Temperature Cycling Process - Continual Improvement and ISO 13485
somashekar
We have temperature cycling as a process in the medical sensor cable assembly and we have been doing this over several years. The data from this has been no failures.
The processes involved in the cable assembly are also validated and I have faced no customer complaints or any early failure feedbacks.
Under this background and based on our data and other technical judgements, could we eliminate this temperature cycling process, there by bring in improvement on energy and throughput time ?
DannyK
Re: Continual improvement and ISO 13485 ?
If you have no technical or customer requirements to perform the temperature cycling, you can probably remove it.
Make sure you update your risk management process and re-assess the risks involved.
Marc
We have temperature cycling as a process in the medical sensor cable assembly...
Do you have any idea why you started doing temperature cycling in the first place?
Marc
OK - Yeah, I've been down that road. Typically if you've had a few years of no failures/returns you can point to that as evidence that your soldering process is robust and temperature cycling isn't buying you anything. Do be sure it's not a documented requirement before you drop it, though. I've see times where temperature cycling or thermal shock never produced failures, but the equipment was high reliability and it was required on 100% of product by the contract regardless of whether it induced failures or not.
H
hitesh
You might want to check your design input requirements. For example,if solder joint reliability shall be intact for the product life cycle, before you remove thermal cycling requirement, Enough products must have passed through product life cycle to consider product confirmity data as a rationale to remove thermal cycling.
Marc
True, and let's not forget data from field returns. That is a data aspect which has to be considered.
G Validity of Extending Temperature Cycling Various Other Specifications, Standards, and related Requirements 3
J Temperature cycling thermal chamber accuracy and Gage R&R Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 2
M GMW3172 - Electrical/Electronic - Temperature cycling and Thermal shock Customer and Company Specific Requirements 4
J Warehouse temperature and humidity control EU Medical Device Regulations 7
T Temperature Requirements For In House Calibration - AS9100 AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 16
C One Time Service Supplier - Temperature and Humidity Testing Service ISO 13485:2016 - Medical Device Quality Management Systems 5
R SAT (System Accuracy Test) temperature according to AMS2750E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 1
W Insulation requirements of BF type infant skin temperature probe IEC 60601 - Medical Electrical Equipment Safety Standards Series 0
B Temperature loop: partial calibration General Measurement Device and Calibration Topics 10
D ASTM C1064 Temperature Measuring Device General Measurement Device and Calibration Topics 2
C ETO Sterilised Class II Medical Device - Required Temperature Storage ISO 13485:2016 - Medical Device Quality Management Systems 1
M Microscope slide warmer temperature control General Measurement Device and Calibration Topics 4
S ISO 18223-2 Temperature symbol in the Legend table in the Instructions for Use Other ISO and International Standards and European Regulations 4
B ASTM F2924-14 /ASTM F3001 - Room temperature classification criteria Other US Medical Device Regulations 0
O Selection of a calibrator (pressure generator, temperature) for my field devices General Measurement Device and Calibration Topics 2
shimonv Symbol for temperature limit EU Medical Device Regulations 6
M Pharmaceuticals Warehouse temperature control (room temperature) Pharmaceuticals (21 CFR Part 210, 21 CFR Part 211 and related Regulations) 3
T 5ºC Storage Temperature Validation - Urology disposable devices EU Medical Device Regulations 3
Q Weigh Scale Electronics - Suggested Storage Humidity and Temperature Manufacturing and Related Processes 7
M CQI-9 monitoring of temperature in die Various Other Specifications, Standards, and related Requirements 2
C ISO 11607-1 Packaging System - Temperature and shipping conditioning Other Medical Device Related Standards 4
G Calibration Laboratory Dimensional Temperature Question General Measurement Device and Calibration Topics 2
V Temperature/humidity requirements in internal metrology laboratory General Measurement Device and Calibration Topics 3
P Qualification of a Temperature Controlled Unit have to be empty? Qualification and Validation (including 21 CFR Part 11) 5
D What is the exact Temperature Coefficient Value to be set on a Conductivity Meter? General Measurement Device and Calibration Topics 2
K Temperature Uniformity Survey (TUS) - Vacuum Furnace General Measurement Device and Calibration Topics 1
R Saudi Arabia - Temperature Storage Requirements Other Medical Device Regulations World-Wide 4
H Performing a TUS on a Furnace used at One Temperature - AMS2750-E AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 27
D Sterile Medical Device Temperature and Humidity (Storage) Labeling Requirements EU Medical Device Regulations 5
R Temperature Profile for Transport Simulation (Sterile Eye Drops) Other Medical Device Related Standards 1
N How to Convert Micro Volts into Temperature Uncertainty Measurement Uncertainty (MU) 1
N Calibrate temperature on refrigerated Centrifuges? General Measurement Device and Calibration Topics 2
C IQ and OQ for a "Dormitory" Refrigerator - Temperature Stability Qualification and Validation (including 21 CFR Part 11) 4
R Medical Device Storage Temperature Tests and Requirements EU Medical Device Regulations 14
R NADCAP AC7108 - Tank Ambient Room Temperature Requirements AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 2
S What is the service temperature of Polysurlyn Laminated Aluminium Alloy 3105? Manufacturing and Related Processes 2
C Saboraud's Agar Incubation Temperature and Validation Requirements Other Medical Device and Orthopedic Related Topics 10
N Load Thermocouple requirement for TUS (Temperature Uniformity Survey) AS9100, IAQG, NADCAP and Aerospace related Standards and Requirements 8
B Do I need to perform Gage R&R on a Digital Temperature & Humidity Gage Gage R&R (GR&R) and MSA (Measurement Systems Analysis) 18
B What is the safe Temperature Limit for Medical Device with blood contact ? IEC 60601 - Medical Electrical Equipment Safety Standards Series 3
S Paperless Temperature Recorder for Temperature Surveys on Heat Treatment Ovens General Measurement Device and Calibration Topics 6
M Daily Temperature and Humidity Graph template Document Control Systems, Procedures, Forms and Templates 2
A Calibrating a PRT sensor - Temperature Sensitivity Coefficient General Measurement Device and Calibration Topics 2
V Packaging as Medical Device (temperature sensitive products) ISO 13485:2016 - Medical Device Quality Management Systems 6
G Temperature Requirements for a Manufacturing Inspection Lab General Measurement Device and Calibration Topics 5
M Fixed Point Temperature Cells Calibration Method General Measurement Device and Calibration Topics 3
H Resistance Temperature Detectors (RTD) Manufacturer's Specifications General Measurement Device and Calibration Topics 3
M Calculating Coefficients for an RTD (Resistance Temperature Detector) Probe General Measurement Device and Calibration Topics 3