ISO 13485:2016 and the QSR both require a statistical based justification behind the selection of the sampling size in V&V testing.
For one of our V&V tests, we tested 3 samples because we followed an harmonized standard which is specific for this type of test and requires the use of 3 samples.
My question is: will the NB accept the reference to the harmonized standard to justify the selection of 3 samples without any further "statistical" justification? It makes sense to me that the regulatory body that accepts that standard should also accept the sample size of the standard. What do you think about it? What is your experience with you NB or FDA in relation to statistical justification? Thank you for your feedback!
shimonv
It's difficult to say without knowing what the product is and to which harmonized standard you are referring to, but I can say this:
Sample size is risk-based; a high risk requires a statistical prediction with high confidence level and high reliability, which means a larger sample size.
You need to look carefully in the harmonized standard and see whether the recommended sample size of 3 matches the description of your V&V plan. If so, you can reference it in your V&V protocol and that should be sufficient.
Shimon
Ronen E
Problem Solver
Moderator
I would carefully look into the standard's wording. 3 datapoints is definitely not enough for any meaningful statistical prediction; however, 3 samples doesn't necessarily mean 3 datapoints as you could test each unit more than once (and there are statistical methods based on such an approach).
shimonv
Come to think of it, I have never seen a standard that is bold enough to prescribe a sample size for a specific test. Even if the test procedure is described in detail in the standard - the medical device can be used in a number of ways and in different clinical application. The risk levels might be different from one case to another, which means different levels statistical confidence might be required, which means different sample size...
Highground
Shimon
This makes sense and helps me out...V&V plan actually has the sample size rationale. :)
V Statistical basis and justification while comparing / changing sampling plans Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 11
R Statistical Methods for comparing test and reference product equivalence for quality attributes US Food and Drug Administration (FDA) 3
S MSA clause 7.1.5.1.1 - Statistical studies on counting weight scale Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 12
S How to perform verification of the Statistical Analysis Software? Qualification and Validation (including 21 CFR Part 11) 7
Bev D Statistical Alchemy Misc. Quality Assurance and Business Systems Related Topics 1
K Looking for guidance to write an SOP on Statistical Methodologies? Statistical Analysis Tools, Techniques and SPC 7
M Minimum sample size - Guidance and statistical rationale Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
N Design Verification & Process Validation - Statistical sample sizes Design and Development of Products and Processes 2
John Predmore Interactive visualization through graphical simulation of statistical concepts Statistical Analysis Tools, Techniques and SPC 3
Bev D Profound Statistical Concepts Misc. Quality Assurance and Business Systems Related Topics 63
A Statistical Analysis - Check if these organisms at different concentrations affect the growth of wheat seedlings Using Minitab Software 4
H Statistical Techniques Procedure - What should be included Document Control Systems, Procedures, Forms and Templates 4
optomist1 It’s time to talk about ditching statistical significance Statistical Analysis Tools, Techniques and SPC 6
Marc Steve Prevette's Statistical Process Control (SPC) "Library" Statistical Analysis Tools, Techniques and SPC 0
John Predmore A Balanced view of statistical tests Statistical Analysis Tools, Techniques and SPC 3
S SPC (Statistical Process Control) for Unilateral Tolerance - Questions Statistical Analysis Tools, Techniques and SPC 6
S IATF 16949 9.1.1.3 Application of statistical concepts - Our technicians are quizzed for statistical knowledge IATF 16949 - Automotive Quality Systems Standard 3
K Please help identify appropriate statistical treatment Statistical Analysis Tools, Techniques and SPC 13
ScottK Statistical basis for 30 pieces for FAI 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
B IATF 16949 clause 7.1.5.1.1 - Statistical studies shall be conducted IATF 16949 - Automotive Quality Systems Standard 3
A Statistical Process Control and Inspection in Footwear Production Statistical Analysis Tools, Techniques and SPC 0
M IATF 16949 Cl. 7.1.5.1.1 - Statistical studies shall be conducted IATF 16949 - Automotive Quality Systems Standard 3
Steve Prevette Statistical Process Control Library Statistical Analysis Tools, Techniques and SPC 17
Marc Happy Birthday Statistical Steven - 2015 Covegratulations 10
L When are Statistical techniques not applicable? Service Industry Specific Topics 16
M FDA 21 CFR 820.250 - Does "valid statistical" always mean math? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
R Common Statistical Errors Using Minitab Software 1
Y Statistical Analysis of Road Traffic Data Statistical Analysis Tools, Techniques and SPC 11
B Class II Medical Device Manufacturer - SOP for 820.250 Statistical 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
E Correct Statistical Test comparing 2 Groups Statistical Analysis Tools, Techniques and SPC 14
A Statistical Software Calibration using Ford's "Sample Calibration File" Statistical Analysis Tools, Techniques and SPC 8
J Defining Martial Arts and Gymnastics Statistical Techniques Statistical Analysis Tools, Techniques and SPC 4
J Capability Analysis - Unusual Statistical Distribution of my Proccess Capability, Accuracy and Stability - Processes, Machines, etc. 5
M PhD Thesis Data Statistical Analysis Methods Statistical Analysis Tools, Techniques and SPC 2
J Statistical Significance and SPC Control Chart Reports Statistical Analysis Tools, Techniques and SPC 9
N Statistical Quality Improvement Action for Small Batch Production Statistical Analysis Tools, Techniques and SPC 17
V Validation of macro - scripts - programs used in statistical software (Minitab-SAS... Qualification and Validation (including 21 CFR Part 11) 5
Moncia Statistical Process Control Crash Course - Question Quality Manager and Management Related Issues 10
H Statistical Models for Predictive Management of Software Processes Software Quality Assurance 2
A Statistical Correlation between ordered SKUs Statistical Analysis Tools, Techniques and SPC 8
S Minitab and Crystal Ball Statistical Analysis Software Using Minitab Software 13
M Determining if two different X's have any Statistical Significance on the Y's Statistical Analysis Tools, Techniques and SPC 4
I Statistical Stability for the PQ of Analytical Equipment Qualification and Validation (including 21 CFR Part 11) 1
F Statistical Comparison of Product: High Average vs. Low Range Capability, Accuracy and Stability - Processes, Machines, etc. 13
E Using ANOVA during the PQ Validation Run to evaluate Statistical Differences Statistical Analysis Tools, Techniques and SPC 4
W Gage R&R for gage pins used to inspect a hole ID called a Statistical Tolerance Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
B AS 9100C sec 8.2.4 - " Recognized Statistical Principles" meaning AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 7
R What is PSW - Statistical Process Package + Level 5 APQP and PPAP 7
O Is SPC (Statistical Process Control) always required? Statistical Analysis Tools, Techniques and SPC 4