Sterile medical device expiration date

#1
hello,

I would appreciate any help with the following topics:

1. if I want to change the expiration date for a medical device (for example from 1 year to 2 years) - what testing is required?

2. can anyone please help interpret this statement from ISO 11607:
6.4.7 When it is demonstrated that the product does not interact with the specified sterile barrier system over time, previously documented data for stability testing shall be sufficient to be in accordance with 6.4.1.

thank you!
 
Elsmar Forum Sponsor

shimonv

Trusted Information Resource
#2
Hi Jane,
You have already proved that the sterilisation process works. Now you need to show that the packaging system (combination of the sterile barrier system and protective packaging) holds for a longer period of time, i.e. you need to show that the package integrity and device performance meet the predeterimed specifications after two year.

Shimon
 
#3
Hi Jane,
You have already proved that the sterilisation process works. Now you need to show that the packaging system (combination of the sterile barrier system and protective packaging) holds for a longer period of time, i.e. you need to show that the package integrity and device performance meet the predeterimed specifications after two year.

Shimon
thank you very much for your reply. could you please direct me to a gauidance\standard were it is stated that for each new\change in expiration date the packaging integrity tests need to be repeated (I'm asking because I want to have a source for showing that it is required not only because of common sense but also by regulation).
 

shimonv

Trusted Information Resource
#4
The standard does not say explicitly that each change in expiration date warrants a repeat of packaging integrity tests; it couldn't. As you can imagine some changes in expiration date will also require a change in device material.
Every change needs to be evaluated through the change control process. See ISO 11607-2, section 5.7, "Process changes and revalidation":

The AAMI TIR22:2007/A1:2008, section 13.3 says the following:
13.3 Revalidation is often affiliated with design control and the associated change control procedure, the extent of the revalidation chosen will depend on the nature of the change and how it affects the process or device.

The governing principle is found it ISO 11607-1, section 6.1.4 and 6.4 (Especially 6.4.1)

6.1.4 The sterile barrier system shall maintain sterility until the point of use or until the expiry date.

6.4.1 Stability testing shall demonstrate that the sterile barrier system maintains integrity over time.



Good luck!
Shimon
 

SKM.Sunil

Involved In Discussions
#5
hello,

I would appreciate any help with the following topics:

1. if I want to change the expiration date for a medical device (for example from 1 year to 2 years) - what testing is required?

2. can anyone please help interpret this statement from ISO 11607:
6.4.7 When it is demonstrated that the product does not interact with the specified sterile barrier system over time, previously documented data for stability testing shall be sufficient to be in accordance with 6.4.1.

thank you!

As your device is sterile....You required to produce "real time stability data"
 

Ronen E

Problem Solver
Staff member
Moderator
#6
@Jane1
Please don't forget that extending the device's expiration involves not only validating the packaging/sterile barrier (11607 etc.) but also validating the extended life of the device itself. You won't find any published standard for that, because there is none, not really. Beware of attempting accelerated ageing based on raised temperature, it doesn't necessarily apply to device ageing.
 
Last edited:

indubioush

Quite Involved in Discussions
#7
See ASTM F1980-16 Accelerated Aging of Sterile Barrier Systems.

Even though it is true that you need real-time aging data to prove the shelf life of the device, often companies will do accelerated aging for device and packaging together and submit this data to support a preliminary shelf life pending real-time data.
 

Ronen E

Problem Solver
Staff member
Moderator
#8
See ASTM F1980-16 ... often companies will do accelerated aging for device and packaging together and submit this data to support a preliminary shelf life pending real-time data.
Yes, and often there's very little scientific/engineering substantiation behind that approach (which seems to bother reviewers quite little).
F-1980 specifically states in its title what its purpose is.
 
Last edited:
Thread starter Similar threads Forum Replies Date
K CE Marking Class 1 (Non sterile) medical device CE Marking (Conformité Européene) / CB Scheme 3
A FDA guidance on non-sterile Medical Device Packaging Medical Device and FDA Regulations and Standards News 7
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
R Deciding if a medical device shall be supplied sterile or not Other Medical Device and Orthopedic Related Topics 5
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
D What is your Sterile Medical Device Pouch Peel Strength Spec? Other Medical Device Related Standards 3
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
N Medical device non sterile packaging requirement CE Marking (Conformité Européene) / CB Scheme 1
N Registering a Class 1 Sterile Medical Device - Currently selling a Non-Sterile 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
3 Egypt Medical Device Registration: Only required for Sterile Medical Devices? Other Medical Device Regulations World-Wide 10
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
F Date of Manufacture and Expiration (Sterile Medical Device) question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
D Sterile Medical Device Temperature and Humidity (Storage) Labeling Requirements EU Medical Device Regulations 5
T Sterile Medical Device Storage Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
J 510k file requirements for non-sterile medical device accessories 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
J Non-Sterile Class 2 Medical Device Shelf Life Other Medical Device Related Standards 3
R Shipping Stability for Class III IVD (Non Sterile) Medical Device Other Medical Device Related Standards 7
J0anne MOC - Sterile Medical Device Marketing - India Other Medical Device Regulations World-Wide 1
J Sterile Medical Device Labeling Requirments - ISO 1041 Other Medical Device and Orthopedic Related Topics 1
J Biocompatibility for Class I Non-Sterile Medical Device EU Medical Device Regulations 2
M Distribution Center has shipped Sterile Medical Device products that have Expired EU Medical Device Regulations 2
H Medical Device RUO (Research Use Only) and 'Sterile' Label Claims 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Symbol for Medical Device Sterile Packaging - "Inspect Seal Before Opening" Other Medical Device Related Standards 5
Q Expiry Dates on a Non-Sterile Class 1 Medical Device (5 Year Shelf Life). Other Medical Device and Orthopedic Related Topics 11
A Need to register with CA for Class I (Non-Sterile and Non-Measuring) Medical Device EU Medical Device Regulations 4
M Process Validation/Equipment Qualification - Class 1A (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 12
M Shelf Life Testing of a Sterile Medical Device Other Medical Device and Orthopedic Related Topics 40
M ISO 14971:2007 Risk Management - Class I A Sterile Medical Device ISO 14971 - Medical Device Risk Management 4
S Sterile Product Ship Testing (Medical Device Packaging) EU Medical Device Regulations 3
J Addition of a Class 1 (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
D Selling Class IIa medical device (sterile syringes) in USA ISO 13485:2016 - Medical Device Quality Management Systems 2
A Documentation required for launch of a non-sterile exempt class 1 medical device ISO 13485:2016 - Medical Device Quality Management Systems 7
S Environmental Control Standards - Sterile medical device manufacturing environment ISO 13485:2016 - Medical Device Quality Management Systems 3
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
J Use of expired perminant implantable sterile medical products ISO 13485:2016 - Medical Device Quality Management Systems 2
S Medical Implant Sterile Packaging Design EU Medical Device Regulations 6
P Sterile Medical Packaging and Shelf Life Validation Sample size ISO 13485:2016 - Medical Device Quality Management Systems 10
S Do Medical Sterile Drapes count as MOPP? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Packaging Requirements for Non-Sterile Class II Medical Devices EU Medical Device Regulations 11
Q Acceptable Visible Particulate Size on Sterile Medical Devices Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
C Cleanliness Specifications that apply to non-sterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 12
W Distributor Requirements for Sterile Disposable Medical Devices EU Medical Device Regulations 8
J Medical Devices: Initial Design - Sterile or Non-sterile Design and Development of Products and Processes 10
M Bioburden Frequency (sterile MD) Other Medical Device Related Standards 2
shimonv Single lot release for sterile packaging 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3

Similar threads

Top Bottom