Standards that exist regarding Sterilization of Polypropylene?
T
tdeardorff
#1
Hello! I am looking for some information on any standards that are in existance concerning acceptable and effectiveness of sterilization of polypropylene material. We produce a very simple (but Class IIa according to the EU) dental device called an HVE tip- it is used to attach to the suction unit in a dental office...most are disposable BUT we do have one that the dentists actually autoclave and re-use. We specify the normal autoclaving times of 121 degrees C for 15 mins which is an industry standard autoclave cycle. The problem is that this "industry standard" is not identified anywhere specifically for polypropylene. Does anyone out there have any experience with this or can you point me in the right direction? If we can locate a specific standard or accredited testing(publication in journal of medicing, AMI or ADA publication) already completed in autoclaving polyprop, it will save us(me) a lot of headaches in getting testing completed.
Elsmar Forum Sponsor
Al Rosen
Staff member
Super Moderator
#2
I posted the list of harmonized standards yesterday. You should be able to find a standard within that list. There are also some threads addressing the subject of sterilization including standards that you can search for. Also, scroll down towards the bottom of this page there should be a list of similar threads.
W
wrodnigg
#3
I assume you want to use moist heat, then you need the standard
EN 554:1994 “Sterilisation of medical devices. Validation and routine control of sterilisation by moist heat”.
This is the basic guideline and tells you how to proceed with the design and validation of a sterilization process.
For estimation of the Bioburden you should take a look at EN 1174-1 -2:1996 “Sterilisation of medical devices. Estimation of the population of microorganisms on products. Part 1 Requirements, Part 2: guidance“
It does not help, if you just take standard parameters from a third source, because you have to have records from your sterilization validation process (described in EN 554).
Beside that, the 121°C over 15 minutes is a recommendation from the European Pharmacopoeia. And the temperature shall be on/in the product, not on the controller or display of the sterilizer. Also when using these standard parameters, you have to provide the documented evidence, that these parameters are effective for sterilization of your products (as described in EN 554).
Thread starter Similar threads Forum Replies Date
T Assessing Hazard-Related Use Scenarios where control measures exist through standards IEC 62366 - Medical Device Usability Engineering 32
Raffy No standards exist for 7.6a of ISO 17025 ISO 17025 related Discussions 7
M How detailed should procedures be if standards or guidance documents exist? Document Control Systems, Procedures, Forms and Templates 10
C Electric Signs Certification - UL standards exist - Are there any european ones? Various Other Specifications, Standards, and related Requirements 3
A Where no standards exist the basis used for calibration shall be recorded General Measurement Device and Calibration Topics 4
M Packaging standards - ASTM vs ISTA Other Medical Device Related Standards 0
M Label Making & Printing Standards ISO / ASTM ISO 13485:2016 - Medical Device Quality Management Systems 5
M Why are EU harmonized particular standards so old & out-of-date? IEC 60601 - Medical Electrical Equipment Safety Standards Series 10
R Weld Penetration Standards and Requirements Manufacturing and Related Processes 2
S Any ISO standards around Artificial Intelligence and Machine Learning? Medical Information Technology, Medical Software and Health Informatics 4
M ERP / QMS related software standards for Validation IEC 62304 - Medical Device Software Life Cycle Processes 6
M MDR harmonized standards EU Medical Device Regulations 3
S Where do l get calibration standards to run a calibration lab? Other ISO and International Standards and European Regulations 2
C Appendix Z in Harmonised Standards? EU Medical Device Regulations 13
Ajit Basrur Corporate management standards ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
Revision36 Engineering Change Order Industry Standards Other Medical Device and Orthopedic Related Topics 4
R Applicable ISO standards - packaging and others Medical Device and FDA Regulations and Standards News 4
Q Standards for Shipping by Sea US Food and Drug Administration (FDA) 2
N Database that lists labs accredited to ASTM / ANSI standards Various Other Specifications, Standards, and related Requirements 6
S API Spec.Q1 Standards Version in Purchasing Information Oil and Gas Industry Standards and Regulations 1
B Exclusions or justification for non-applicability of IEC standards Reliability Analysis - Predictions, Testing and Standards 1
P Citing EN standards Other Medical Device Related Standards 6
Ed Panek Compliance with Standards? When a standard is updated/revised CE Marking (Conformité Européene) / CB Scheme 3
N Harmonized Standards for MDR Certification EU Medical Device Regulations 7
A Some questions about voluntary consensus standards in traditional 510(k) submission US Food and Drug Administration (FDA) 6
D Imaging/Image Quality Standards Other Medical Device Related Standards 1
R ECG Risk Analysis Standards ISO 14971 - Medical Device Risk Management 2
I Management review in conformity assessment standards - Certification Bodies Management Review Meetings and related Processes 6
Govind Informational ISO COVID-19 response: freely available ISO standards Other ISO and International Standards and European Regulations 1
C Does a CE mark infer meeting all applicable standards? CE Marking (Conformité Européene) / CB Scheme 4
Richard Regalado Informational ISO makes available relevant BCMS standards free for online access Business Continuity & Resiliency Planning (BCRP) 6
A Technical File - Updating Applied Standards EU Medical Device Regulations 4
A NB (Notified Body) Audit of Standards ISO 13485:2016 - Medical Device Quality Management Systems 3
R Applicability of new non-harmonized standards (MDD/MDR) EU Medical Device Regulations 14
dgrainger Informational EU published new Harmonised Standards List for the Directives (25/3/2020) Medical Device and FDA Regulations and Standards News 0
P Static Electricity/Pacemaker - Standards for static electricity generation CE Marking (Conformité Européene) / CB Scheme 3
M Referencing Harmonised Standards in Declarations of Conformity EU Medical Device Regulations 4
Pmarszal External Standards and Regulations Management Process Document Control Systems, Procedures, Forms and Templates 10
S IVF medical devices standards and test methods Medical Device and FDA Regulations and Standards News 4
M What are the standards for malls management? ISO 14001:2015 Specific Discussions 1
A Complaint alert levels - Industry standards pertaining to alert levels for complaints Customer Complaints 5
R MDR standards - which standards to comply with ? EU Medical Device Regulations 3
D Why does Official Journal list superseded standards? EU Medical Device Regulations 6
M MDR "Common Specifications" (CS) - If not standards, then what? EU Medical Device Regulations 5
L In house calibration - Our CMM's are outside certified can I use them to certify our standards General Measurement Device and Calibration Topics 4
F AS9100D, raw material standards are a requirement for manufacturing? AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 5
S Industry Labelling Standards - non medical Manufacturing and Related Processes 0
J Medical device related standards or regulations - Disposed records log ISO 13485:2016 - Medical Device Quality Management Systems 2
S How many mandays required for the below mentioned standards IEC 27001 - Information Security Management Systems (ISMS) 3
M Standards Subscription Service and Standards Tracking Service Pros and Cons? Other US Medical Device Regulations 3
Similar threads
Top Bottom