What is the difference between the above two?
Bioburden testing:
Bioburden test is performed to determine the number and nature of microorganisms on a product prior to sterilization.
According to the standard ISO/TS 11139:2006 (Sterilization of health care products ? Vocabulary), the definitions of sterility, test for sterility & bioburden are listed below;
Sterility:
State of being free from viable micro-organisms. [ISO/TS 11139:2006]
Bioburden:
Population of viable micro-organisms on a product and/or sterile barrier system. [ISO/TS 11139:2006]
Test for sterility:
Technical operation defined in an official pharmacopoeia performed on product following exposure to a sterilization process [ISO/TS 11139:2006]
Test of sterility:
Technical operation performed as part of development, validation or requalification to determine the presence or absence of viable microorganisms on product or portions thereof. [ISO/TS 11139:2006]
mobylad
Starting to get Involved
#3
So if the both are measuring for bioburden? Do they use the same test method? except one (i.e. the sterility test)is expecting a zero bioburden result?
Ajit Basrur
Staff member
Admin
#5
Yes the intent of both tests is the same to check for microorganisms but Sterility Testing has a wider and deeper implication.
In Sterility Testing, the intent is to find if the batch is sterile / non sterile and only 2 liquid media - Trytic Soy Broth (for aerobic organisms) and Fluid Thioglycollate Broth (for anaerobic organisms) are employed and any growth in the media indicates that the batch could be contaminated and requiring further investigation. There is lot of considerations to be given for sample size too.
Bioburden test is to assess the load of the organisms in the articles and hence different solid agar are employed - each for bacteria, mold and yeast and different objectionable articles.
Incubation periods are different in both tests.
You could also take a look in the European Pharmacopoeia or the British Pharmacopoeia for more information.
Let us know if you have additional questions :)
somashekar
Staff member
Super Moderator
#6
If I can add to Ajit's response, bioburden test is also to assess the challenge in the work environment that could have a direct bearing to the effectiveness of the product sterilization process.
mpfizer
Involved In Discussions
#7
"There is lot of considerations to be given for sample size too."
If there is a validated procedure for sterilization, is there still a need to follow the number of pcs to be tested after sterilization for sterility as per USP, i am referring to sterile medical devices and the total lot size is more than 500 numbers please advise
Thanks
michelle
Y
Yousuri
#8
Hi,
Im not sure if this is the right thread to post this to, but ill try.
I have an SOP for Gamma Sterilization of a Class I device but im lacking the following items in the procedure:
• Include bio-burden monitoring as a stand alone action.
• Add trending
• Establish action alerts and action limits for bio-burden.
Would somebody have a SOP that would have these actions in it? I just dont want to have to re-invent the wheel if its readily available?
Thanks!
Yousuri
Thread starter Similar threads Forum Replies Date
D Sterility and BioBurden Testing for Plastics ISO 13485:2016 - Medical Device Quality Management Systems 5
C Sterility and Bioburden Testing - LAL passed under 0.5EU/mL Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
M How many Pieces to Test - Sterility Testing USP EU Medical Device Regulations 2
N Sterility Testing - USP Chapter 71 and ISO 11737-2 ISO 13485:2016 - Medical Device Quality Management Systems 5
A Shelf Life Testing - We would like to extend the sterility expiration date Misc. Quality Assurance and Business Systems Related Topics 1
D Sterility (bacteriostasis/fungistasis) Testing by Membrane Filtration validation Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 2
B Determining sample size for device sterility Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
B Sterility test on gloves ISO 13485:2016 - Medical Device Quality Management Systems 3
M Medical Devices Sterility - Question from A Regulator Other Medical Device Regulations World-Wide 3
O Freeze-Dried Powder Sterility Test using the Membrane Filtration Technique Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
B IPCD Sterility test failure during MPQ (EO sterilization validation) ISO 13485:2016 - Medical Device Quality Management Systems 2
H ISO 13485 for Sterility Only? ISO 13485:2016 - Medical Device Quality Management Systems 2
S Shelf Life Determination - Sterility test (Stasis test) ISO 13485:2016 - Medical Device Quality Management Systems 12
N Consultant Obligations, CFR 803 and Sterility Quality Manager and Management Related Issues 4
A Sterility Assurance Level? Special 510k US Food and Drug Administration (FDA) 5
D ETO Sterility Requirements for FIM Clinical Study in Europe Other Medical Device Regulations World-Wide 3
M GLP (Good Laboratory Practices) Requirement for Sterility Validations Qualification and Validation (including 21 CFR Part 11) 1
J Container Closure Integrity & Sterility Test - The Dye Immersion Method Quality Assurance and Compliance Software Tools and Solutions 16
B SAL (Sterility Assurance Level) for High Level Disinfection vs. Sterilization US Food and Drug Administration (FDA) 7
S Sterility - Class I or Class II Medical Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S Suitable Sterilization Cycle to achieve Sterility Assurance Level of <10-6 ISO 13485:2016 - Medical Device Quality Management Systems 1
M Gamma Sterility Validation - ISO 11137-2, 4.2 vs. FDA Requirements ISO 13485:2016 - Medical Device Quality Management Systems 6
S Sterility Assurance Level of Liquids (SAL) - Special Culture Medium ISO 13485:2016 - Medical Device Quality Management Systems 3
M Testing criteria - where to place Design and Development of Products and Processes 1
E 60601-1 - Tilt testing - Tensile safety factor IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
V Setup for testing against ISO14708 clause 16 (protection of the patient from herms caused by heat) Other Medical Device Related Standards 0
gramps What do you think about automated QA testing For software app industry? Misc. Quality Assurance and Business Systems Related Topics 5
B In house NIOSH pre Testing accepted by NIOSH? US Food and Drug Administration (FDA) 1
M Bacteriostasis/Fungistasis Testing Other Medical Device and Orthopedic Related Topics 6
P Sample Size for Distribution Simulation Testing Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
N EN 813, EN 12277, EN 1497 - Testing some harness prototypes to an EN standard Various Other Specifications, Standards, and related Requirements 0
S What should i choose for "testing procedure" characteristics? (N95) General Information Resources 0
D Essential performance and EMC immunity testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 4
J Biocompatbility testing on Class 1 device requirements Other US Medical Device Regulations 12
Q Summative Usability Evaluation Testing: prior or during Clinical Investigation? Human Factors and Ergonomics in Engineering 6
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M ECG lead leakage currents - How to specify ECG leads during electrical safety testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
C Stress / Challenge Conditions for Design Verification Testing to Reduce Sample Size 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
D CFR Title 14: Aeronautics and Space PART 120—DRUG AND ALCOHOL TESTING PROGRAM Federal Aviation Administration (FAA) Standards and Requirements 3
lanley liao Purchase Acceptance Criteria - Tensile testing Oil and Gas Industry Standards and Regulations 2
M Device mounted at IV pole - what about mechanical stability testing? IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
A Outsourcing IEC 60601-1 Ed 3.2 Testing IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
J Cochlear Implant Testing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
A Class I (exempt) testing requirements Other Medical Device Related Standards 0
JoCam Electrical Testing for Japan, PSE or CB Scheme Other Medical Device Regulations World-Wide 0
M Who are the go to companies for non-destructive hardness testing? General Measurement Device and Calibration Topics 3
M Determining if an Insulin Pen Testing Machine is a Medical Device? EU Medical Device Regulations 4
P Testing cloud-based backups IT (Information Technology) Service Management 7
I IATF Lab Scope Testing Qualification and Competency Documentation IATF 16949 - Automotive Quality Systems Standard 3
Similar threads
Top Bottom