"Sterile" missing on the CE Certificate - ISO 13485 Certified Contract Manufacturer

NicoleinFlorida

Inactive Registered Visitor
#1
Hello the cove,

I have question about the description on CE/ISO Certificate.

We have ISO 13485 Certified Contract Manufacturer that provide sterile medical device for use. Our consultant noticed that the ISO/CE certificate fails to specify "Sterile". When I talked with the representative of Contract Manufacturer, the lady told me that the NB audited the ISO Certificate of the Sterilizer and also products's sterilization validation report. But she was unaware that the "sterile" is required on the certificate.

I am not very familiar with the NB certificate process. And after googled for a while, I haven't get any clue about the requirement of "Sterile" for class IIb product. Is this required for the certificate if the contract manufacturer provides the sterile device for us? And how is the auditing process for the sterile?

Really appreciate it.
 
Elsmar Forum Sponsor

bjohnsonrli

Starting to get Involved
#2
Re: "Sterile" missing on the CE Certificate

Hi Nicole,

I am no expert in this matter, however what I have seen on our own contract sterilizer's ISO cert includes "sterilization services" in the scope. I would think that sterilization should be listed as one of their registered activities. Hope this helps.

Regards,

Bryan
 

Ronen E

Problem Solver
Staff member
Moderator
#3
Re: "Sterile" missing on the CE Certificate - ISO 13485 Certified Contract Manufactur

Hello the cove,

I have question about the description on CE/ISO Certificate.

We have ISO 13485 Certified Contract Manufacturer that provide sterile medical device for use. Our consultant noticed that the ISO/CE certificate fails to specify "Sterile". When I talked with the representative of Contract Manufacturer, the lady told me that the NB audited the ISO Certificate of the Sterilizer and also products's sterilization validation report. But she was unaware that the "sterile" is required on the certificate.

I am not very familiar with the NB certificate process. And after googled for a while, I haven't get any clue about the requirement of "Sterile" for class IIb product. Is this required for the certificate if the contract manufacturer provides the sterile device for us? And how is the auditing process for the sterile?

Really appreciate it.
Essentially when a manufacturer applies to be certified by a NB they include the product scope in the application. Once this is agreed with the NB, this is the scope the NB will audit and certify to. I guess that they didn't include the "sterile" in their application, for some reason (maybe inadvertently).

It sounds as if the NB did audit them as a sterile devices manufacturer, otherwise they wouldn't have audited the sterilisation validation etc. However, it's better to be safe than sorry, so I would ask your CM if they could apply with their NB for a change of scope based on the audit that already took place. If indeed it's a matter of oversight, the NB wouldn't have a problem fixing the certificate. However, if they have indeed only certified the non-sterile production it will require re-certification. In the latter case it'd actually be a good thing that you brought it up.

Cheers,
Ronen.
 

Sidney Vianna

Post Responsibly
Staff member
Admin
#4
Re: "Sterile" missing on the CE Certificate - ISO 13485 Certified Contract Manufactur

Is this required for the certificate if the contract manufacturer provides the sterile device for us? And how is the auditing process for the sterile?
If your organization really wants the contract manufacturer (CM) to only deliver sterile medical device products, this obviously should be stated in your purchasing documents with the supplier and, if the CM/Supplier only delivers sterile products, their ISO 13485 certificate should mention this key aspect.

Also, you can familiarize yourself with the NBOG’s Best Practice Guide document concerning the certificates issued by NB's. Download the document and search for the 11 instances of the word sterile in it.
 

NicoleinFlorida

Inactive Registered Visitor
#5
Re: "Sterile" missing on the CE Certificate - ISO 13485 Certified Contract Manufactur

Thank you for pointing out about the application scope. I will contact with the CM about this.

As part of our contract, we require our CM to have ISO certificate for the our products, which means Sterile is mandatory.
 

NicoleinFlorida

Inactive Registered Visitor
#6
Re: "Sterile" missing on the CE Certificate - ISO 13485 Certified Contract Manufactur

According to the contract, the CM will provide Sterile Medical Devices for us. Per supplier qualification, they must be ISO 13485 Certified for the manufacturing of our products.

Thank you for the document :)
 
Thread starter Similar threads Forum Replies Date
K Sterile barrier system symbols - Label redlines for a Class III device (Kit Box) EU Medical Device Regulations 0
C Where to place CE/label in Sterile IVD EU Medical Device Regulations 1
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
J Sterile medical device expiration date 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 7
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3
Y Packaging validation for non-sterile Medical Equipment Other Medical Device Related Standards 1
A Single use non-sterile syringe used in the oral cavity - Laboratory test advice US Food and Drug Administration (FDA) 7
D Standard guide for shelf life determination of non sterile component with degradation materials Reliability Analysis - Predictions, Testing and Standards 1
M GSPR 23.3 (h) Manufacturing date on sterile barrier EU Medical Device Regulations 0
M MDR 2017/745 sterile and non-sterile labeling EU Medical Device Regulations 12
R Labels on sterile packaging EU Medical Device Regulations 5
T No Defined Shelf Life/Expiration Date - Disposable single-use, non-invasive, non-sterile Other Medical Device Regulations World-Wide 2
Z Class 2 medical device - Process Validation - Test sterile or non-sterile units? Qualification and Validation (including 21 CFR Part 11) 4
R Deciding if a medical device shall be supplied sterile or not Other Medical Device and Orthopedic Related Topics 5
N Per the FDA regulations can a medical device be class I despite that it is sterile? Other Medical Device Related Standards 4
S Package performance testing for non-sterile Hydrogels US Food and Drug Administration (FDA) 1
D What is your Sterile Medical Device Pouch Peel Strength Spec? Other Medical Device Related Standards 3
E Cleaning Validation Process of Non-Sterile Implants and Instruments Other Medical Device Related Standards 2
R Use of the "Non-Sterile" Icon - ISO7000-2609 EU Medical Device Regulations 4
Y Sterile device packaging under MDR EU Medical Device Regulations 3
A Medical Device Class 2A, non sterile bodyworn packaging requirements CE Marking (Conformité Européene) / CB Scheme 7
N Medical device non sterile packaging requirement CE Marking (Conformité Européene) / CB Scheme 1
A Sterile Garbage Bag for Clean Room Manufacturing and Related Processes 6
F Bioburden Testing requirements for Non-Sterile Dental Products EU Medical Device Regulations 6
N Registering a Class 1 Sterile Medical Device - Currently selling a Non-Sterile 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
Y Shelf Life of Non-Sterile Disposable Drug Inhaler 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
3 Egypt Medical Device Registration: Only required for Sterile Medical Devices? Other Medical Device Regulations World-Wide 10
shimonv Over-Labeling Sterile Packaging - Shelf Life 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
D Ethylene Oxide Sterile Implantable Medical Device Shelf Life Extension Other Medical Device Related Standards 1
P ISO 10993 - Sterile (Irradiation) Polypropylene Syringe Question Other Medical Device Related Standards 6
K AQL for Sterile Product - Laminate Tubes 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
F Date of Manufacture and Expiration (Sterile Medical Device) question 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
shimonv Packaging Validation for Non-Sterile Package EU Medical Device Regulations 5
D Sterile Medical Device Temperature and Humidity (Storage) Labeling Requirements EU Medical Device Regulations 5
R Temperature Profile for Transport Simulation (Sterile Eye Drops) Other Medical Device Related Standards 1
I Use of the Non-Sterile Symbol (EN980/ISO7000/ISO15223-1) Other Medical Device Related Standards 3
D Class II Sterile Product Shelf Life Labeling - Should we do a recall? Nonconformance and Corrective Action 4
V Are microbial limit tests required for sterile products? Qualification and Validation (including 21 CFR Part 11) 3
T Sterile Medical Device Storage Requirements ISO 13485:2016 - Medical Device Quality Management Systems 5
J 510k file requirements for non-sterile medical device accessories 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
J Non-Sterile Class 2 Medical Device Shelf Life Other Medical Device Related Standards 3
O Bioburden Testing - Non Sterile Product / Tapes - ISO 11737-1 Other ISO and International Standards and European Regulations 9
S ECH (ethylene chlorohydrin) residues - Plastic catheter packed in a sterile bag Other Medical Device Related Standards 5
B External Label Content of multi-device (sterile and nonsterile) EU Medical Device Regulations 2
J Grouping Products and Testing Sterile and Biocompatibility - Validating Products Other Medical Device Regulations World-Wide 3
J Change to Outer Packaging Material/Supplier - Sterile, Class IV Device Other Medical Device Related Standards 4
J Use of expired perminant implantable sterile medical products ISO 13485:2016 - Medical Device Quality Management Systems 2
R Shipping Stability for Class III IVD (Non Sterile) Medical Device Other Medical Device Related Standards 7
J0anne MOC - Sterile Medical Device Marketing - India Other Medical Device Regulations World-Wide 1

Similar threads