Technical Documentation Review: Responding to a Nonconformity
Ronen E
Problem Solver
Staff member
Moderator
#11
I do agree, however, the requirement is not that manufacturers only show compliance with essential requirements, as, related to the standards, there's this requirement:
It only requires a description, not a justification, so it's still not a requirement as I mentioned anyway.
Marcelo,
Your quote from the MDD (reproduced below) is from Annex II, so it only applies in those cases where the manufacturer chooses the Annex II route. Besides, it only requires listing the standards that will be applied. In my understanding, if the manufacturer chooses NOT to apply some standards, or even just doesn't apply some standards without a clear decision/consideration, this clause doesn't require doing anything about it.
3.2. Application of the quality system must ensure that the products conform to the provisions of this Directive which apply to them at every stage, from design to final inspection. All the elements, requirements and provisions adopted by the manufacturer for his quality system must be documented in a systematic and orderly manner in the form of written policies and procedures such as quality programmes, quality plans, quality manuals and quality records.
It shall include in particular the corresponding documentation, data and records arising from the procedures referred to in point (c).
It shall include in particular an adequate description of:
(c) the procedures for monitoring and verifying the design of the products, including the corresponding documentation, and in particular:
— the design specifications, including the standards which will be applied and the results of the risk analysis, and also a description of the solutions adopted to fulfil the essential requirements which apply to the products if the standards referred to in Article 5 are not applied in full,
Elsmar Forum Sponsor
Marcelo
Inactive Registered Visitor
#12
In fact, the same text with a slight variation is present in Annexes II, III, VII and VII, which means in practice it is required in for all routes.
In the case of not using harmonized standards, it requires a description of the solution, not a list of standards. The list of standards is in the first part of the text.
But again, it does not require that you compare your "alternative" solutions if not using harmonized standard, it only requires that you describe then. The NB would need to verify the alternate solutions, but as you mentioned, they usually do not have enough expertise.
Ronen E
Problem Solver
Staff member
Moderator
#13
I'm glad that we agree on the essence.
Coming back to the original question, there is no requirement in the MDD to list or justify standards that are not applied.
Mark Meer
Trusted Information Resource
#14
I agree with Ronen.
Mark, my statement that "Compliance with the MDD equals compliance with all the relevant harmonized standards to your organization" is not incorrect.
It's NB gospel truth :)
I think we're all in agreement here that:
1. In principle, and strictly by regulations, standards are voluntary, and you don't have to justify not using them as long as ER compliance can be demonstrated.
2. Many NBs are increasingly choosing to treat standards as defacto mandatory (due, in part, to lack of expertise to adequately evaluate alternate compliance evidence).
..and I'd go so far as to add:
3. This places increasing burden on Manufaturers, as they don't know what to expect when interacting with NBs. The "safe" route (explicit compliance to all applicable harmonized standards) often involves greater resources (with little or no value-added - except for NB appeasement), whereas using your own solutions is a gamble, depending on the NB.
Overall, a bad state of affairs for the industry, IMO.
It is unfortunate that dealings with NBs are becoming increasingly more and more bureaucratic in nature. That is my opinion.
Agreed.
sriramsl
Involved In Discussions
#15
Thanks a lot all of you for your insights on this thread. Very useful information and answers for a very important and relevant question.
Thanks again
Sriram
Thread starter Similar threads Forum Replies Date
H MD Technical Documentation and SALES possibility EU Medical Device Regulations 0
A DMDIV-Variants : definition and technical documentation CE Marking (Conformité Européene) / CB Scheme 4
P MDR/IVDR Annex III - Technical documentation on PMS EU Medical Device Regulations 1
I Maintaining Technical Documentation under MDD and MDR EU Medical Device Regulations 1
M Basic UDI and technical documentation EU Medical Device Regulations 1
C Maintenance of Product Technical Documentation EU Medical Device Regulations 3
C Access to Technical Documentation - MDR Chapter VI Article 61 EU Medical Device Regulations 4
N EU MDR Basic UDI-DI and Technical Documentation for Systems EU Medical Device Regulations 22
I MDD Class I software technical documentation sample EU Medical Device Regulations 2
M Annex II - Technical Documentation. V&V, Performance and Safety EU Medical Device Regulations 3
D MDR and Summary Technical Documentation Files EU Medical Device Regulations 18
B Gap analysis of new start-up company technical documentation CE Marking (Conformité Européene) / CB Scheme 1
A Technical Documentation - Manufacturing Specifications EU Medical Device Regulations 4
J Software for maintaining of Technical files and medical device documentation(RA need) Medical Information Technology, Medical Software and Health Informatics 2
F Technical File - RoHS2 Technical Documentation RoHS, REACH, ELV, IMDS and Restricted Substances 2
Q Use of standard references in documentation (PRS, Technical File) Document Control Systems, Procedures, Forms and Templates 1
T Information and Structure for Technical Documentation for Firmware ISO 13485:2016 - Medical Device Quality Management Systems 1
A ISO 16016 - Technical product documentation Other ISO and International Standards and European Regulations 2
S Translation Software for Documentation with Technical Words - Recommendations wanted Quality Tools, Improvement and Analysis 5
S Medical Device Technical Files (Class IIa & IIb) ? Design Development Documentation EU Medical Device Regulations 7
M Documentation of OEM Device - Technical File for a Class IIa Medical Device ISO 13485:2016 - Medical Device Quality Management Systems 4
Le Chiffre Health Canada to accept Summary Technical Documentation (STED) for Class III & IV ISO 13485:2016 - Medical Device Quality Management Systems 1
C Technical Documentation, DMR, ISO 13485 vs. FDA, STED Document Control Systems, Procedures, Forms and Templates 10
C Structure and Content of DMR / Technical Documentation (FDA vs. ISO 13485)? Document Control Systems, Procedures, Forms and Templates 6
A Technical Documentation Records - Customer drawing, Revision Level, PPAP, etc. Document Control Systems, Procedures, Forms and Templates 1
D Technical Documentation Package (TDP) - Listing of the types of documents contained Document Control Systems, Procedures, Forms and Templates 4
M Does anyone know of a forum that discusses Technical Documentation? Document Control Systems, Procedures, Forms and Templates 4
K Technical file for mhra CE Marking (Conformité Européene) / CB Scheme 2
O Soft and technical skills for a VP of Quality Career and Occupation Discussions 1
A CE-IVD Technical File with Multiple IFUs EU Medical Device Regulations 4
M Office and manufacturing site relocation effect on Device Technical File CE Marking (Conformité Européene) / CB Scheme 2
R MDR & depth of technical files review by NB's EU Medical Device Regulations 18
J Saudi Arabia Technical File Requirements Other Medical Device Regulations World-Wide 1
DuncanGibbons Technical Data Package vs Digital Product Definition APQP and PPAP 0
D Hospital Decontamination and the MDR - Technical Files, UDI and PMS EU Medical Device Regulations 3
F Software as a Medical Device (SaMD) Technical File Requirements Manufacturing and Related Processes 3
S Technical File for Surgical Mask in Class 1 MDD Elsmar Cove Forum ToS and Forum Policies 1
A Technical File - Updating Applied Standards EU Medical Device Regulations 4
M What to be careful about/focus on when doing a Technical File review EU Medical Device Regulations 4
Watchcat Anyone had an MDR technical file review/audit yet? EU Medical Device Regulations 13
G Same intended use devices - Technical file EU Medical Device Regulations 5
B Clause 5.1.12 of Technical Standard IEC 62304/A1 IEC 62304 - Medical Device Software Life Cycle Processes 5
V Iso13485 certification vs CE technical audit ISO 13485:2016 - Medical Device Quality Management Systems 3
F Management of Software version while NB reviews Technical file CE Marking (Conformité Européene) / CB Scheme 7
D Risk Analysis & Technical File - What detail goes in the Risk Management Report ISO 14971 - Medical Device Risk Management 5
F Recommended practice for furnishing Technical Procedures ISO 17025 related Discussions 1
S How to find technical specification for facilities maintainance ISO 9000, ISO 9001, and ISO 9004 Quality Management Systems Standards 2
L Special Processes - Technical reason concerning torque - Anaerobic compounds Manufacturing and Related Processes 10
N Technical file requirements class 1 device MDR Other Medical Device Regulations World-Wide 6
R Implementing Design History Documents/ Technical Documents in the QMS ISO 13485:2016 - Medical Device Quality Management Systems 3
Similar threads
Top Bottom