Ok, so I have got little problem that I can't seem to find the solution for.

We have a validated sterilization process where the master product does no longer exist (out of production since a long time..) and we would like to identify a new such master product. We have a good candidate product that we can use.

The current validated uses a PCD which is shown to be more challenging than the old master product, and all current products are adopted into this process.

We use an extreme over-kill cycle for EO but during audits we sometimes get questions about the master product that we do no longer have.

So the question is, can we, without full re-validation, switch the master product to our new candidate. If yes, what steps should we do?

We have the data for the selection of the PCD, if we can show that the new master product is equivalent or easier than the old master product to sterilize, would this fact suffice to make the switch?

Any help here would be highly appreciated!
 
Elsmar Forum Sponsor
M

MIREGMGR

#2
(out of production since a long time..)
I'm surprised that your Notified Body didn't red-flag this no later than when your first periodic re-validation after discontinuance of that product was due.

My take would be that your periodic re-validations since then would be at least questionable, and therefore you might reasonably be expected to do a new full validation.
 

planB

Trusted Information Resource
#3
Dojla,

refer to ISO 11135:2014, section D.7.1.6

"A means of demonstrating equivalence to a previously qualified product or internal PCD is the comparison of the relative rates of inactivation of BIs placed in a challenge location within the new or modified product and previously qualified product/master product(see D.8.6 and D.12.5.2) when both are exposed to a fractional cycle. Equivalence studies should compare the new or modified product to the internal PCD used to validate the process. If a PCD is used for this comparison, this resistance of the PCD should be assessed as part of the annual review."

So you are on the right track with
> We have the data for the selection of the PCD, if we can show that the new master product is equivalent or easier than the old master product to sterilize, would this fact suffice to make the switch?

This can be done experimentally via fractional-cycle studies or you may also be able to demonstrate equivalency/no-greater challenge by following ISO 11135:2014, section D.7.1 and /or by making use of the framework provided by AAMI TIR 28:2009, Product adoption and process equivalence for ethylene oxide sterilization.

Also note that you do not have to define an actual product being your master product, but you can also define a simulated product or even your PCD being the master.

HTH,

Gerhard
 

big boss

Posts Moderated
#5
ear all
i have a question, as per the iso 11135:2014 in OQ part
1- do we need to run humidity mapping ?? as in this phrase in standard it refer to table C1 which releate to number of data logger for temp only , also is it make sense as humidity injected to be 60 % as an example so in empty chamber this not supposed to be decreased or increased !!! so can any one help to clarify this point
2- also i need to understand (( In empty chamber OQ exercises, the recorded temperature range, within the usable chamber volume during EO or inert gas exposure, of ± 3 °C of the average recorded chamber temperature at each time point should be obtained after an equilibration period)) as this not clear for me
 
Thread starter Similar threads Forum Replies Date
A ISO 2859-1 - 9.3.3.2 Switching score - Sampling Plans and PPM AQL - Acceptable Quality Level 0
M Switching (Migrating) to a new QMS Software Package Document Control Systems, Procedures, Forms and Templates 10
S Change notice while switching to injection molding technique for manufacturing EU Medical Device Regulations 1
R What is the expectation for Biocompatibility Testing when switching Suppliers? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 11
A Attribute Sampling Plans with Switching Rules - Software Recommendations Inspection, Prints (Drawings), Testing, Sampling and Related Topics 1
R Long Range OC Curves - Acceptance Sampling Switching Rules Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
I Microbiological Monitoring for Surfaces - Switching from Swab to Contact Plates Other Medical Device Related Standards 2
S Switching 510(k) product from Prescription to OTC US Food and Drug Administration (FDA) 2
I Switching Serial Numbers on Repaired Medical Devices Other US Medical Device Regulations 2
R Switching Power Supplies Certified to IEC 60601 3rd Edition IEC 60601 - Medical Electrical Equipment Safety Standards Series 5
R Reference Voltage for Switching Power Supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 11
R Windings Temperature exceeds the Limit - Overload Test of Switching Power Supply IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
D Switching to a new Quality Management software - Seeking recommendations Quality Assurance and Compliance Software Tools and Solutions 4
G Customer Notification requirements when Switching Facility APQP and PPAP 10
W ANSI Z1.4 Switching Rule Applications Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
M Switching over to Management - Is the MBA an Asset? Career and Occupation Discussions 10
R Sampling Plans - ISO 2859-1: 1999 - How to calculate the switching score? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
A Is it acceptable to use a Sampling Plan and not use the Switching Rules? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
M Switching to new problem solving methodology Quality Tools, Improvement and Analysis 7
E ISO 2895/3951 Switching Inspection Rules Inspection, Prints (Drawings), Testing, Sampling and Related Topics 11
C BS6001 sampling tables - Seeking software with Alert for switching rules Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
T PPAP and Raw Material - Is a PPAP required when switching raw material suppliers? APQP and PPAP 4
Marc Registrar Comparisons - We're thinking of switching registrars Registrars and Notified Bodies 6
Marc Switching Registrars (A 'Life Long' Relationship?) and Problems for Auditors Registrars and Notified Bodies 16
I Master Document Access - ISO 9001:2015 clause 7.5.3 Document Control Systems, Procedures, Forms and Templates 5
Q Forms Master List versus Record Matrix ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 3
E Part 11 Compliance, Excel living documents (i.e. document master list, equipment list, approved supplier list) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
R No design history file or device master record ISO 13485:2016 - Medical Device Quality Management Systems 5
Sam Lazzara Medical Device File (MDF per 13485:2016 4.2.3) versus FDA Device Master Record (DMR) ISO 13485:2016 - Medical Device Quality Management Systems 3
D Virtual Documentation via Master Flow Chart? Process Maps, Process Mapping and Turtle Diagrams 4
M Master Dot for quality planning/improvement Manufacturing and Related Processes 1
M Medical Device Master File - We are a contract manufacturer ISO 13485:2016 - Medical Device Quality Management Systems 7
G Comparing Gauge Blocks - Translating the calibration between the master set and the QA set General Measurement Device and Calibration Topics 2
R Are DMR (device master record) and DHR (device history record) inspected by FDA or audited by an NB ISO 13485:2016 - Medical Device Quality Management Systems 6
S Foreign Manufacturer and DMR - Who keeps the DMR (Device Master Record)? ISO 13485:2016 - Medical Device Quality Management Systems 2
S DMR (Device Master Record) for Medical Device Software IEC 62304 - Medical Device Software Life Cycle Processes 3
D Design Outputs vs DMR (Design Master Record) 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
V Carbon Black (Master Batch) - Basic Requirements Manufacturing and Related Processes 5
H Korea DMF (Drug Master File) Fees Various Other Specifications, Standards, and related Requirements 5
K Computerized system Master Validation Plan, IQ-OQ-PQ Draft or Template Qualification and Validation (including 21 CFR Part 11) 4
R Master Gage Block Calibration (We have 2 sets) question General Measurement Device and Calibration Topics 3
P DMF (Drug Master File) type V for Sterilization Processes for USFDA Submission Other US Medical Device Regulations 3
Q Master FMEA for "like processes" - Does an FMEA have to match the CP and PM directly? FMEA and Control Plans 7
S Looking for templates of Master Production Record and Master Batch record Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
T Master Document List - Obsolete Documents Document Control Systems, Procedures, Forms and Templates 23
V Site Master File for multiple business units within same premises/facility Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 1
rob73 Digital DMF (Device Master File) Records EU Medical Device Regulations 3
I DMR (Device Master Records) and similar for standalone software device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 2
S Device Master Records for a Contract Manufacturer 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 4
S DMR (Device Master Record) Management and Index questions Other US Medical Device Regulations 18

Similar threads

Top Bottom