I would like feedback on the following scenario:
For e-beam sterilization, it is acceptable to do a verification dose with sterilization provider -X- and then, the dose mapping with another sterilization provider -Y- with the intention to use sterilization provider -Y- for routine sterilization processing.
I would like feedback on the following scenario:
For e-beam sterilization, it is acceptable to do a verification dose with sterilization provider -X- and then, the dose mapping with another sterilization provider -Y- with the intention to use sterilization provider -Y- for routine sterilization processing.
Absolutely yes. That is actually the typical scenario. Verification doses have a very tight tolerance (+/- 10% I think) that cannot typically (for gamma at least) be met in a typical production process. Also, all verification doses must include their own dose mapping.
In your example, provider Y is your production provider and they would need to perform dose mapping in support of your routine processing with them. This is typically done initially and only updated if you change your load configuration.