO 9001 and have a question about clause 7.6. We have identified equipment to be sent for external calibration. I understand that we must specify the calibration requirements to the external calibration requirements - we plan to do this on the PO. My question is: is it acceptable to the auditors to place a blanket statement on calibration PO's that says: "to be calibrated to manufacturer's specifications and be traceable to NIST standard"? Or do we have to actually specify the NIST standard?

Any advice would be appreciated!
The definition of the requirements is for the organization not for the auditors. I do not know which kind of equipment/tools you send out for calibration. I presume the external lab knows the manufactures'specification for their calibration or that you provided them .
Consider to specify which kind of certifcate will be provided to you with error and accuracy data as well as the traceability to NIST. If you select ISO17025 Lab service it is ok, verifying if they have measurements requested in their scope.
 
G

Gmzita

#3
Re: Specification of measurement standards on a external calibration PO?

Nova -

I agree with Qusys' comments above. It is up to you to specify the calibration requirements of your devices. As a ISO 17025 accredited calibration Company, we state the requirements on the Purchase Orders that are issued. We use a combination of "blanket statements" as well as specific requirements based on the device. In the majority of the cases - we will specify calibration tolerances but we also have devices where we can simply specify "in accordance with the latest revision of ASTM E 74" in the case of a Force device or to "Manufacturer specifications" when it is a device like a Micrometer.

If you contract with an ISO 17025 accredited calibration lab, measurements will be traceable to the SI unit (NIST traceability) as that is a requirement for accreditation. But be sure to check their Scope of Accreditation to make sure their CMC meets your needs and they are accredited for the Range/Scale that you require. For instance - if you need a 48 inch length Ruler calibrated but their accreditation only is valid up to 24 inch - you need to keep looking.

Good Luck and if you need any more assistance - please feel free to ask.
 
N

nova hotcakes

#4
Re: Specification of measurement standards on a external calibration PO?

Thanks much! In the past we've always sent out our equipment to the same calibration company, it comes back calibrated, everyone's happy. So the requirement to specify the calibration standard to the calibration feels like additional work to satisfy the auditor. Hence my question about whether the auditor would find the statement acceptable. :)
 

qusys

Trusted Information Resource
#5
Re: Specification of measurement standards on a external calibration PO?

Thanks much! In the past we've always sent out our equipment to the same calibration company, it comes back calibrated, everyone's happy. So the requirement to specify the calibration standard to the calibration feels like additional work to satisfy the auditor. Hence my question about whether the auditor would find the statement acceptable. :)
ISO 9001 certification foresees processes in place for this ( clause 7.6). The fact that in the past everyone was happy for this could change because you are pursuing now the certification process and something could change. Awareness, roles and responbilities could vary and new task/duties should be accomplished. This does not mean you are not already working in compliance . To come back to your question, the auditor could investigate which are these manufacturer's specification in case you specify the reported statement in the PO and how you agreed upon them with your calibration service.
Consider also to establish the acceptance criteria of the calibrated tools/equipment that return to your siteafter calibration. In thsi case the certificate is key . ISO 17025 lab usage should be the best choice to meet the clause , always considering the capability of the external lab to provide measurement in their accreditation lab scope.
 
Thread starter Similar threads Forum Replies Date
N Specifying a MINITAB PC Hardware - What matters? Using Minitab Software 2
P Specifying the building of a new Calibration Laboratory General Measurement Device and Calibration Topics 6
B Registrar Specifying Audit Frequency ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
ScottK Specifying standards you don't own - non conformance or observation? Nonconformance and Corrective Action 25
P Dimensional Characteristics Rating:Guidelines for specifying Special Characteristics? Statistical Analysis Tools, Techniques and SPC 3
I QMS monitoring, measurement, analysis and evaluation requirement - Template ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
F Physical Part Measurement General Measurement Device and Calibration Topics 5
M Measurement Uncertainty in Optical Microscopy Measurement Uncertainty (MU) 1
B ISO 6508 and portable hardness measurement instruments General Measurement Device and Calibration Topics 0
C How to Establish the Calibration & Measurement Capability (CMC)? ISO 17025 related Discussions 1
Ron Rompen Measurement of residual magnetism General Measurement Device and Calibration Topics 2
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 1
M Disabling measurement data during fault conditions IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
J Most versatile measurement device General Measurement Device and Calibration Topics 2
P Surface scratch measurement General Measurement Device and Calibration Topics 1
N Drawing tolerance vs. Measurement device General Measurement Device and Calibration Topics 7
Ron Rompen MSA on automated measurement system - Multiple Step Vision System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
P Thermo-Hygrometer - Measurement of 10DegC @10%RH ISO 17025 related Discussions 4
hogheavenfarm GDT Flatness measurement question Inspection, Prints (Drawings), Testing, Sampling and Related Topics 10
K Software Validation for Measurement Tools used in Process Validation ISO 13485:2016 - Medical Device Quality Management Systems 2
K Best Measurement Systems Demos in California? General Measurement Device and Calibration Topics 1
M Measurement Error - How to determine what is acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
K Process Mapping - Inputs/Outputs/Detail Activities/Control points/Measurement? Process Maps, Process Mapping and Turtle Diagrams 4
B IEC 60601-2-10 - Accuracy of Pulse Parameters - Required Measurement Uncertainty IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N Program or application for standard time measurement in industry from video surveillance? Human Factors and Ergonomics in Engineering 1
M Defining frequency of measurement tools callibration Calibration and Metrology Software and Hardware 3
T How to “validate/qualify” a new measurement equipment? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
Bev D Verification and Validation of Measurement Systems Misc. Quality Assurance and Business Systems Related Topics 0
F How do you add accuracies for tools with different measurement devices? General Measurement Device and Calibration Topics 1
M Viable count Measurement ISO Standard Other Medical Device Related Standards 1
J Gauge R&R on a torque measurement Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
W Air Quality Measurement Hardware and Software General Measurement Device and Calibration Topics 11
O Dimension Measurement Tool Recommendation General Measurement Device and Calibration Topics 21
G Does pitch/increment/resolution of a ruled scale apply to measurement uncertainty as line item? Measurement Uncertainty (MU) 10
Marco Bernardi Plane measurement with a 3D touch machine like a CMM. Calibration and Metrology Software and Hardware 3
Marco Bernardi Diameters measurement with 3D touch machine like a CMM Calibration and Metrology Software and Hardware 4
R Uncertainty in measurement larger than tolerance Measurement Uncertainty (MU) 2
R Measurement Uncertainties Budget for Thermometer Measurement Uncertainty (MU) 6
O Performance Measurement ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
O Monitoring performance - How do I determine performance measurement basis within my organization? Misc. Quality Assurance and Business Systems Related Topics 4
G What information to put on measurement Dimensional Results APQP and PPAP 7
D O Ring capability and measurement - What is the automotive 'norm' for capability studies on O Rings? General Measurement Device and Calibration Topics 3
F Electrical Metrology (Source/Generation and Measurement) General Measurement Device and Calibration Topics 1
G Gauge R&R on multiple dimensions using 3D measurement system Capability, Accuracy and Stability - Processes, Machines, etc. 6
F Measurement Audit and ILC for ISO 17025 Clause 7.7.2 - Comparison with results of other laboratories ISO 17025 related Discussions 0
Emran.mi Measurement system analysis - Can you help me about implementation MSA for CMM device Manufacturing and Related Processes 2
K ISO 17025:2017 clause 7.6.2 - Performing calibration of its own equipment shall evaluate the measurement uncertainty ISO 17025 related Discussions 6
Q How should I analyze measurement correlation between me and customer? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
I Determining Calibration Tolerance of a Measurement Device General Measurement Device and Calibration Topics 2
Fernando Drumond Calibration and Control of Plastic Volumetric Measurement Instruments General Measurement Device and Calibration Topics 1

Similar threads

Top Bottom