Stop Watch Calibration Reality Check

Marc

Fully vaccinated are you?
Leader
#1
I thought this was good. You can guess the gist of the previous 'advice' on calibrating a stopwatch....

---------snippo---------

Date: Thu, 4 Feb 1999 09:00:00 -0500
From: "Klouse Kenneth P." KlouseKP MSHA.gov
To: Greg Gogates
Subject: RE: Re1: Stop Watch Cal (fwd)

Unless I'm missing something, anyone who spends more than 15 minutes calibrating a stop watch, or claims accuracies of .001%, better smell what their peddling, get a life, or start measuring the reflexes of the stop watch operators. I'm not saying that stop watches so not need to be calibrated, I'm saying that one should look at the accuracy needed for the measurement, and calibrate the instrument accordingly. As Jeff Gust so eloquently put it, " As you would note, the random error from the person dominates the uncertainty budget, and in fact is 41 times greater than that of the unit under test tolerances. " If that person hasn't had his coffee that morning or he didn't get enough sleep, his reaction time could be all over the map.

As an assessor of product certification schemes, it is enlightening to read responses from different people on these varied subjects. But sometimes I think we have to step back and see the practicality of what we are doing or requiring.
 
Elsmar Forum Sponsor

Marc

Fully vaccinated are you?
Leader
#2
I sorta figured this also could fit here. Another reality check. What if there is no 'standard'?

----------snippo---------

Date: Sat, 20 Feb 1999 05:54:03 EST
From: Rcveale
To: iso25
Subject: Re: Linear Displacement Potentiometer

Pierre,

There are no American nor ISO standards for linear displacement potentiometers. There might me a military standard that I am not aware of.

A group was formed several yars ago to write a standard for linear transducers, and a draft was written; however, it was never finished nor published as a standard. A couple of years ago there was an attempt to revive the Group, but there appeared to be little interest or need for such a standard.

If a need can be shown to exist and someone is willing to put together a group to do the work, there is a procedure in place for producing the standard.

Ralph Veale, Chairman
ASME B89.1 Subcommittee for Length
 
K

Ken K

#3
Ah yes...stopwatches and timers:notme:

We have 8 timers and 4 stopwatches whose calibrations are due shortly. In the past we have just tossed them into the garbage and replaced them. During our last audit, the auditor asked why we don't just use the internet to calibrate them. OK...I'll bite:rolleyes:

So, I'm looking at the calibration certificates that came with them. It states:

The maximum error of this Digital Timer at the time of calibration did not exceed 0.01%.

As measured: +/- 0.367 seconds per 24hrs.

Expanded measurement uncertainty at K=2: +/- 0.13 seconds per 24hrs.


So then I went to thise site: http://www.time.gov/timezone.cgi?Central/d/-6/java

I assume this is the site the auditor meant. So now what? Check each one against "The Clock"? For how long? How would one document the results? How could one prove this was actually performed without a certificate?

;)
 

BradM

Leader
Admin
#4
Probably there won't be a certificate. But you could accept World Time as a "Naturally Accepted Standard".

Verifying stopwatches is an interesting concept. One thing I have seen is a test rig. This "starts" the timers and the standard timer at the same time. It does minimize some of the human error.

BTW... Has anyone ever seen a stopwatch be wrong ? :D
 
#7
If you try and calibrate a timer using the internet wouldn't you have to consider the speed of the internet connection as well as the computer speed?
How can you verify what you are seeing on the internet is really accurate?
 
S

skappesser

#8
You are welcome Brad.
Most of the cal labs I have worked in over the years had plenty of homegrown cal procedures for these devices, some of which were pretty good and others that were pretty hokey. If I remember correctly, our uncertainty for this measurement was like 0.3 or 0.4 seconds (K=2), mainly due to the human reflex component. Some procedures had electronic pick-ups attached to the stopwatch circuitry - gating electronic counters. I never failed an electronic stopwatch. There was an old mechanical timer I failed once..."back in the day".
When I heard that NIST published this one, I encouraged the large commercial calibration company I worked for at the time to adopt it. All the calibration techs really liked it.
 
S

skappesser

#9
If you try and calibrate a timer using the internet wouldn't you have to consider the speed of the internet connection as well as the computer speed?
How can you verify what you are seeing on the internet is really accurate?
The NIST procedure I attached earlier addresses your question nicely...
 

Hershal

Metrologist-Auditor
Trusted Information Resource
#10
The NIST guide provided has a lot of good information, and is written by well known Metrology professionals.

I have assessed timer calibrations using GPS-sync'd time bases and that often results in MU under 0.1 seconds.

Hershal
 
Thread starter Similar threads Forum Replies Date
C Calibration Of Stop Watch using the NIST Clock General Measurement Device and Calibration Topics 4
J Machinery Directive : Emergency stop clause : definition of 'portable' or 'hand-guided' machinery Other ISO and International Standards and European Regulations 0
H Should we stop inspecting a part if its never been rejected? Reliability Analysis - Predictions, Testing and Standards 6
qualprod Ideas to stop bad practices in daily work ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 13
G Bad Parts cause Customer line stop IATF 16949 - Automotive Quality Systems Standard 13
M Supplier Audits - Stop Wasting Everyone's Time! General Auditing Discussions 24
qualprod Records where apply and stop recording everything? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 4
D Where does "as far as possible" stop? FMEA - EN 14971 ISO 14971 - Medical Device Risk Management 29
L Clause 0.4 of ISO 9001 and EHS - Where should I stop the inclusion of EHS in my QMS ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
Marc A 20 Hour Non-Stop Flight - (October 2019) Travel - Hotels, Motels, Planes and Trains 19
M Informational FDA order – All manufacturers of surgical mesh intended for transvaginal repair of anterior compartment prolapse (cystocele) to stop selling and distr Medical Device and FDA Regulations and Standards News 0
N What subcommands can be used with "Stop" in Minitab 18? Using Minitab Software 8
P Terms, "Yard Holds", and "Stop Ships" in TS 16949 IATF 16949 - Automotive Quality Systems Standard 1
smryan Excel Printing Glitch - How to STOP Printing One Cell per Page Excel .xls Spreadsheet Templates and Tools 10
Anerol C Procedure to stop a machine due to safety Occupational Health & Safety Management Standards 4
N How to Stop Packaging Choice Mistakes on a Form Other Medical Device Related Standards 6
Z When to stop Post Market Surveillance 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
C ISO 14001 Emergency Stop Information ISO 14001:2015 Specific Discussions 4
S How many production runs will it take for SPC monitoring to stop? Statistical Analysis Tools, Techniques and SPC 10
Marc Ray Harryhausen - The Stop Motion Animation King Coffee Break and Water Cooler Discussions 4
N How to stop this abuse..... Imported Legacy Blogs 12
F When TSA stop to issue new Foreign FAA Part 145 Repair Station? Federal Aviation Administration (FAA) Standards and Requirements 1
L When does Medical Device New Product R&D stop and Design start? Design and Development of Products and Processes 6
B Delay in Payment Posting - Customer would like to stop paying monthly amortization Customer Complaints 5
B Can we stop Customer Surveys as they are generating low return rates ? Customer Complaints 16
G Customer Property Cl. 7.5.4 - Where does ISO 9001 stop and ISO 27001 start? ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 33
Marc Stop Online Piracy Act (SOPA) and the PROTECT IP Act (PIPA) After Work and Weekend Discussion Topics 21
1 Stop Ship/Build or something like a Quality Alert template or procedure example Document Control Systems, Procedures, Forms and Templates 5
Marc What happens when computers stop shrinking? After Work and Weekend Discussion Topics 3
Anerol C What protocol do you follow to stop the production line? Nonconformance and Corrective Action 9
P Regulations concerning Calibration of Stop Watches used for Testing Medical Devices 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
J Press Brake Tooling / Back Gauging - Holes for the back stop instead of the Part Edge Manufacturing and Related Processes 8
D Process Flowchart - Can the Flowchart of one Process have two END (Stop) Points Process Maps, Process Mapping and Turtle Diagrams 5
Wes Bucey It's Friday - don't stop the job hunt! Career and Occupation Discussions 2
Stijloor Stop Forum Spam - Helpful site? After Work and Weekend Discussion Topics 1
D Production start after stop IATF 16949 - Automotive Quality Systems Standard 12
Sidney Vianna Aerospace auditors are told to stop soft grading NC's. Or else! AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 42
D Definition Stop Gap - Definition - What does "stop gap" mean? Definitions, Acronyms, Abbreviations and Interpretations Listed Alphabetically 9
M Can I Stop the Internal Audit Program? Internal Auditing 21
Manix Ideas On Wiring Configuration Control - Help STOP a Non Conformance! Manufacturing and Related Processes 7
Sidney Vianna Electronic wedge brakes will change the way we stop World News 6
ScottK Foreign Languages in the workplace - can you stop it? Career and Occupation Discussions 14
C First Audit Stop - Employees Washroom - Learn all you need to know about a company General Auditing Discussions 83
Marc Will the weather service be forced to stop giving weather information for free? World News 4
J IATF 3th auditor training course stop? Training - Internal, External, Online and Distance Learning 2
A Supplier 8D does not stop the problem - Critical defects - How to make 8D effective? Supplier Quality Assurance and other Supplier Issues 4
Sidney Vianna ISO/IEC standard for "one-stop accreditation" to boost cross-border trade ASQ, ANAB, UKAS, IAF, IRCA, Exemplar Global and Related Organizations 2
S Huh? What is "Stop Shipment Notification" IATF 16949 - Automotive Quality Systems Standard 6
N Why do 'we' and ?our? QMS?s stop working? What are the Failure Modes? Misc. Quality Assurance and Business Systems Related Topics 62
M High Cpk's and Control Charts - Is it justified to stop using control charts? Capability, Accuracy and Stability - Processes, Machines, etc. 7

Similar threads

Top Bottom