Submit a 510(k) with a Predicate cleared but not Listed

Berni

Starting to get Involved
#1
We have received the 510(k) clearance letter. We also have prepared the submission of the new version of the cleared device, a new 510(k) which uses our cleared device as substantially equivalent. The new device includes wireless communication and some other changes but similar intended use.
This is the device we want to bring to the US market. Our idea is to submit the new 510(k) saving the costs of the registration and listing. Do you think this is possible? I haven't found anything that makes me think that we can't do it.
 
Elsmar Forum Sponsor

Ronen E

Problem Solver
Staff member
Moderator
#2
21 CFR 807.92 (3) indicates that a device which has been found to be substantially equivalent through the 510(k) premarket notification process can serve as the legally marketed device to which a new device may be compared for a determination regarding substantial equivalence.
 

Sullen-gent

Starting to get Involved
#3
Would it matter if that device had not been used in clinical practice in the US? Is it still suitable as a predicate without any clinical use data?
 

mihzago

Trusted Information Resource
#4
There is no requirement to sell or use a device in the market after a clearance. Some companies dissolve soon after a clearance, and some use a step-wise approach to getting a clearance (start with basic claims and expand in subsequent submissions).

There's also no requirement to list a device after receiving a clearance if you do not plan to market the device.
 

Sullen-gent

Starting to get Involved
#5
21 CFR 807.92 (3) indicates that a device which has been found to be substantially equivalent through the 510(k) premarket notification process can serve as the legally marketed device to which a new device may be compared for a determination regarding substantial equivalence.
Would it matter if that device had not been used in clinical practice in the US? Is it still suitable as a predicate without any clinical use data?
 

mihzago

Trusted Information Resource
#6
It would not matter.
Other than recalls and MDRs, the FDA has no good way of knowing how a cleared device is used in clinical practice or if it's even used.
So yes, even if a cleared device has not been used in clinical practice, it is still a suitable predicate.
 

Ronen E

Problem Solver
Staff member
Moderator
#8
Would it matter if that device had not been used in clinical practice in the US? Is it still suitable as a predicate without any clinical use data?
510k submissions don’t necessarily require clinical data. The base concept is that it’s enough to show identity in intended use and technological equivalence to demonstrate substantial equivalence.
 

Berni

Starting to get Involved
#9
Thanks a lot for your answers. It's weird they don't request some kind of clinical data or device usage for a predicate statement but in my case, is a good point. We want to reach the market with the new device which substantially improves the predicate in usability.
 

Ronen E

Problem Solver
Staff member
Moderator
#10
It's weird they don't request some kind of clinical data or device usage for a predicate statement
510k's are only intended for low and moderate risk devices (classes I & II). The requirements are commensurate with the risk. Where the risk is higher (class III) clinical data is certainly required. I think it all makes sense.

We want to reach the market with the new device which substantially improves the predicate in usability.
(emphasis added)
You need to be careful - you need to ensure that:
(a) you don't deviate form the predicate's intended use (Remember: Intended use is the one implied from everything, not necessarily what one conveniently writes in their statement); and
(b) any technological differences don't raise new questions regarding safety & effectiveness.
 
Thread starter Similar threads Forum Replies Date
Ronen E FDA releases new draft guidance on when to submit a 510(k) for device modification Other US Medical Device Regulations 10
B Submit a Special 510(k) or "Documentation by our own" 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
J Can we submit data with Special 510(k)? US Food and Drug Administration (FDA) 8
C When to submit a 510(k)? After the DMR (Device Master Record) is finished? Other US Medical Device Regulations 4
A Should I submit the entire copy of my software in a 510(k) submission?? US Food and Drug Administration (FDA) 7
shimonv Deciding When to Submit a 510(k) for a Change to an Existing Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 17
sagai New Draft Guidance - When to Submit a 510(k) for a Change to an Existing Device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B Deciding when to submit a Special 510(k) US Food and Drug Administration (FDA) 10
C How to Submit an Expansion of Claims 510(k)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 9
N Can I submit 510(k) without Biocompatibility Test Results? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C When to Submit a 510(k) in the Product Life Cycle 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
W Is Foreign Exporter required to submit a 510(k)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
C Do we have to submit a new 510(k) for a device with a new MCU? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 12
E Do I need submit a traditional 510(k) or Abbreviated 510(k) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
C When to submit a 510(k) for a change to an exisiting medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 29
P Can I submit a 510(k) without aging data? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
C Only appearance and PCB are changed - Do we need to submit a new 510(k)? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
C Application for special 510(k) - What are the documents to submit to FDA? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
A Ownership of 510(k) to transfer or submit new one - Class 1 medical device ISO 13485:2016 - Medical Device Quality Management Systems 4
A Submit a 510(k) to an existing device vs. Just Documenting a Change ISO 13485:2016 - Medical Device Quality Management Systems 2
F How to submit PPAP without IMDS APQP and PPAP 10
M Informational TGA presentation: How to submit an effective good manufacturing practice clearance application Medical Device and FDA Regulations and Standards News 0
Watchcat EU Certification - What do you submit to an NB EU Medical Device Regulations 5
L Reasons to submit a new PPAP or notify the customer APQP and PPAP 3
L 510k exemption process - Do you still need to submit technical dossier? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
N How to deal with catalog parts suppliers who refuse to submit PPAP documents? APQP and PPAP 0
I Where to submit Adverse Events Reports in Europe EU Medical Device Regulations 7
B Should I Request/Submit a PPAP from/to a Supplier/Customer for only Repackaging Parts APQP and PPAP 7
L 20 days to submit PPAP APQP and PPAP 9
A Using Regulations.gov to Submit Comments on FDA Guidance Documents Other US Medical Device Regulations 5
E When to submit a Technical File for a Class IIb Medical Device EU Medical Device Regulations 3
P Do we need to submit Special 510K for a Blood Glucose System? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
D Customer will not submit Purchase Orders - System for ISO 9001 Compliance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 16
L Can we submit an AS9102 if we are not AS9100 certified? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 10
S Guidance to prepare and submit a Level 3 PPAP to a Customer APQP and PPAP 6
I Where do we submit the FAA Preapplication forms? Federal Aviation Administration (FAA) Standards and Requirements 1
O Medical Device Registration in EU - Which Ministry to submit Technical File? EU Medical Device Regulations 9
S How to Submit ISO/TS 16949 & ISO14001 Certificates to HMMA IATF 16949 - Automotive Quality Systems Standard 4
W Submit MDR for a device cleared by FDA but not sold in US? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
P Part Submission Form - Need to submit PSW only once? APQP and PPAP 5
M If I update a DFMEA do I need to submit to my Customer? FMEA and Control Plans 7
Howard Atkins Is there a time frame for CBs to submit the TS 16949 Audit Report? Registrars and Notified Bodies 2
J What is the time limit for an auditor to respond to NCR's and submit audit report? AS9100, IAQG 9100, Nadcap and related Aerospace Standards and Requirements 7
S How to locate a Sponsor in Australia to submit Product Approval Applications by TGA Other Medical Device Regulations World-Wide 3
R In 510k application do we submit documents showing compliance with 21CFRPart820 Other US Medical Device Regulations 12
P Does our company need to be certified in order to submit a 510k Other US Medical Device Regulations 4
A Before the ISO-9001 certification audit can we submit the documents to our customer ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 8
N Supplier PPAP - How many parts should they submit? APQP and PPAP 6
A What is PPAP (Production Part Approval Process) Why submit it and/with PSW APQP and PPAP 2
R When to submit a new FDA Device Listing 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
Similar threads


















































Top Bottom