Symbol Colours and Sizes on Medical Devices

Elsmar Forum Sponsor
Z

zhang126

#12
In MDD 93/42/EEC,
ANNEX I
ESSENTIAL REQUIREMENTS
...
..
.
13.2. Where appropriate, this information should take the form of symbols.
Any symbol or identification colour used must conform to the
harmonized standards.
In areas for which no standards exist, the
symbols and colours must be described in the documentation
supplied with the device.
 
M

MIREGMGR

#14
(...) although i think FDA recognizes the EN 980 fonts (Symbols for labeling medical devices).
It's hard to tell, when one is informed of "the FDA's stance" or "interpretation" during the 510(k) process by an examiner, whether the information being presented is idiosyncratically unique to that examiner or actually "the FDA's stance". That notwithstanding, we have been told by two FDA examiners now that EN 980 and ISO 15223 specifically are not recognized.

One of those two examiners then told us that the only symbols the FDA will accept on device labels without explanatory and self-sufficient subtext (in English only) were those that are defined in the IVD labeling guidance (whether or not the device in question is an IVD).

The other examiner told us that the FDA will not accept any symbol on a non-IVD medical device label, or elsewhere on the device, without explanatory and self-sufficient subtext in English only.

Note that in both cases, definition of the symbol elsewhere in an IFU or other accompanying information was not acceptable. The information had to be on the label, where the symbol appears.

I asked the second examiner how one would be expected to meet international labeling requirements in regard to multiple languages and symbol preference, and was told that the FDA does not consider regulatory requirements outside of the US to be relevant to US requirements.

It's generally known, I think, that there is an internal conflict within the FDA between what I would call the pro-harmonization faction, which includes Tim Ulatowski of CDRH, and a traditionalist faction. My impression is that the 510(k) examiners' office must be in the latter camp.
 
R

RetroVirus

#15
i have heard about instances that MIREGMGR spoke about - but in my work experience within companies whose labelling of FDA approved products carried symbols on the device labelling, the FDA inspector who subsequently audited us never mentioned anything against the use of symbols and certainly nothing was mentioned by the 510(k) reviewers.

i did not work in the regulatory side of these companies but i do know that labelling with symbols (i can now post the url below) were indeed acceptable as long as a glossary is provided in the IFU/manual. certainly this is true for small labels in which full text would be either difficult to print or too small to read. Note though that the guidance reads:

"FDA recognizes 25 symbols for IVD devices for professional use from the following two international consensus standards:
*ISO 15223, Medical Devices - Symbols to be used with medical device labels, labeling and information to be supplied
*EN 980, Graphical symbols for use in the labeling of medical devices"

the symbols that the companies i worked in used were within the recognized symbols. the approval of FDA guidance though expired in 2007 and no replacement has been published by the FDA. perhaps that is what the examiners (of the poster above) were siding with either that of the symbols being used were not any of the 25 recognized by the FDA.

http://www.fda.gov/downloads/Biolog...toryInformation/Guidances/Blood/UCM085430.pdf

RV

It's hard to tell, when one is informed of "the FDA's stance" or "interpretation" during the 510(k) process by an examiner, whether the information being presented is idiosyncratically unique to that examiner or actually "the FDA's stance". That notwithstanding, we have been told by two FDA examiners now that EN 980 and ISO 15223 specifically are not recognized.

One of those two examiners then told us that the only symbols the FDA will accept on device labels without explanatory and self-sufficient subtext (in English only) were those that are defined in the IVD labeling guidance (whether or not the device in question is an IVD).

The other examiner told us that the FDA will not accept any symbol on a non-IVD medical device label, or elsewhere on the device, without explanatory and self-sufficient subtext in English only.

Note that in both cases, definition of the symbol elsewhere in an IFU or other accompanying information was not acceptable. The information had to be on the label, where the symbol appears.

I asked the second examiner how one would be expected to meet international labeling requirements in regard to multiple languages and symbol preference, and was told that the FDA does not consider regulatory requirements outside of the US to be relevant to US requirements.

It's generally known, I think, that there is an internal conflict within the FDA between what I would call the pro-harmonization faction, which includes Tim Ulatowski of CDRH, and a traditionalist faction. My impression is that the 510(k) examiners' office must be in the latter camp.
 
Thread starter Similar threads Forum Replies Date
A ISO/DIS 15223-1:2020 - Country of manufacture label (IEC 60417 No. 6049) - Which national law requires this symbol? Other Medical Device Related Standards 0
P Do we need to retrospectively use the "MD" symbol (indicating device is a medical device) on labels, e.g. finished devices within expiration date? EU Medical Device Regulations 2
D Tolerance symbol Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
K Contains hazardous substance symbol EU Medical Device Regulations 1
K Is UDI (Unique Device Identifier) symbol mandatory under EU MDR? EU Medical Device Regulations 3
A What does this sentence "this symbol shall be used in the orientation shown" mean in ISO 780:2015? Other Medical Device Related Standards 4
A Read instruction manual - What Graphic Symbol must I use? US Food and Drug Administration (FDA) 7
E Stand-by vs. ON/OFF symbol IEC 60601 - Medical Electrical Equipment Safety Standards Series 2
K ISO 15223-1 Sterile barrier system - The 3-layer symbol EU Medical Device Regulations 1
Q LOT or Serial Number Symbol not used when the information is contained in the UDI? 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 1
K EIFU indicator symbol - Class I labels as per EU MDR EU Medical Device Regulations 3
J Create your own symbol? Other Medical Device Related Standards 7
K REF symbol with product code in the IFU EU Medical Device Regulations 1
S ISO 18223-2 Temperature symbol in the Legend table in the Instructions for Use Other ISO and International Standards and European Regulations 4
K Medical Device Symbol(s) on Accessories EU Medical Device Regulations 28
shimonv Symbol for temperature limit EU Medical Device Regulations 6
echo7tango Circled-Z symbol on a syringe — what is it for, what does it mean? China Medical Device Regulations 21
V "Manufacturer Symbol on Device is missed due to space constraint" EU requirement EU Medical Device Regulations 11
J ISO 15223-1: Keep dry vs. humidity limitation label symbol Other Medical Device Related Standards 2
Z IFU (Instructions for Use) Symbol - 7010:M002 - can it be in black? IEC 60601 - Medical Electrical Equipment Safety Standards Series 1
M GD&T tolerance or band? Using a symbol like parallelism // Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
J Use of Caduceus symbol on Class II medical device 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 6
B Old GD&T Symbol "~" Inspection, Prints (Drawings), Testing, Sampling and Related Topics 2
O Can a Profile of a Surface have a Diameter Symbol? Inspection, Prints (Drawings), Testing, Sampling and Related Topics 3
Q Labeling/Marking Location for BF Symbol IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
F RFID Symbol for Medical Devices EU Medical Device Regulations 3
I Use of the Non-Sterile Symbol (EN980/ISO7000/ISO15223-1) Other Medical Device Related Standards 3
P CE Marked IVD Labelling Canada - Conformity symbol? Canada Medical Device Regulations 6
E Unfamiliar Blueprint Symbol - Hex Symbols Inspection, Prints (Drawings), Testing, Sampling and Related Topics 8
A Reference Source for Polarity Symbol Other ISO and International Standards and European Regulations 1
C CE Mark Label Symbol Requirements on a Medical Device CE Marking (Conformité Européene) / CB Scheme 3
A Does anybody know in which Standard can this attached Alarm Symbol be found? Other Medical Device Related Standards 3
B ISO 3864-4 Symbol Color Regulations Other ISO and International Standards and European Regulations 4
N Latex Free symbol for use in UK? EU Medical Device Regulations 3
K Class Ins CE Mark Labeling Requirements (particularly Lot or Batch symbol and number) EU Medical Device Regulations 7
R Storage Symbol for Labeling ISO 13485:2016 - Medical Device Quality Management Systems 3
M US FDA Harmonization of Label Symbol Rules 21 CFR Part 820 - US FDA Quality System Regulations (QSR) 3
M Appropriate use of the REF symbol in IVD kits and kit components Other Medical Device Related Standards 2
Q Symbol Identification Help Other Medical Device Related Standards 14
J Symbol for manufacturing date - Is ISO 980 expired? CE Marking (Conformité Européene) / CB Scheme 11
N Symbol for Medical Device Sterile Packaging - "Inspect Seal Before Opening" Other Medical Device Related Standards 5
C Tolerance and Uncertainty have the same Symbol - How to Calculate the Tolerance Measurement Uncertainty (MU) 1
S Medical Device Labeling - Is there a Thermo Disinfector/Washer Symbol Other Medical Device Related Standards 7
B Do we need to put the box around the SN symbol listed in EN 980? Other ISO and International Standards and European Regulations 5
A Medical Device Labeling - Non-Pyrogenic Symbol Other Medical Device Related Standards 6
N Unknown Medical Symbol in Product Label ISO 13485:2016 - Medical Device Quality Management Systems 1
B Equipotentiality Symbol - IEC 60417-5021 IEC 60601 - Medical Electrical Equipment Safety Standards Series 9
M How to make an X Median symbol (capital X with a tilde on top) in MS Office Quality Assurance and Compliance Software Tools and Solutions 11
E Is there a need to pay for the symbol quoted in ISO 3826-2:2008? Other ISO and International Standards and European Regulations 1
L Use of the EN 980 Manufacturer Symbol CE Marking (Conformité Européene) / CB Scheme 3

Similar threads