Stability of the gage to be determined first and then gage R&R?

H

hemant

#1
in the survillance audit auditor has raised the issue that stability of the gage to be done first and then only R&R.
as per me this is hen & egg situation.
if R&R is not less than 10% how stability can be exibited?

[This message has been edited by hemant (edited 06 November 2000).]
 
Elsmar Forum Sponsor

Atul Khandekar

Quite Involved in Discussions
#2
One has to determine the inherent "Statistical Stability or Control" of the measurement process over a period of time. The methodology used for this is the same as that for conducting an SPC study. ie.
1. Determine Sample size and frequency
2. Plot control charts
3. Analyse / separate "common cause" and "special cause" variation
4. Look for out-of-control and trend signals

An out-of-control Range chart indicates unstable repeatability (too much variation) and an out-of-control X-bar chart may mean that the instrument is due for calibration (bias).

A measurement process must be statistically under control (no special cause variations) for it to be used effectively to control the manufacturing process it measures. If there are no out-of-control signals, Standard Deviation can be estimated and compared with process standard deviation to determine suitability of the measurement process for the application.

Once the statistical stability is established, R&R study is then conducted to 'quantify various errors' in the measurement system. Based on these (r&r percentages), one can determine the suitability of the measurement system for a particular measurement application.

Quoting from MSA manual "...Without data-based knowledge of the state of control of a measurement process, figures of 'repeatability' , 'reproducibility', etc. are only descriptions of data obtained during the (r&r) study. They may have no meaning for future performance. Assessing the repeatability, reproducibility etc. of a measurement system for which the state of stability is unknown may cause more harm than good..."

- Atul
 
H

hemant

#3
on reflecting further to my original post, i have gone through the MSAMEASUREMENT SYSTEM ANALYSIS (MSA) book ifeel that

STABILITY:

When considering the subject of stability in connection with measurement system, it becomes extremely important to differentiate between what is generally referred to as measurement system stability ----

a ) The amount of total variation in the system’s bias over time on a given part or master part : Known as "Stability over Time"

and

statistical stability, the more general term which is applied to not only stability ,but to repeatability, bias, process in general., etc.

to understand the difference between the two stability, let us consider that there can be 2-measurement systems, measuring exactly the same master part, both of which demonstrate statistical stability, yet one system may have significantly higher variation in its bias over time than the other. From a statistical standpoint, they are equally stable. From a traditional gage stability stand point, the system with greater bias variations over time is considered less “ STABLE “ than the one with lower bias variation. ----------------- (page-21,chapter-2-section-2 )

Statistical properties of measurement systems :

The measurement system must be in statistical control. This means that variation in the measurement system is due to common causes only and not due to special causes. This can be referred to as statistical stability. ---------- (page-5, chapter –1 , section-2 )

Quality of measurement data
The statistical properties most commonly used to characterize the quality of data are

Bias and Variance.
The property called bias refers to the location of the data relative to the master value and the property called variance refers to the spread of the data.
---------- ( page-3 , chapter-1,section-1)

specifically, the procedures assess the following statistical properties;

Repeatability, reproducibility, bias, stability, and linearity.

Collectively, the procedures are sometimes referred to as “gage R&R” procedures. ------------ (page-15,chapter-15,-section-1)
DOES THE GAGE R&R ASSESS THE SAID STATISTICAL PROPERTIES OTHER THAN REPEATABILITY & REPRODUCIBILITY. YES/NO.

If yes -------------- how

Analysis of results ---graphical analysis


Stability
From range chart stability is determined by: a point or points beyond the control limit ; within operator or within part patterns, ------- ( page-46 , chapter 2- section 4 )
linearity

the averages of the multiple readings by each appraiser on each part are plotted with the reference value or overall part average as the index. This plot can assist in determining :

linearity (if the reference value is used ) ------- (page-52, chapter 2-section4 )

CONCLUSION

GAGE R&R study is an apt exercise for statistical stability and if used, as said above all statistical properties are revealed. Like stability, linearity, repeatability & reproducibility. GRR should be done first and GRR values should be brought below 10% . This can be achieved by understanding the graphical representation and taking appropriate steps. By doing so we are achieving the "Statistical stability".

Without data –based knowledge of the state of control of a measuring process, R&R figures are only descriptions of the data obtained during study. They have no meaning for future performance. Assessing the repeatability ,reproducibility, etc, of a measurement system for which the state of stability is unknown may cause more harm than good. When talking of measurement system statistical stability, the length of time a system is stable is often a major point of discussion. However by means of TIME STABILITY, the length of time a system is stable can be found by using x-bar r-bar control chart. This time stability is to be done after statistical stability in other words called gage R&R. Incase if time stability is performed prior to gage R&R the bias readings will not be exact as readings are contaminated with repeatability and, reproducibility errors.

Above all, any manufacturing process is supposed to be statistically stable if CP & CPK are controlled as they are representing spread & bias (centrality. ).The normal practice to control SPREAD first then to CPK the bias or centering OPN.

Similarly, we have to look at measurement system.

First control repeatability and reproducibility errors of SPREAD by doing gage R&R STUDY and then go to TIME STABILITY to know the extent of DRIFT OR BIAS the CPK.


the book reference r indicated in the brackets.

invite views on this
 
L

Lyndon

#4
i do not agree that gage r&r ensures statistical stability....in my opinion,stability studies have to be done simultaneously with r&r studies,so that special causes occuring in between 2 r&r studies can be noticed.if such a stability study is not carried out simultaneously,then any special cause occuring wil not be noticed,which can give u a lower value of %r&r inthe second study,thus masking the improvement u have made in reducing the variation.
e.g i have done msa studies where by cleaning a comtor gage(forchecking bore diameter),i have got better %r&r,in reality what i have done is improved stabiliy by reducing a special cause(that of dust,dirt etc),and not improved either the gage repeatibilty or the appraiser reproducibility
but i agree with hemant that..a time stability has to be ALSO carried out later once u have got 10%r&f.....but i repeat my point that stability studies should be carried out simultanouesly with any r&r studies,in order to be aware of the occurence of any special cause
 
#5
i do not agree that gage r&r ensures statistical stability....in my opinion,stability studies have to be done simultaneously with r&r studies,so that special causes occuring in between 2 r&r studies can be noticed.if such a stability study is not carried out simultaneously,then any special cause occuring wil not be noticed,which can give u a lower value of %r&r inthe second study,thus masking the improvement u have made in reducing the variation.
I know its an old thread,... but anyone still there.....Can someone tell me if I get %GRR under 10%, then how long do I need to do stability tests to make sure that my gage works for e.g. an year without giving me an error. How is the time period for a stability test decided????

Thanks guys!
 

Miner

Forum Moderator
Staff member
Admin
#6
The stability test should have a duration that exposes the gage to the extremes of environmental variation that might be expected while in use. This could be a day or a week.
 
#7
Under what conditions these statistical stability requirements discussed in above posts can be waived????
 

Jim Wynne

Staff member
Admin
#8
Under what conditions these statistical stability requirements discussed in above posts can be waived????
That would be up to your customer(s). If you can provide sensible rationale for not doing stability studies, present it to the customer and see what happens.
 

Miner

Forum Moderator
Staff member
Admin
#9
If you have similar gages with a proven stability there should be no reason to duplicate a stability study. It is more important a with new style gage with which you have no experience.
 
B

blair_liu

#10
Hello everyone,
who could help me on this?
how about the bias,linearity? what's the right time to carry them out?
is there any relationship with GRR?
best regards and thanks
Helly
 
Thread starter Similar threads Forum Replies Date
Y Product Stability (Machine Capability) vs. Equipment Repeatability (Gage R&R) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
D Measurement System Analysis (MSA) - Requirement to also do gage stability studies? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Scales - Gage R&R, Bias, Linearity, Stability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
H Gage Stability Study Requirements - Gage R&R Study Data Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
B Stability study- Plastic injection molded parts with fixture gage & digital indicator Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
F 1. Suggest a Gage for Checking Rounding of Gears 2. When to conduct Stability Study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
F Meeting the AIAG's MSA Manual Gage R&R Bias, Linearity, Stability requirements Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
D Gage R&R - Acceptance Criteria for Linearity, Bias, and Stability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
G Auditor wants our Gage R&R Procedures to address Linearity, Accuracy and Stability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 22
G Engineering retest in line - Machine stability issues IATF 16949 - Automotive Quality Systems Standard 8
G Accelerated stability of medical devices Other Medical Device Related Standards 1
R Revalidation machine line and Stability Studies impact. Medical Device and FDA Regulations and Standards News 3
S Stability accelerated vs. real-time Other Medical Device Related Standards 3
S Standards for real time stability for medical devices Other Medical Device Related Standards 2
C Medical Device Stability Studies - Non-validated pilot production Other Medical Device Related Standards 1
B Stability testing of medical devices as per ASTM F1980-16 CE Marking (Conformité Européene) / CB Scheme 0
lilybef Number of lots required for stability testing of a Class II device in the 510(k) Qualification and Validation (including 21 CFR Part 11) 3
V System suitability or stability during continuous usage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
F Extension of a stability study for an API Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
O New stability - Yes or No - Similar product but for different intended use EU Medical Device Regulations 1
A Accelerated Stability Studies - Class III Medical Gloves Other Medical Device and Orthopedic Related Topics 1
E On going Stability for IVDs Other US Medical Device Regulations 5
A Risks related to Method Validation and Stability Studies Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
N Sampling for stability studies - Class II surgically invasive devices Other Medical Device and Orthopedic Related Topics 1
K Food Shelf Life Stability Study Statistical Analysis Tools, Techniques and SPC 1
M Electrosurgical Electrodes - Stability Test Other Medical Device Related Standards 1
C IQ and OQ for a "Dormitory" Refrigerator - Temperature Stability Qualification and Validation (including 21 CFR Part 11) 4
R Holding time stability study of intermediates of drug substances US Food and Drug Administration (FDA) 2
S Stability test - primary packaging EU Medical Device Regulations 7
L MSA - Stability - Part of our PQ (performance qualification) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
C Sample Size for Stability Testing Other Medical Device and Orthopedic Related Topics 2
J EU Medical Device Class I - Stability (Shelf Life) Test EU Medical Device Regulations 2
E EN 13640:2002 or EN23640:2013 For stability testing of in vitro diagnostic reagents ISO 13485:2016 - Medical Device Quality Management Systems 2
M Is a Real Time Stability Study required to determine Expiration Date? EU Medical Device Regulations 9
P Specification and Tolerance in Stability Control Chart Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 9
P The correct way to do a Linearity and Stability Study on a Critical Dimension Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 8
M How to provide Stability Study for micron-precision tool Misc. Quality Assurance and Business Systems Related Topics 2
R Convincing Management Cpk without Stability is Meaningless Capability, Accuracy and Stability - Processes, Machines, etc. 7
M A question on Measurement System Analysis - Stability to measure got Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
R Drug Stability Data at the time of Filing US Food and Drug Administration (FDA) 3
V Can we apply bracketing & matrixing simultaneously for stability studies @ ICH Q1D US Food and Drug Administration (FDA) 1
M Ownership of Pre-Clinical cGMP Stability Testing Other Medical Device and Orthopedic Related Topics 2
R Shipping Stability for Class III IVD (Non Sterile) Medical Device Other Medical Device Related Standards 7
B How to Demonstrate Stability and Repeatability in very Short Run Process AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 10
J How do you perform a S3 Stability Study for a Non-Replicable Measurement System Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
E Company-wide MSA - Measurement Equipment Bias, Linearity & Stability Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
V ICH - Stability studies, alternate conditions... Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
L Stability of the Process vs Improvement during Auditing General Auditing Discussions 8
B Interpretation of Stability Study Data Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3