Study Type 1 on a CMM using a measuring support

#1
Hello Everyone.

I need to perform a Study Type 1 on a CMM using a measuring support.

First step measure 1 part 15x on a CMM without moving the part to obtain the reference value. Then measure the same part at least 50x or more times, moving the part in the measuring support.

Question:

This measuring is supposed to be 1 time per day totalizing 50x, moving the part from the measuring support? Or several times in the same day, moving the part from the measuring support???

Thanks in advance.
Cláudio Rosa
 
Elsmar Forum Sponsor

Miner

Forum Moderator
Staff member
Admin
#2
The Type 1 study is a repeatability study, so the key requirement is the removal and replacement of the product from the measuring support. Introducing the element of time changes this from a Type 1 study into a stability study, which answers a different question.
 
#4
The Type 1 study is a repeatability study, so the key requirement is the removal and replacement of the product from the measuring support. Introducing the element of time changes this from a Type 1 study into a stability study, which answers a different question.
Hello Miner. Other question:

I have a MSA plan with all measuring/ Control supports.

I made a study type 1 when they were new in the plant for the device acceptance. After that, should I make a study type 1 every year? Or only a control chart for the stability?

Thanks in advance.
 

Miner

Forum Moderator
Staff member
Admin
#5
While the accuracy of a measurement device can change over time due to wear (calibration) or environmental conditions (stability study), the repeatability and reproducibility typically do not change unless there has been a change to the measurement system (including methods), so there is no need to perform Type 1 R or R&R studies on specified intervals. Rather, repeat the study when, and if, a change is made to the system.
 
Last edited:
Thread starter Similar threads Forum Replies Date
Nihls MSA Study Type 1 (CMM) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
A Gage type 1 study on CMM Capability, Accuracy and Stability - Processes, Machines, etc. 2
M MSA Study Type 1 not capable. We are at the limit. And manufacturing wants to continue producing. Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
N % Tolerance - Type 1 study on the gages, then a gage R&R (ANOVA) Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 4
M Definitive answer on Type 1 vs Type 2 vs Type 3 Gage Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 0
B Minitab Type 1 Gage Study on True Position Question Measurement Uncertainty (MU) 1
W Using MPE (Mean Percentage Error) for Linearity and Type-1 study Measurement Uncertainty (MU) 1
E Bias and Type 1 Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
P What is a Type 3 Gage Study? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
G Ford - Type 1 Gage Study MSA (Measurement Systems Analysis) acceptable? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 7
H Type 1 Gage Study Interpretation Question Using Minitab Software 7
S Minitab: Type 1 Gage Study vs. Gage R&R Study Using Minitab Software 5
B NDC relates which type of MSA Study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 16
S Which type of study would be better for Solder Paste Height Measurement system Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 10
M Type 1 Gage Study - What's your opinion on this gage? Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
D Ford MSA - Ford TYPE 1 study on a Marposs post process gauge Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
J Type 1 Gage Study Excel Template Excel .xls Spreadsheet Templates and Tools 5
S 95% UCL - Attaching an excel for recovery study Statistical Analysis Tools, Techniques and SPC 0
D Low risk IVD study in the UK, do I need MHRA approval? UK Medical Device Regulations 1
M Lean: Conducting Capacity Study, calculating Cycle times on laser cutting machines Lean in Manufacturing and Service Industries 2
P MSA study for visual system with artifical inteligence Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
A Capability Study - in the beginning of your career what should you have known about the tool Quality Tools, Improvement and Analysis 11
Q Capability study with a minimum spec Statistical Analysis Tools, Techniques and SPC 8
E Use of OQ cleared batch for stability and clinical study use ISO 13485:2016 - Medical Device Quality Management Systems 0
S Study sign off question / responsibilities ISO 13485:2016 - Medical Device Quality Management Systems 3
M ISO 9001:2015 case study sample ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
M Case study help as per ISO 9001: 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 18
M Case study solution help required as per ISO 9001 : 2015 ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
M Case study resources ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 1
S A clinical performance evaluation study with an IVD product as Investagional Use product - Clinical Monitor requirements 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
T Attributes SPC study - Attributive control (Go gage) Statistical Analysis Tools, Techniques and SPC 5
S Capability or Gage R&R Study for Leak Tester? Reliability Analysis - Predictions, Testing and Standards 15
D Gage R&R Study on Load Cells - Large chemical blending tanks IATF 16949 - Automotive Quality Systems Standard 1
A CE Mark and PMCF study EU Medical Device Regulations 3
Q % Study variation low, % tolerance high - GR&R Interpretation help Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 3
A Gauge Study for Small Lots Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 5
K IATF & Attribute Study requirements IATF 16949 - Automotive Quality Systems Standard 9
T Correct MSA study for an automated camera system which makes attribute inspection Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 1
DuncanGibbons Looking for example aerospace part CAD files to be used for a case study Career and Occupation Discussions 2
L ASQ CBA biomedical auditor - CBA primer material is enough to study? ISO 13485:2016 - Medical Device Quality Management Systems 6
S Machine Setup for a CNC Machine Capability Study Capability, Accuracy and Stability - Processes, Machines, etc. 1
M Brazil forced degradation study requirements Reliability Analysis - Predictions, Testing and Standards 0
P RAC-US Spring 2020 Study group_Discussions Professional Certifications and Degrees 19
M Custom Medical Device For Clinical Study Other Medical Device Regulations World-Wide 1
G Capability Study for Tapped Hole Reliability Analysis - Predictions, Testing and Standards 3
R PMCF plan to justify that our product not require PMCF study EU Medical Device Regulations 2
E How much a bioequivalence study will cost in the UK Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 0
N MSA Study for a Leak Testing Device Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 6
D Do I need part variation while doing Destructive Variable Gage R&R MSA study Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 19
Prashant G MSA Study - AS 9100 and and our customer want us to do MSA study for their parts dimension Reliability Analysis - Predictions, Testing and Standards 4

Similar threads

Top Bottom