I was hoping to get some help on a classification query I have. We sell instrument trays used to hold the instruments during reprocessing and Sterilisation. I’m not sure whether they should be classified under MDR as a Class I or Ir or ClassIIA. Any opinions/advice?
Elsmar Forum Sponsor
planB
Super Moderator
#2
drwallice,
classification depends on your defined intended use: if your instrument tray is intended to just _hold_ the instruments in place during reprocessing, then your tray is class Ir. If you claim that your tray is "intended specifically to be used for disinfecting or sterilising medical devices" (quote from Rule 16), then your tray would be class IIa. Typically, an instrument tray is classified as 1r device.
HTH,
RosaB
Registered
#4
Hello,
thank you planB for your answer. May I ask though, why a tray is supposed to be class Ir, if this class is intended for reusable surgical instruments? A tray is definetely not a reusable surgical instruments, as thos are defined as follows: "‘Reusable surgical instrument’ means an instrument intended for surgical use in cutting, drilling, sawing, scratching, scraping, clamping, retracting, clipping or similar procedures, without a connection to an active device and which is intended by the manufacturer to be reused after appropriate procedures such as cleaning, disinfection and sterilisation have been carried out."
I agree though on the classification in class IIa if Rule 16 applies. As I have the same problem as drwallice, I would classify it as an accessory of class I. Would you two agree?
Thanks!
Asia78
Involved In Discussions
#5
Hi, they are definitely not.class ir because they are not surgical devices.
They are also not following rule 16 because they do not act as disinfectant ( art 16 is.followed by other type of devices).
I would classify them as class I, rule 1
#7
Accessories for a medical device shall be classified in their own right separately from the device with which they are used. Chapter II 3.2
Asia78
#8
Exactly.
The accessory take the same class if it is driving or influencing the performance of a (hardware) medical device
Thread starter Similar threads Forum Replies Date
J Reusable surgical Instrument is under class IIa? EU Medical Device Regulations 4
B Bioburden monitoring for surgical instrument provided non-sterile EU Medical Device Regulations 3
M Informational Is my device a Reusable Surgical Instrument under the EU MDR(2017/745)? Medical Device and FDA Regulations and Standards News 0
J IEC 60601-2-2 - EMC Requirements - Surgical instrument manufacturer IEC 60601 - Medical Electrical Equipment Safety Standards Series 6
1 Reprocessing of Surgical Instrument Sets Other Medical Device and Orthopedic Related Topics 4
S Surgical Instrument Luminaries Standards - Single Use Disposable Surgical Instrument IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
L Stainless Steel Grades - Nomenclature for Surgical Instrument Stainless Steel ISO 13485:2016 - Medical Device Quality Management Systems 2
C Use with surgical robot Medical Device and FDA Regulations and Standards News 1
marmotte Help with Surgical Class I + 510(k) exemption US Food and Drug Administration (FDA) 0
D Cleanroom requirements for surgical apparel EU Medical Device Regulations 2
S Requirements for Certificate of Destruction for Surgical Instruments Medical Device and FDA Regulations and Standards News 2
P UDI-PI requirements on reusable surgical device, do we need serialisation? ISO 13485:2016 - Medical Device Quality Management Systems 3
C Non-sterile reusable surgical instruments - FDA sterilization requirement Other Medical Device Related Standards 2
D ISO 13485 & CE Certification for Surgical Gloves CE Marking (Conformité Européene) / CB Scheme 0
S Required tests for Surgical gown US Food and Drug Administration (FDA) 1
B NIOSH Approval for Surgical N95 Respirators - Required testing US Food and Drug Administration (FDA) 2
M Surgical angle measurement guide device with an application software Medical Device and FDA Regulations and Standards News 4
R HV Test of HF Surgical Unit IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
E MRI and Surgical Needle Other Medical Device Related Standards 0
C Surgical mask stability testing (CE mark) EU Medical Device Regulations 2
Z MDR Class Ir deadline (Reusable surgical instruments) EU Medical Device Regulations 27
P Surgical masks ( type IIR) - Cleanroom Requirements EU Medical Device Regulations 3
A Is there any Light for a surgical microscope? Other ISO and International Standards and European Regulations 3
M Risk managment report of Surgical Mask Example ISO 14971 - Medical Device Risk Management 14
J Guidance on manufacturing non-surgical face masks US Food and Drug Administration (FDA) 3
M Regulatory clearance pathways for surgical masks/N95 respirators Other US Medical Device Regulations 3
M Are Electrical Laparoscopic Surgical Devices ME equipment? IEC 60601 - Medical Electrical Equipment Safety Standards Series 8
Z Labeling for Medical Device Samples - Surgical instruments 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
B ISO 11607-1 for surgical instruments sterilized by end user Other Medical Device Related Standards 1
M Informational US FSA – Surgical staples – Guidance and proposed rules Medical Device and FDA Regulations and Standards News 0
M Informational FDA order – All manufacturers of surgical mesh intended for transvaginal repair of anterior compartment prolapse (cystocele) to stop selling and distr Medical Device and FDA Regulations and Standards News 0
J0anne Implantable Devices - The effects of surgical scaffold implanted into the human body Other Medical Device and Orthopedic Related Topics 8
M FDA News FDA announces panel meeting on surgical mesh placed transvaginally to treat pelvic organ prolapse Medical Device and FDA Regulations and Standards News 0
J IEC60601-2-18 - Endoscopic Surgical Monitor System Images IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
N Strength verification of surgical instruments ISO 13485:2016 - Medical Device Quality Management Systems 5
E Canada classification of surgical displays, MDL needed? Canada Medical Device Regulations 2
C Class I Reusable Surgical Instruments - CE Mark with the Notified Body number EU Medical Device Regulations 14
P IFU for Sterling Silver Surgical devices ISO 13485:2016 - Medical Device Quality Management Systems 3
C What is the most appropriate Bioburden Test Method for Surgical Gloves ISO 13485:2016 - Medical Device Quality Management Systems 10
R FDA Biocompatibility questions for Surgical Gown 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
R FDA Shelf Life Requirements for EO Surgical gown 510k 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 10
I Defending "No Essential Performance" for a Surgical Video Camera IEC 60601 - Medical Electrical Equipment Safety Standards Series 13
S Is double packaging obligatory for surgical single-use scalpels? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 5
H Stainless Steel Surface Quality for Ophthalmic Surgical Device Other Medical Device Related Standards 11
E Vaccinations for Repair of Returned Surgical Devices US Food and Drug Administration (FDA) 3
T Electrical Surgical Tool with Big Metal Applied Part IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
E Open Sourced Surgical Tray Sterilization Database and Software wanted Medical Information Technology, Medical Software and Health Informatics 2
K Regulations for sending Veterinary Surgical Instruments to Qatar Other Medical Device Regulations World-Wide 4
L Does IEC 61010-1 apply for Surgical Microscopes? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
N Reusable Surgical Instruments - Definition per Annex IX of Directive 93/42/EEC EU Medical Device Regulations 5
Similar threads
Top Bottom