Technical File Maintenance - Old generation product
Aaria
My question is related to maintaining technical file for a legacy product. We have CE mark for V(x) of the medical device and now V(x.1) is in the design and development phase. We never launched V(x) but were only using the CE marked version in clinical studies.
The manufacturing of V(x) has been decommissioned last year because of which few documents such as Router/BOM were marked 'obsolete' in the change management system. Because we have an active CE mark certification, do we have to keep the supporting documents in the "approved/released" status to demonstrate an active "DMR". Typically record retention would kick in but the questions is more about "approved-relased" vs. "obsolete" status of supporting documentation to maintain the tech. file.
Sorry, for the long text, but really hoping to get some input.
Thanks!
Ronen E
Problem Solver
Moderator
A. I'm not aware of a specific requirement to maintain the tech file documents in a specific status. the requirement / expectation is that those documents will be available for review in a reasonably timely manner (in my opinion regardless of status "active" vs. "obsolete").
B. If you don't intend to introduce V(x) any more then I would recommend contacting your NB and just notifying them that the product is discontinued. That would minimise residual requirements related to its tech file.
Aaria
Thank you for your reply and it makes sense that regardless of the document status, the tech. file can be maintained if those documents are available.
For response, B., I've a follow-up question, hypothetically, if we were to contact NB and inform that the product is now discontinued, we would lose the CE mark status as a company correct? If it were to stay active, for the next version of the product it would be a change impact assessment vs. an entire new tech. file. What are your thoughts?
Thanks!
Ronen E
Problem Solver
Moderator
There's no such thing as "CE mark status" for a company. CE marking clearance (EC certificate) is granted for products. So in my understanding you can't lose "your status" as a company. In my opinion the revised device version would be assessed as an incremental change from the old, cleared version, regardless of whether the latter is discontinued or not, based on your communications with the NB. They should know that the old version has been cleared and that it wasn't discontinued due to some failure or lack of compliance, but simply due to business considerations, and therefore it's still a good baseline for the upgrade (in the regulatory sense). To be on the safe side, I recommend open communication with the NB to that extent, upfront. I estimate that they will cooperate.
Cheers,
Ronen.
fialor
Thanks!
Aaria,
Three points are worth bearing oi mind:
1 - unless your product is a class III, the CE mark will be for either the generic product group or product family. Meaning you are able to include devices of generational increase or same focus (use, user group, function, manufacturing processes etc). This means the number of products in the group or family increases or decreases depending on what you add in or take out. When you advise your NB that you have discountinued the said product in that group, they will remove it from the relevant CE mark certificate. Some NBs actually include the products list as an annex to the certificate and will charge for revision, some do not include it.
2 - if the technical file is for both versions of the device then you would need to maintain it as it would include maintaining for the newer version of device.
If the technical files are separate, then no you do not need to maintain obsolote product's.
3 - However you would need to keep the record of the obsolete device for 10 years after the last device was placed on the market if your device falls under the remit of RoHS.
I hope this helps.
A Technical Files/Dossier/STED File Maintenance for Notified Body Audiits Other Medical Device Regulations World-Wide 2
K Technical file for mhra CE Marking (Conformité Européene) / CB Scheme 2
A CE-IVD Technical File with Multiple IFUs EU Medical Device Regulations 4
M Office and manufacturing site relocation effect on Device Technical File CE Marking (Conformité Européene) / CB Scheme 2
J Saudi Arabia Technical File Requirements Other Medical Device Regulations World-Wide 1
F Software as a Medical Device (SaMD) Technical File Requirements Manufacturing and Related Processes 3
S Technical File for Surgical Mask in Class 1 MDD Elsmar Cove Forum ToS and Forum Policies 1
A Technical File - Updating Applied Standards EU Medical Device Regulations 4
M What to be careful about/focus on when doing a Technical File review EU Medical Device Regulations 4
Watchcat Anyone had an MDR technical file review/audit yet? EU Medical Device Regulations 13
G Same intended use devices - Technical file EU Medical Device Regulations 5
F Management of Software version while NB reviews Technical file CE Marking (Conformité Européene) / CB Scheme 7
D Risk Analysis & Technical File - What detail goes in the Risk Management Report ISO 14971 - Medical Device Risk Management 5
N Technical file requirements class 1 device MDR Other Medical Device Regulations World-Wide 6
O Medical Device Technical File Confidentiality - On site reviews EU Medical Device Regulations 14
N How to Write and Compile a Technical File for CE Marking CE Marking (Conformité Européene) / CB Scheme 4
N Technical File Reviewer has requested more testing to ISO 10993 Other Medical Device Related Standards 10
M Technical File (STED) labelling - When you compile tech files do we have to include all the product labels? EU Medical Device Regulations 2
S MDR Technical File Structure Requirements EU Medical Device Regulations 12
D MDD to MDR - Technical File conversion EU Medical Device Regulations 4
J Connection between Technical File and Declaration of Conformity CE Marking (Conformité Européene) / CB Scheme 2
T Technical File Review - Transfer to a new NB EU Medical Device Regulations 3
J One Medical Device, two indications, two classes, one technical file? EU Medical Device Regulations 4
A Technical file requirement for Class IIb medical device CE Marking (Conformité Européene) / CB Scheme 3
D Technical file structure/content requirements for Class IIa devices & new MDR EU Medical Device Regulations 14
J Update Technical File for EU Class IIa Medical Software Products EU Medical Device Regulations 3
T IVD - Self Declaration - Technical File - Template wanted Various Other Specifications, Standards, and related Requirements 9
T Technical File - Previous Device Generations EU Medical Device Regulations 2
K Questions on Technical File to show it is current EU Medical Device Regulations 5
Sam Lazzara Expedited, fast-track Notified Body Technical File reviews EU Medical Device Regulations 3
S Question about Australia - TGA and MDSAP - Technical file ISO 13485:2016 - Medical Device Quality Management Systems 3
A Technical File Procedure (for Company) - Example wanted EU Medical Device Regulations 1
C Technical File Assessment from a NB - Clinical Equivalence CE Marking (Conformité Européene) / CB Scheme 6
M Merge Technical File, DMR, and Device File into a single document? Other Medical Device and Orthopedic Related Topics 3
R Technical File Guidance for Preclinical Evaluations EU Medical Device Regulations 12
S On site Technical File Review by our NB EU Medical Device Regulations 3
R ISO13485:2016 Clause 4.2.3 - Changes in Technical File Requirements ISO 13485:2016 - Medical Device Quality Management Systems 18
shimonv Which class I accessory information should be kept in the Technical File EU Medical Device Regulations 6
JoCam Review of Technical File for CE mark EU Medical Device Regulations 6
F Technical File - RoHS2 Technical Documentation RoHS, REACH, ELV, IMDS and Restricted Substances 2
Q Use of standard references in documentation (PRS, Technical File) Document Control Systems, Procedures, Forms and Templates 1
N CE Marking Checklist Technical File Under MRA EU Medical Device Regulations 5
A CE Mark - Maintaining Technical File and Regulatory Assessment CE Marking (Conformité Européene) / CB Scheme 1
O CE TCF (Technical Construction File) Template for LVD/EMC/Pressure Directive CE Marking (Conformité Européene) / CB Scheme 5
M Technical File Locations - Class I and Class IIa Medical Devices EU Medical Device Regulations 3
D TENS Device (Transcutaneous Electrical Nerve Stimulator) Technical File CE Marking (Conformité Européene) / CB Scheme 4
B Which Department does the Technical File reside? EU Medical Device Regulations 9
Q Class I Technical File reviewed during ISO 13485 Certification Audit by NB CE Marking (Conformité Européene) / CB Scheme 10
F MDD, Technical File, Clinical Evaluation for PACS/DICOM Viewer ISO 13485:2016 - Medical Device Quality Management Systems 2
E When to submit a Technical File for a Class IIb Medical Device EU Medical Device Regulations 3