I've got a sterile device whose shelf life of four-years is based on completed real-time package integrity studies. Right now we do not have temperature and humidity labeling on our product. In fact in our IFU we say "no special storage conditions are necessary". Usually, when customers ask, they are okay with this... That all changed when today we had a customer that wanted the labeling to include storage and humidity requirements. Before we embark on testing to establish a temperature and humidity range, I need to know: Is there a requirement for storage conditions to be on a sterile medical device package label?
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
Storage conditions limits need to be indicated where the device or the sterile packaging integrity might be compromised and exceeding those limits is a probable scenario.

I'm not quoting a single specific requirement that spells it out, however this can be derived from the Essential Requirements in the MDD.

This is a bit vague, but ER 13.3(i) states that the label must bear "any special storage and/or handling conditions".
 

dr1vn

Involved In Discussions
#3
Thanks Ronen. That reg, unfortunately, is a bit vague.

I can't see why humidity and temperature extremes would affect package integrity, which is likely why they were not put on in the first place. But, until testing is completed, we can't be 100% sure... I'm afraid this could be a slippery slope which is why I'm hesitant to go down that route. i.e. different temp/humidity combinations, high/low altitude??
 

Ronen E

Problem Solver
Staff member
Moderator
#4
You could consult about the technical aspect with the manufacturer of the unsealed pouches / blisters etc.
 

planB

Trusted Information Resource
#5
dr1vn,

some items to consider:

  1. At which temperature / temperature ranges did you perform your real-time package integrity study? Did you do an accelerated study as well? What was your temperature set-point there? Both "real-time" and "accelerated" temperature will provide you with an indication maximum storage temperature to label.
  2. In shelf life / stability studies, humidity is usually a parameter you monitor, but you do _not_ set - refer e.g. to ASTM 1980- 07, which says in section 6.5:

6.5 A humidity factor to calculate the accelerated aging time (AAT) is not applicable for accelerated aging protocols. Unrealistic or extreme temperature and humidity conditions may be of interest in overall sterile barrier system performance. However,this must be evaluated in a separate study and is not related to aging of the materials. See Appendix X3 for more
details on the use of humidity in accelerated aging protocols.
So in case your package integrity is really not affected by humidity, you can tell your customer that there actually are _no_ humidity requirements, which means that everything below 100 %rH would be OK.

For temperature, "no special requirements" usually means room temperature, which typically means a max. temp of 25°C If this is also the case for your device, you could either label room temperature or a max Temp of 25°C - if this correlates with your packaging integrity study set points.

HTH,
 

Quarma

Starting to get Involved
#6
Hi,
I would base the requirement on labeling temperature and humidity range on a sterile device on
[FONT=&quot]Annex 1 section 8.3:[/FONT]
[FONT=&quot]Devices delivered in a sterile state must be designed, manufactured and[/FONT]
[FONT=&quot]packed in a non-reusable pack and/or according to appropriate[/FONT]
[FONT=&quot]procedures to ensure that they are sterile when placed on the market[/FONT]
[FONT=&quot]and remain sterile, under the storage and transport conditions laid[/FONT]
[FONT=&quot]down, until the protective packaging is damaged or opened.


[/FONT]If you do not define any e.g. temperature conditions, you are implicating, that it will remain sterile under any condition which means the sterile barrier will remain intact under any conditions. I am thinking of 2 situations: a) sales people leaving products in a car in summer, which may easily reach 75°C, b) uncontrolled unisolated warehouses which heat up in summer . So i believe, a temperature range on the product has to be given. [Otherwise I expect proof, that the packing withstands conditions - which ? high temperatures ? .]

Do the manufacturers of blister packing provide guiding values for that, or what are common values used ?
I do not see that there is a room temperature symbol for packaging.

I hope this helps the discussion[FONT=&quot]
[/FONT]
 
Thread starter Similar threads Forum Replies Date
J Sterile medical device expiration date 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
R Deciding if a medical device shall be supplied sterile or not Other Medical Device and Orthopedic Related Topics 5
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
D What is your Sterile Medical Device Pouch Peel Strength Spec? Other Medical Device Related Standards 3
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
N Medical device non sterile packaging requirement CE Marking (Conformité Européene) / CB Scheme 1
N Registering a Class 1 Sterile Medical Device - Currently selling a Non-Sterile 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
3 Egypt Medical Device Registration: Only required for Sterile Medical Devices? Other Medical Device Regulations World-Wide 10
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
F Date of Manufacture and Expiration (Sterile Medical Device) question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
T Sterile Medical Device Storage Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
J 510k file requirements for non-sterile medical device accessories 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
J Non-Sterile Class 2 Medical Device Shelf Life Other Medical Device Related Standards 3
R Shipping Stability for Class III IVD (Non Sterile) Medical Device Other Medical Device Related Standards 7
J0anne MOC - Sterile Medical Device Marketing - India Other Medical Device Regulations World-Wide 1
J Sterile Medical Device Labeling Requirments - ISO 1041 Other Medical Device and Orthopedic Related Topics 1
J Biocompatibility for Class I Non-Sterile Medical Device EU Medical Device Regulations 2
M Distribution Center has shipped Sterile Medical Device products that have Expired EU Medical Device Regulations 2
H Medical Device RUO (Research Use Only) and 'Sterile' Label Claims 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
N Symbol for Medical Device Sterile Packaging - "Inspect Seal Before Opening" Other Medical Device Related Standards 5
Q Expiry Dates on a Non-Sterile Class 1 Medical Device (5 Year Shelf Life). Other Medical Device and Orthopedic Related Topics 11
A Need to register with CA for Class I (Non-Sterile and Non-Measuring) Medical Device EU Medical Device Regulations 4
M Process Validation/Equipment Qualification - Class 1A (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 12
M Shelf Life Testing of a Sterile Medical Device Other Medical Device and Orthopedic Related Topics 40
M ISO 14971:2007 Risk Management - Class I A Sterile Medical Device ISO 14971 - Medical Device Risk Management 4
S Sterile Product Ship Testing (Medical Device Packaging) EU Medical Device Regulations 3
J Addition of a Class 1 (Sterile) Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 3
D Selling Class IIa medical device (sterile syringes) in USA ISO 13485:2016 - Medical Device Quality Management Systems 2
A Documentation required for launch of a non-sterile exempt class 1 medical device ISO 13485:2016 - Medical Device Quality Management Systems 7
S Environmental Control Standards - Sterile medical device manufacturing environment ISO 13485:2016 - Medical Device Quality Management Systems 3
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
J Use of expired perminant implantable sterile medical products ISO 13485:2016 - Medical Device Quality Management Systems 2
S Medical Implant Sterile Packaging Design EU Medical Device Regulations 6
P Sterile Medical Packaging and Shelf Life Validation Sample size ISO 13485:2016 - Medical Device Quality Management Systems 10
S Do Medical Sterile Drapes count as MOPP? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
M Packaging Requirements for Non-Sterile Class II Medical Devices EU Medical Device Regulations 11
Q Acceptable Visible Particulate Size on Sterile Medical Devices Inspection, Prints (Drawings), Testing, Sampling and Related Topics 13
C Cleanliness Specifications that apply to non-sterile medical devices ISO 13485:2016 - Medical Device Quality Management Systems 12
W Distributor Requirements for Sterile Disposable Medical Devices EU Medical Device Regulations 8
J Medical Devices: Initial Design - Sterile or Non-sterile Design and Development of Products and Processes 10
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
D Standard guide for shelf life determination of non sterile component with degradation materials Reliability Analysis - Predictions, Testing and Standards 1
M GSPR 23.3 (h) Manufacturing date on sterile barrier EU Medical Device Regulations 0
M MDR 2017/745 sterile and non-sterile labeling EU Medical Device Regulations 12

Similar threads

Top Bottom