Sterilization Re-Validation after Manufacturer Change

Qara123

Involved In Discussions
#11
The first lot of products from the new manufacturer will be tested for bioburden as part of a routine lot release. The question is whether it's enough or would you suggest anything else?
In all likelihood, bioburden testing of the first lot of products will not be even close to the 10^6 cfu neighborhood. If an overkill method is used, there will be an even lower likelihood that the product is non-sterile. Furthur, the bacteria identified will not be bacillus atrophaeus so your bioburden test requirement should be easily met. This looks like a step towards justifying a reason for not re-validating a sterilization process that has not been changed since there is no change to validate.

If you can show that bioburden at the previous and new manufacturers are similar and are very low, and that the devices from the new manufactuer meet all release specifications (including EtO residue, & endotoxin levels) you may have a fighting chance.

Do you plan to include your justification (or sterilization validation if you proceed with re-validation) in your submission to the FDA or have you already notified FDA about the subcontractor change?

Also, you mentioned that there is no change to the packaging subcontractor. Are products sent "naked" to the subcontractor who then packs the devices in Tyvek pouches? If so, at what stage do you perform bioburden testing since your devices could easily pick up some bugs while packaging.

Q
 
Elsmar Forum Sponsor
G

gitgiv

#12
Thank you very much Qara123 for the informative (and reassuring) answer!
We already notified the FDA. For you second question - with both manufacturers - the former and the current - the catheters were/are being put in Tyveks in the clean room at the end of assembly and sent to the packaging sub-contractor for closure, so nothing has changed in terms of packaging. You're right to think that bugs can enter on the way, but it has been proven before that even then the sterilization is valid.
 

Qara123

Involved In Discussions
#13
If possible, could you let us know whether FDA requested a sterilization re-validation or FDA accepted your rationale for not re-validating?

It would be valuable (non-proprietary) information that we can add to our knowledge.

Hope you get the FDA response that you want.

Q
 
C

csterling

#14
Hello guys,

This discussion is in perfect timing for me, because we are in the same process of changing manufacturing site, but keeping the sterilization site. We have being debating if a rationale for not doing a full sterilization revalidation is enoug or not.

Has the FDA responded back to you on this issue?
It would be great to receive feedback on this.

Thanks in advanced.
 

medwise

Involved In Discussions
#15
In addition to DougTrof statement "Per ISO 11135-2, a requalification study may be required if a change has been made in materials, manufacturing location, or processing method which may impact the product bioburden resistance. The study results should determine the extent of physical or microbiological requalification required, i.e. full qualification, no requalification, or reduced microbiolocical performance qualification (fractional or half cycle exposures). It is common to do an annual validation review to ensure inadventant process changes have not occurred. " I will do bioburden test on 3 different batches produced by the new manufacturer and compare it with the old manufacturer.

In my opinion, the manufacturer of this device in your case is just a raw material supplier not a legal manufacturer. Therefore, the onus of risk management is on you as you are releasing the finished sterile product in the market. In the risk assessment, if you can establish that raw material made up of same material, design is identical, manufacturing process is similar including supplier's environmental monitoring, same packaging material with same packaging process and your ETO process is as per the validated method. Then you have a good case of not performing the re-validation for now. This should all be covered when you do you annual/ periodic validation.


Hope I have been of help.

Regards
Romit :tg:
 
M

MIREGMGR

#16
A numerically comparative bioburden study comparing two different-manufacturing-site product versions, by itself, is not a valid way to determine their comparative sterilizability.

Numerical bioburden studies do not establish the character of the bioburden...only its numbers. Some kinds of bioburden are much more difficult to sterilize than other kinds. You cannot assume, without evidence, that two manufacturing sites, likely with different material suppliers, different layouts and housekeeping procedures, different personnel, etc., have qualitatively similar bioburden environments.

Only a numerical comparison plus a comprehensive type-characterization (i.e. microbial identification and quantitative apportionment) would support a conclusion of comparable sterilizability...and that would be at least as difficult, and still less reliable, than just doing a mini-re-validation in the first place.
 
J

JulieChisholm

#17
Has your new CMO provided any guidance? I've found that some of my vendors work with many, many companies and have seen the gamut in terms of what people are doing when they make the switch. Recently I had one that provided a suggestion to mitigate risk without adhering fully to a standard.
 
I

iamtroll

#18
Has your new CMO provided any guidance? I've found that some of my vendors work with many, many companies and have seen the gamut in terms of what people are doing when they make the switch. Recently I had one that provided a suggestion to mitigate risk without adhering fully to a standard.
Along the same line of thought that Julie has taken, consider consulting with your sterilization provider. If you have a good relationship with them, they should be able to point you in the right direction as I'm sure that they have also seen many variations of re-validation or justification not to re-validate.
 
Thread starter Similar threads Forum Replies Date
N Sterilization Protocol Change in Validation Process and further impacts ISO 13485:2016 - Medical Device Quality Management Systems 1
S Sterilization validation after changing sterilization process provider Qualification and Validation (including 21 CFR Part 11) 3
B Sterilization Validation Plan Other Medical Device Related Standards 3
H EO Sterilization Validation - Sterility Testing and Load Configuration Other Medical Device Related Standards 1
A Reprocessing, Cleaning, Disinfection and Sterilization Validation EU Medical Device Regulations 4
G Pyrogen Test - Sterilization Validation - MDR 2017/745 Other ISO and International Standards and European Regulations 6
F Irradiation Sterilization Validation Maximum Dose - VDmax 25 method (ISO 11137) EU Medical Device Regulations 2
E Ethylene Oxide Sterilization Validation-Reference Load Other Medical Device Related Standards 2
Q Use of parts in gamma sterilization validation - Reprocessing implants EU Medical Device Regulations 1
R ISO 11137 Gamma sterilization validation, dose setting, dose audit Other Medical Device Related Standards 1
S Requirement for Sterilization Validation completion before issuance of ISO cert ISO 13485:2016 - Medical Device Quality Management Systems 0
A Is Aeration Process in ETO Sterilization Validation Required? Manufacturing and Related Processes 7
B Product definition for Ethylene Oxide Sterilization Validation Other Medical Device Related Standards 2
K Re-validation of EtO sterilization after change of tape Other Medical Device and Orthopedic Related Topics 1
B IPCD Sterility test failure during MPQ (EO sterilization validation) ISO 13485:2016 - Medical Device Quality Management Systems 2
J Sterilization Validation Procedure Needed? ISO 13485:2016 - Medical Device Quality Management Systems 3
Sam Lazzara Ethylene Oxide Sterilization Validation question Other Medical Device Related Standards 11
L Regulatory Submittal During 3 Lot Sterilization Validation EU Medical Device Regulations 3
S Fractional cycle fail in Eto sterilization validation ISO 13485:2016 - Medical Device Quality Management Systems 3
L Steam Sterilization Validation per ANSI ST 79 Other Medical Device Related Standards 2
suildur Sterilization Validation of a Unique Medical Device Other Medical Device Related Standards 5
S Sterilization Validation Accreditation - A mandated standard requirement? ISO 13485:2016 - Medical Device Quality Management Systems 5
K Re-validation of Out Sourced Medical Device Sterilization Process ISO 13485:2016 - Medical Device Quality Management Systems 3
C Validation of Sterilization Procedures - ISO 13485 Section 7.5.2.2 Document Control Systems, Procedures, Forms and Templates 4
Y 3rd Party Sterilization Validation for Reusable Medical Devices Other Medical Device Related Standards 7
S Steam Sterilization Validation Other Medical Device and Orthopedic Related Topics 4
T Missed Quarterly Dose Audit! Sterilization Validation of Medical Devices Other Medical Device and Orthopedic Related Topics 15
L Sterilization validation for reusable medical device class II Other Medical Device Related Standards 9
E Design for Sterilization Companies - The design of sterilization validation programs ISO 13485:2016 - Medical Device Quality Management Systems 12
M Sterilization Validation request for 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
S Validation of Sterilization Expiry Date - Accelerated Aging, etc ISO 13485:2016 - Medical Device Quality Management Systems 6
S Gamma Sterilization Validation Documents - 13485:2003 section 7.5.1.2.1 ISO 13485:2016 - Medical Device Quality Management Systems 5
C Sterilization Validation ISO 14937 - Sterilizer that uses EO gas diffusion Qualification and Validation (including 21 CFR Part 11) 7
P ETO Sterilization Validation Process and an NCR - ISO 13485:2003 CMDCAS audit ISO 13485:2016 - Medical Device Quality Management Systems 1
K Iso 11135:1994 - Validation and Control of Ethylene Oxide Sterilization ISO 13485:2016 - Medical Device Quality Management Systems 5
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
chris1price Sterilization using beta radiation Other Medical Device Related Standards 2
A ISO 13485 Sterilization Clause Applicability ISO 13485:2016 - Medical Device Quality Management Systems 7
S Orthopedic Implants (Knee system) Sterilization Indicators Manufacturing and Related Processes 6
chris1price Contract Sterilization Service Providers ISO 13485:2016 - Medical Device Quality Management Systems 5
R Scientific Operating Procedures For Sterilization Practices in India PSA/2012/1 Other Medical Device Regulations World-Wide 0
N Looking for Gamma sterilization facility in China ISO 13485:2016 - Medical Device Quality Management Systems 4
I Release checklist EO sterilization process Medical Device and FDA Regulations and Standards News 3
MrTetris Are GLP required for testing cytotoxicity and soil remaining after sterilization of MD? Other Medical Device Related Standards 8
Z EN 868-8 "required tests" - Rigid sterilization containers Other Medical Device Related Standards 0
Judy Abbott Increase the maximum allowable dose for E-beam sterilization Reliability Analysis - Predictions, Testing and Standards 2
W EO sterilization and pallets Other Medical Device and Orthopedic Related Topics 4
M Informational Device Shortages Update – The US FDA Announces Two New Innovation Challenges on Device Sterilization Medical Device and FDA Regulations and Standards News 0
W EO sterilization - what about the pallet? Other Medical Device Related Standards 1
B Sterilization of a custom medical device Other Medical Device and Orthopedic Related Topics 1

Similar threads

Top Bottom