Hello Everyone,

I have a client that has a piece of plastic (PA-46) that will be attached to a surgical tray. The client (hospital clinical nurse) would like us to prove 1. the plastic can be sterilized and 2. the plastic will not harbor bioburden after sterilization. We talked to a lab who quoted us several tests at 100 cycles per test. This seems like overkill provided we were told industry standard is cycle of 1.

My questions are:

1. Does anybody have a recommended lab to perform sterility and bioburden testing?
2. For sterilization/sterile processing of med dev in the hospital setting, which authority or body sets the acceptable level of bioburden after steam sterilization?

Thank you for all of your help in advance!

L
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#4
Hello Everyone,

I have a client that has a piece of plastic (PA-46) that will be attached to a surgical tray. The client (hospital clinical nurse) would like us to prove 1. the plastic can be sterilized and 2. the plastic will not harbor bioburden after sterilization. We talked to a lab who quoted us several tests at 100 cycles per test. This seems like overkill provided we were told industry standard is cycle of 1.

My questions are:

1. Does anybody have a recommended lab to perform sterility and bioburden testing?
2. For sterilization/sterile processing of med dev in the hospital setting, which authority or body sets the acceptable level of bioburden after steam sterilization?

Thank you for all of your help in advance!

L
Hi,

"1. the plastic can be sterilized" can have two meanings:
(a) Sterility of that piece is achievable
(b) The plastic piece will not be harmed (get out of spec) by the cycle necessary for achieving sterility

(a) is thoroughly covered by published standards, and you definitely don't need 100 cycles for that. (b) might require repeated cycles, the number of which depends on the intended use (how many cycles the part is intended for).

"2. the plastic will not harbor bioburden after sterilization" is mostly a function of how they package it for sterilisation and how they handle it afterwards. It's a part of the reprocessing validation, which is also throughly covered by published standards.

The "authority or body that sets the acceptable level of bioburden after steam sterilization" would be the hospital itself, and that level is also supported by published standards.
 

Ajit Basrur

Staff member
Admin
#5
Hello Everyone,

I have a client that has a piece of plastic (PA-46) that will be attached to a surgical tray. The client (hospital clinical nurse) would like us to prove 1. the plastic can be sterilized and 2. the plastic will not harbor bioburden after sterilization. We talked to a lab who quoted us several tests at 100 cycles per test. This seems like overkill provided we were told industry standard is cycle of 1.

My questions are:

1. Does anybody have a recommended lab to perform sterility and bioburden testing?
2. For sterilization/sterile processing of med dev in the hospital setting, which authority or body sets the acceptable level of bioburden after steam sterilization?

Thank you for all of your help in advance!

L
I am unclear on your question as sterility and bioburden are different concepts.

When you "sterilize" an article, it has to be "free from viable micro-organisms" and proved by a sterility test so there is no "bioburden". Typically all articles intended to be used as sterile undergo this sterilization process and should not have any microbial load.

You can refer the following thread for more info:

Sterility Testing vs Bioburden Testing - Differences
 
Thread starter Similar threads Forum Replies Date
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
M Sterility Testing vs Bioburden Testing - Differences Qualification and Validation (including 21 CFR Part 11) 7
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
B Sterility test on gloves ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical Devices Sterility - Question from A Regulator Other Medical Device Regulations World-Wide 3
O Freeze-Dried Powder Sterility Test using the Membrane Filtration Technique Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
M How many Pieces to Test - Sterility Testing USP EU Medical Device Regulations 2
B IPCD Sterility test failure during MPQ (EO sterilization validation) ISO 13485:2016 - Medical Device Quality Management Systems 2
N Sterility Testing - USP Chapter 71 and ISO 11737-2 ISO 13485:2016 - Medical Device Quality Management Systems 5
H ISO 13485 for Sterility Only? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Shelf Life Determination - Sterility test (Stasis test) ISO 13485:2016 - Medical Device Quality Management Systems 12
N Consultant Obligations, CFR 803 and Sterility Quality Manager and Management Related Issues 4
A Shelf Life Testing - We would like to extend the sterility expiration date Misc. Quality Assurance and Business Systems Related Topics 1
A Sterility Assurance Level? Special 510k US Food and Drug Administration (FDA) 5
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
M GLP (Good Laboratory Practices) Requirement for Sterility Validations Qualification and Validation (including 21 CFR Part 11) 1
J Container Closure Integrity & Sterility Test - The Dye Immersion Method Quality Assurance and Compliance Software Tools and Solutions 16
B SAL (Sterility Assurance Level) for High Level Disinfection vs. Sterilization US Food and Drug Administration (FDA) 7
S Sterility - Class I or Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Suitable Sterilization Cycle to achieve Sterility Assurance Level of <10-6 ISO 13485:2016 - Medical Device Quality Management Systems 1
D Sterility (bacteriostasis/fungistasis) Testing by Membrane Filtration validation Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
S Sterility Assurance Level of Liquids (SAL) - Special Culture Medium ISO 13485:2016 - Medical Device Quality Management Systems 3
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3
B Radiation Sterilization Dose Audit: Bioburden frequency Other Medical Device Related Standards 1
F Bioburden Testing requirements for Non-Sterile Dental Products EU Medical Device Regulations 6
V Cleanliness/Bioburden for Non-Sterile Gas Path Medical Devices Other Medical Device Regulations World-Wide 7
K Bioburden 0 cfu for Dose Setting using the VDmax25 Other Medical Device Related Standards 9
N Must Bioburden Sampling Procedures be Documented? Other Medical Device Related Standards 5
R Simulated product for bioburden testing Other ISO and International Standards and European Regulations 2
C What is the most appropriate Bioburden Test Method for Surgical Gloves ISO 13485:2016 - Medical Device Quality Management Systems 10
M Bioburden numbers and ECA gloves 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
M Bioburden testing for Class 10000 (Grade C) clean room ISO 13485:2016 - Medical Device Quality Management Systems 1
O Bioburden Testing - Non Sterile Product / Tapes - ISO 11737-1 Other ISO and International Standards and European Regulations 9
P Serial Testing of Bioburden for Medical Devices Other Medical Device Related Standards 2
M Biological Monitoring Program Environmental Bioburden Test Method Needed ISO 13485:2016 - Medical Device Quality Management Systems 1
C What is an acceptable bioburden level or count when working in cleanroom ISO class7 ISO 13485:2016 - Medical Device Quality Management Systems 1
P Frequency of Bioburden Testing EU Medical Device Regulations 3
B BioBurden and EndoToxin Questions Other Medical Device Related Standards 6
S Bioburden Monitoring Sample Size Increase ISO 13485:2016 - Medical Device Quality Management Systems 4
I Bioburden and Endotoxin (LAL) Testing proposal help 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
I Bioburden Count Question - Bioburden is above the alert limit >1 CFU/component avg. 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
J Bioburden Monitoring - 'Off the Shelf' Bioburden/Microbiological Monitoring Software ISO 13485:2016 - Medical Device Quality Management Systems 4
L GB15980-2003 Bioburden Count Trend on Operator Hands Other ISO and International Standards and European Regulations 1
0 Guideline/Method for Raw Material Roll Stock Bioburden Sampling Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
Q Bioburden Requirements on Device used on Unbroken Skin Other Medical Device and Orthopedic Related Topics 4
A Microbial Limit (Bioburden)on dry nonwoven that used for making Hand sanitizing wipes ISO 13485:2016 - Medical Device Quality Management Systems 1
D Bioburden and Product Labels - VD Max studies on all the BOMs Other Medical Device and Orthopedic Related Topics 9
H Determining a New Product Bioburden Limit ISO 13485:2016 - Medical Device Quality Management Systems 11
Similar threads


















































Top Bottom