I would start with a raising a change control and following your procedures. Perform an impact analysis. Look at your
dFMEA, design inputs/outputs and
pFMEA to see where the material change may have an effect. Also check through the Essential Requirements and clinical evaluation. Then perform the necessary verification to ensure all requirements are met. If you have a notified body, assess whether they need to be informed.